I am sending myself a little mad going round in circles here.
So.......we have come up against what I fear is a common issue; our contract manufacturers have decided that they fear the FDA too much and they don't wish to be listed with them anymore. They supply us with a class II device that has had 510K clearance for over two years but are yet to be audited.
The current contract manufacturer is happy to supply all of the components, so we would only need them finished, packaged and sterilised with new manufacturer (who is already listed with the FDA and has been audited many times).

I have three questions:

1) Am I right in thinking that in order to implement this change we would need to run new biocompatibility AND performance testing AND to submit a new 510K with the 'new device'?

2) We are concerned that there will now be a break in the supply chain whilst waiting for the above mentioned testing and 510K clearance; can our US distributor buy extra stock from us prior to the changes which can then be sold after the end of this year (if necessary) and before the new device has gained clearance? So our current contract manufacturer would not renew their listing for this next year but the device would have been manufactured whilst they were listed. Can the US distributor then sell the device in say January while we wait for the new product to gain clearance?

3) Have I mis-read the guidance and in fact we won't need a new 510K submission as the components of the device have remained the same?

Many thanks for taking the time to read this.

:) Katie
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
Few thoughts:
It seems you are moving the final stages of manufacturing from one approved supplier to a new one. Of course, there are a lot of details involved such as new supplier qualification, work instructions, process qualification, revise pFMEA, and repeat some relevant validation activities.
BUT in essence - if the product specifications remains unchanged, you may not need to submit a new 510(k) application.

Have a look at FDA guidance on "Deciding When to Submit a 510(k) for a Change to an Existing Device".

Good luck,
Shimon
 

katiesam

Starting to get Involved
#3
Thank you for your response, I had a look at the guidance but because we are keeping everything the same materials wise and it is the manufacturer that has changed, who is putting the pieces together, packaging and sterilising I am left a bit confused.

I am presuming that biocomp testing will need to be re-done as the product is being finished in a different location, the same for the packaging and shelf life?
 

Rincewind

Involved In Discussions
#4
I don't see why a new Biocomp would be needed if everything stays the same (it might be needed if there are other changes like different processing aids or such).
 

monoj mon

Involved In Discussions
#6
3) Have I mis-read the guidance and in fact we won't need a new 510K submission as the components of the device have remained the same?
This flowchart is from the FDA Guidance (of course I re-do it for my personal use where I put my justification beside each "No"), and if all the answers to this flowchart is "No" in your case, then you don't need to submit a new 510(k).
1595579844725.png
 

katiesam

Starting to get Involved
#7
This flowchart is from the FDA Guidance (of course I re-do it for my personal use where I put my justification beside each "No"), and if all the answers to this flowchart is "No" in your case, then you don't need to submit a new 510(k).
View attachment 27017
Thank you for your response. So a change in the manufacturer will also mean a change in the packaging and steriliser, which could affect the biocompatibility of the product, so surely that comes under flow chart B, B3.2? If the new packaging and steriliser are already used by FDA cleared devices and the SAL is lower etc then I can say 'No' to B3.2 and B4.1?
 
Thread starter Similar threads Forum Replies Date
M Licensing of a 510K to another manufacturer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
M Contract Manufacturer 510K Starting Point 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13
R FDA Inspections - Questions from a Device Manufacturer Submitting 510k Other US Medical Device Regulations 3
K FDA 510k electrosurgery 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J Leveraging FDA 510k Clearance for International Registrations Other Medical Device Regulations World-Wide 2
S Two or more predicates suitable? 510K submission US Food and Drug Administration (FDA) 5
S 510k: What to include if your own device is the predicate US Food and Drug Administration (FDA) 5
L 510K - Trade Name Change 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
Ed Panek Adverse Event Clinical Trial using a 510K approved Device Other US Medical Device Regulations 6
R Countries that accept 510k approvals - Class II nonsterile medical device Other Medical Device Regulations World-Wide 2
M Timing of 510k submission Other US Medical Device Regulations 9
S Examples of software changes that required a 510k US Food and Drug Administration (FDA) 2
D 510K and Changes to Verification and Validation US Food and Drug Administration (FDA) 2
Ronen E Can anyone please explain this? (510k submission requirement) Other US Medical Device Regulations 14
H New 510K VS Documents about design change. Other US Medical Device Regulations 1
P Medical Device not on commercial distribution but installed in the establishment. 510k? Other US Medical Device Regulations 10
V Presubmission FDA for 510k US Food and Drug Administration (FDA) 4
A How to deal with changed shared components in 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Can we take a part from 510k cleared medical device and use it in class I device? Other US Medical Device Regulations 3
D Contents of a 510k Q submission Other US Medical Device Regulations 1
L Class IIa Medical Device, CE Mark & 510k - Labeling Content CE Marking (Conformité Européene) / CB Scheme 4
Ed Panek Other company wants to use our FDA 510K bundled with their product 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
L 510k exemption process - Do you still need to submit technical dossier? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
S IEC 62133 - CB Scheme Required by FDA 510k Other Medical Device Related Standards 3
S The key differences in 510k types - Traditional vs. Abbreviated vs. Special US Food and Drug Administration (FDA) 0
P V&V for Special 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Clarity on requirements for a 510k exempt Class I device - Suture removal kit US Food and Drug Administration (FDA) 9
T Is 510k Anesthesia group particularly picky/difficult to work with during 510k? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J 510k - Product shelf life validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
K 510K statement Questions 21 CFR 807.93 - Discontinued or never released for sale US Food and Drug Administration (FDA) 0
J Applying for 510K - What is a significant process change and what is not US Food and Drug Administration (FDA) 5
K 510K Interactive Review 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
I Celsius and Fahrenheit in 510k US Food and Drug Administration (FDA) 4
R FDA 510K - Does the FDA accept the testing report from other countries? Other US Medical Device Regulations 1
I 510k Performance Testing Studies (Raw Data) - What does the FDA expect 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
P Is a different 510k is required for each (stand alone) software configurations? Other US Medical Device Regulations 3
K ISO 10993 Requirements For 510k - Substantial Equivalence Other US Medical Device Regulations 11
P Two separate 510k submissions for one device from the same company Other US Medical Device Regulations 10
I Special 510k Imported Legacy Blogs 0
K 510k "Premarket" Submission for Existing Class II Software Medical Device US Food and Drug Administration (FDA) 3
N Is a new 510K required for change of manufacturing location for an existing device? US Food and Drug Administration (FDA) 1
Q 510k Submission for a New Medical Device Product Other US Medical Device Regulations 3
W New 510k required for a material change? Other US Medical Device Regulations 9
M What if information in predicate 510k is obviously wrong? Other US Medical Device Regulations 5
E 'Special' 510k as a Predicate Medical Device? Other US Medical Device Regulations 3
A Does anyone have experience providing a summary of Pre-IDE discussions in the 510k US Food and Drug Administration (FDA) 1
J US import requirements for bench testing without 510k Other US Medical Device Regulations 3
4 Customer Making Changes to a 510k Device They Purchased 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R Finding specific 510k guidance for Keratoprosthesis (MLP) Other US Medical Device Regulations 1
R Why not use abbreviated 510k submissions? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2

Similar threads

Top Bottom