Good evening everyone,
I'm currently tackling a non-conformity raised by our Notified Body and it pertains IEC 60601-1-8 so I'd really appreciate any insight or shared experience from you...
Our medical device includes visual indicators (LED lights) that illuminates in the event of a fault. The device's design relies on a fail-safe architecture, so that whenever a hazardous or abnormal condition is detected, the system automatically shuts down and enters a safe state, independently of user intervention.
In our Technical Documentation, we initally proposed a rationale that these lights did not constitute an alarm system according to the standard since no user response is required to reset a condition as they were mere visual indicators that serve to inform rather than alert in a formalised sense.
Our NB disagreed and pushed it back, since according to our Risk Analysis, these luminous indicators (plus the error message pops up in user interface) were in fact risk controls.
My first instinct was eliminating any references to these indicators in my Risk Analysis thus confirming the IEC's non-applicability (though I doubt it will be acceptable).
The alternative is mapping all possible warnings and relative "alarm conditions, classifying them by alarm type and priority, and specify the required actions - treating them as formal alarms in line with the standard's definition. Will this be sufficient?
At this point, what I'm trying to comprehend is what specific evidence or structure does my NB expect when evaluating conformity to IEC 60601-1-8. Has anyone used a checklist or structured template for evaluating conformity?
I thank you in advance for any opinions, strategies or tools
I'm currently tackling a non-conformity raised by our Notified Body and it pertains IEC 60601-1-8 so I'd really appreciate any insight or shared experience from you...
Our medical device includes visual indicators (LED lights) that illuminates in the event of a fault. The device's design relies on a fail-safe architecture, so that whenever a hazardous or abnormal condition is detected, the system automatically shuts down and enters a safe state, independently of user intervention.
In our Technical Documentation, we initally proposed a rationale that these lights did not constitute an alarm system according to the standard since no user response is required to reset a condition as they were mere visual indicators that serve to inform rather than alert in a formalised sense.
Our NB disagreed and pushed it back, since according to our Risk Analysis, these luminous indicators (plus the error message pops up in user interface) were in fact risk controls.
My first instinct was eliminating any references to these indicators in my Risk Analysis thus confirming the IEC's non-applicability (though I doubt it will be acceptable).
The alternative is mapping all possible warnings and relative "alarm conditions, classifying them by alarm type and priority, and specify the required actions - treating them as formal alarms in line with the standard's definition. Will this be sufficient?
At this point, what I'm trying to comprehend is what specific evidence or structure does my NB expect when evaluating conformity to IEC 60601-1-8. Has anyone used a checklist or structured template for evaluating conformity?
I thank you in advance for any opinions, strategies or tools