60601-2-10 clause 201.12.4.103 Output Indicator (itself in SFC?)

Loekje

Involved In Discussions
Hi Forum,

The clause states that the device must indicate with a yellow lamp when it can produce more than 10mA, 10V or 10mJ over 1k, in normal and single fault condition.

Does this mean the designer has to either use a LED with high integrity characteristics (plus the rest of the steering electronics) or double the components that can have the light itself fail?
Or is it enough to have the circuit go though a Power On Self Test in which the operator must verify that this LED indicates during the appropriate self test?

As I never encountered this indicator doubled my guess is either the latter or this clause is not taken too strict?

Kind regards,
Loek
 

Peter Selvey

Leader
Super Moderator
My guess is that the authors of the standard were thinking of faults resulting in unexpected output rather than faults in the indicator. However, strictly speaking faults in the indicator should also be considered.

In general an indicator would be unstressed modern electronics which has a very low failure rate, the fault in the indicator would have to be combined with either a user error (turned the output on unintentionally) or a second electrical fault, and the unplanned output would then need to flow through dangerous path (e.g. through the heart) and even then there is a low probability of a serious outcome (e.g. heart actually stops). So it may be that overall the risk is acceptable without any risk control measures. At the same time, fairly low cost solutions exist (e.g. have the MPU monitor the current through the LED, use two LEDs) so it's not crazy to use a risk control either. It's a grey zone item!
 

Loekje

Involved In Discussions
Thanks for your reply Peter,

Gray zones are the place to go for asking agreement up front with your test house: "Do you agree with our solution on how to implement this clause?".
(and of course you start off with the implementation that suits you best)
 

Peter Selvey

Leader
Super Moderator
Ignoring the test house angle, this is also probably a good place to test the "risk minimum theory".

This theory places less emphasis on the absolute value of risk (which is near impossible to measure or establish criteria) and more on the impact of resources, and in particular if the resources is small then just do it, don't argue the point.

So for example, instead of asking if a 13 meter high tower is acceptable risk for a backup generator for the Fukushima nuclear power plant, we should first ask the question "at what height does the risk become negligible?". If, for example, 15m (yes, just an extra 2m) makes it a once in a million year event, then why waste time on endless discussions about whether 13m tower estimated to be ok for everything except 1 in a 1000 year event is OK, whether the benefit to society justifies the risk, blah blah blah. Just build a 15m tower and be done with it.

So, coming back down to earth and looking at the original case study, the cost (resources) of the electronics to monitor the LED current is so small, it is really not worth the discussion. Just do it.

Unfortunately, our preponderance towards estimating risk and comparing against acceptable criteria (ala ISO 14971) encourages avoidance of low cost solutions that virtually eliminates risk. It will almost certainly be a case study in the future for how human beings can be silly. For decades.
 
Thread starter Similar threads Forum Replies Date
N 60601-2-2 Ed 6 Clause 201.8.8.3.102 - HF instrument leakage IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R How to calculate the Air Flow Rate in clause 201.12.4.4.105 of IEC 60601-2-16:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
I Dielectric Test for HF Surgical Equipment (IEC 60601-2-2 clause 201.8.5.1.2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1-8 clause 6.4.1 (statistics of the distribution of the sum of the maximum ALARM CONDITION + SIGNAL GENERATION DELAY) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601-1 Clause 7.9.3.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Clause 8.9.1.2 of AMD2:2020 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
W IEC 60601-1 clause 8.8.3 - Dielectric Strength IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
B IEC 60601-1 Clause 14.8 - Architecture IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A 60601-1 clause 8.11.3.6 Test Clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
K What is mean by Oxygen Rich Environment as per the IEC 60601-1 clause no 11.2.2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
S ME system clause 16 of IEC 60601-1, 3.1 ed IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T What is PESS? Cited in Edition 3.1 of IEC 60601, clause 14.1 note 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E ME (Medical Equipment) Systems - IEC 60601-1 Clause 16.1 Interpretation IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Question on IEC 60601 clause 15.5.1.1 (transformer overheating) IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L IEC 60601-1 3rd edition Clause 9.2.2 Trapping Zone (Acceptable Gaps) IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E IEC 60601-1 Clause 14 Questions IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R How to understand the exception of IEC 60601 sub-clause 8.11.5 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M MEEE Expected Service Life - IEC 60601-1 Clause 4.4 Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P IEC 60601-1 Clause 11.1.3 Winding Temperature Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J EMC (Electromagnetic Compatibility) Requirements - IEC 60601 3rd edition Clause 17 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601 Clause 7.1.3 - Durability of Markings - Methylated Spirits IEC 60601 - Medical Electrical Equipment Safety Standards Series 19
P Understanding IEC 60601-1:2005, Clause 8.1 b) - Short Circuiting Double Insulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
P IEC 60601 Clause 57.3 / 8.11.3.2 - PVC Flexible Cords IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Single Fault Condition for 3-Phase Medical System - IEC 60601 Clause 8.1b IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
R Can manufacturer separately evaluate the clause 14 (PEMS) of 60601-1: 2005? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B IEC 60601-1 2nd Clause 17 and 3rd edition - Electrical Isolation IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
G Test Plan - Pass/Fail Criteria for TR 60601-4-2 Other Medical Device Related Standards 0
H 60601 - Patient leakage current via the patient connection test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
F Defining the need to perform the Proximity magnetic fields immunity tests (IEC 61000-4-39) according to the IEC 60601-1-2 ed.4.1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O ESD Testing IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
Q Interpretation of "external part of implant" in 60601-1 in relation to EN-45502 IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E Meaning of ALARM PRESETS (subclause 6.5.1 of IEC 60601-1-8:2020) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Franky Zhang123 The Electrical Standards GB9706.1:2020 (same as IEC 60601-1:2012) etc. will be implemented on May 1st, 2023 in China. China Medical Device Regulations 0
A Where 60601-1 represented in the 510k submission? US Medical Device Regulations 6
H IEC 62368-1 used in IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Qualify EEG surface electrodes as AP acc 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T Is TR 60601-4-2 mandatory? IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
E Artificial hand (IEC 60601-1-2) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Differences between IEC 60601-1-2 and IEC TR 60601-4-2 US Food and Drug Administration (FDA) 3
S PDF version of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
P IEC 60601-2-25 requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
P Application of 60601-1 and 60601-2-2 to a hand-held laparoscopic unstrument with motorisation in the handle IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Why is Protective Earth Resistance required <0.1ohm or 0.2ohm in 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
I MOPP required for SIP when considering Single Fault Condition (SFC) in 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
J Environmental conditions 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom