60601-2-2 Section 50.2 Accuracy of Controls and Instruments - Help needed

Elsmar Forum Sponsor

Marcelo

Inactive Registered Visitor
#12
Is't the above mentioned example wrong?
60W setting should then be within 48-72W?
Yes, it´s wrong and nobody noticed, not even me :p

It should be "For example, if the diagram shows that at the 50 ohms control setting and with X load resistance the output would be 60 Watts, when you measure it according to the standard the real output has to be between 40-72 Watts."

40-72 Watts is +- 20% of 60 W.

Hope they did understand back them though.

Thanks for this!
 
Thread starter Similar threads Forum Replies Date
R IEC 60601-1 - Power Supply Cords (Section 8.11.3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J IEC 60601 Third Edition Labeling Requirements - Section 5.2.2.1 (g) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Z IEC 60601 Testing - Section 13.2.7 - Impairment of Cooling IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
E Accredited vs. non-accredited labs for 60601 compliance in the US IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Accredited vs. non-accredited labs for 60601 compliance in the US Other Medical Device Related Standards 0
M Class II type machine , and its compliance with 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
Ed Panek 60601-1-2:2015/A1:2021 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E 60601-1 - Tilt testing - Tensile safety factor IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
JoCam IEC 60601-1 and 60601-1-2 retest after PCBA change IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 60601-1:2005/AMD2:2020, Why this standard version is 3.0? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A 60601-1 : Integrated Dry ECG Electrodes = 2 Patient connections inside 1 applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A 60601-1-2 : Integrated Dry Electrodes = PATIENT COUPLED PORT? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N 60601-2-2 Ed 6 Clause 201.8.8.3.102 - HF instrument leakage IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
N Maximum mains voltage used when applying 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J IEC 60601-1-11 Home Class II With Ballasts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Coverage and differences: EN 60601-1:2006+A12:2014 Vs AAMI/IEC 60601-1:2005+AMD1:2012 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
T IEC 60601-1-8:2020 Is it necessary to change the alarm melody? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
S Applicability of 60601 vs. 60335 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Is it normal / sufficient to have only the IEC 60601-1-2 test report without indicating IEC 60601-1? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-2-30 - Is it mandatory to claim alarms? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Y Auditory alarm standard IEC 60601-1-8 Reliability Analysis - Predictions, Testing and Standards 0
A 60601-1 understanding of stationary device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Complex IEC 60601-1 gap assessment IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M How does IEC-60601-1 apply to a non-medical device in the patient vicinity? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R IEC 60601-1 - 11.1.3 e) Test criteria - Temperature Measurements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R IEC 60601-1 - Magnesium oxide used for the electrical insulation of heating elements IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Is IEC 60601-1-2 required by FDA for all electronic medical devices? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
Z IEC 60601-2-25; Frequency response test Medical Device and FDA Regulations and Standards News 1
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A ANSI/AAMI versions of 60601-1-2 and related testing requirements Other Medical Device Related Standards 3
R IEC 60601-1:2005+AMD1:2012+AMD2:2020 CSV IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B IEC 60601 - Creepage Distance - Relay that acts as a means of physical mechanical protection Process Maps, Process Mapping and Turtle Diagrams 0
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M What to Expect from Next IEC 60601-1 and IEC 60601-1-2 Amendments? IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
D IEC 60601-1 - Service life testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
R Hand transmitted vibration 9.6.3 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A IEC 60601 11.2.2.1 Risk of Fire in an Oxygen Rich Environment, Source of Ignition IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
shimonv Minimum 60601 requirements for a clinical trial IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E PEMS Hazards - IEC 60601 Clause 14.6 - Internal data use - Pressure sensor IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W IEC 60601 - Essential performance c.2.34 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B IEC 60601-2-43 - Clause 203.6.103 - Physical button? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M IEC 60601-1 1988 - Device developed in 2012 with standard of 1988 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom