The medical device company I work for uses Jira to develop, and maintain software used in the device. In terms of complying to IEC 62304, there needs to be code review, unit tests and integration tests along with requirements verification tests.
For the last one, we create a protocol with test cases, acceptance criteria, test results etc. But for the first three ( code, unit and integration) testing, how does one document them from Jira? Do you keep them within Jira and create a summary report within a word document to comply with IEC 62304? Or do you actually create a word document that identifies unit, test coverage ..... If so, would anyone be willing to share an example/template for these?
TIA
Elsmar Forum Sponsor
Tidge
Trusted Information Resource
#2
I have only used JIRA for issue tracking (and resolution), and not project management (of deliverables). My teams have treated the test deliverables (code review, unit testing, integration testing) as stand-alone documents that get approved and then executed.
It should be possible to configure JIRA to manage these sorts of activities, but I haven't personally witnessed JIRA being used in a diligent enough way that would do this:
  • record both pre-approval of test goals (with traceability) and post-execution review/approval of actual test results, and then a report summarizing results
I seem to recall that a default configuration for JIRA "sort of" has places were these hooks could exist, but there is (from my PoV) a lot of overhead that would have to be entered in each JIRA item, such that I'm not sure that this would be practical.
This may be just because I am somewhat set in my ways, but here goes: In my mind, JIRA is for "one-and-done" tasks, whereas formal testing is the sort of activity that can be revised... because of new implementations, new test methods, bug fixes, test setups, etc. JIRA tasks start to look really sloppy when things go wrong. This is fine for development, but when generating objective evidence to support 62304 required activities I think it would become a nightmare to untangle.
mihzago
#3
You can attach things in JIRA, so you could add the protocol/results of each feature review/testing to the JIRA issue. If you don't want to clutter your JIRA issues, then you could create tasks where you could keep track and record testing activities.
Creating a summary report is also a good idea; at least that's what do in some cases.
You could also explore integrations with Confluence and Bitbucket/Github, or test case management tools (Using JIRA Software for Test Case Management | Jira | Atlassian Documentation)
Thread starter Similar threads Forum Replies Date
K IEC 62304 compliance - Code reviews as part of verification strategy IEC 62304 - Medical Device Software Life Cycle Processes 5
W CPU BIST IEC 62304 - Embedded code has CPU instruction tests IEC 62304 - Medical Device Software Life Cycle Processes 2
L EN 62304:2006/AC:2008 vs IEC 62304:2006/Amd 1:2015 IEC 62304 - Medical Device Software Life Cycle Processes 1
MaHoDie Is it possible to assign medical software to security class A (according to IEC 62304)? EU Medical Device Regulations 8
T FDA level of concern vs IEC 62304 safety classification - 510(k) exempt device Other US Medical Device Regulations 13
Q Artificial Intelligence/Machine Learning and 62304 IEC 62304 - Medical Device Software Life Cycle Processes 7
Y Software to Manage IEC 62304 Traceability Requirement IEC 62304 - Medical Device Software Life Cycle Processes 3
B Should I buy IEC 62304:2006, IEC 62304:2006/AMD 1:2015 or both? EU Medical Device Regulations 1
L ISO 62304 IEC 62304 - Medical Device Software Life Cycle Processes 4
M How to apply IEC 62304 Problem Resolution Process for bugs discovered prior to release i.e. during initial development? IEC 62304 - Medical Device Software Life Cycle Processes 2
T IEC 62304 : Risk control for SaMD IEC 62304 - Medical Device Software Life Cycle Processes 8
F IEC 62304 agile development EU Medical Device Regulations 1
shimonv Working with a software developer who is not setup for IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
P Examples of quality plans in IEC 62304 US Food and Drug Administration (FDA) 2
E Test report to certify compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 4
G Adopting old product - compliance with IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 12
A IEC 62304 safety classification, External Controls and off-label use related risks IEC 62304 - Medical Device Software Life Cycle Processes 5
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 9
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
K IEC 62304 - Testing Independance IEC 62304 - Medical Device Software Life Cycle Processes 5
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2
M Risk Analysis Flow - Confusion between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
D BS EN 62304 - Medical-Relevant Data C.5 - Definition of IEC 62304 - Medical Device Software Life Cycle Processes 5
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M IEC 62304 Class A Project IEC 62304 - Medical Device Software Life Cycle Processes 15
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
P SOUP anomaly evaluation for MMA (Mobile Medical Application) IEC 62304 clause 7.1.3 IEC 62304 - Medical Device Software Life Cycle Processes 6
P IEC 62304 - evaluation of integration and system testing IEC 62304 - Medical Device Software Life Cycle Processes 4
P Risk acceptability alignment between ISO 14971 and IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 6
D Required Checklist Showing Compliance to IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 11
P Proposed revision of IEC 62304 - 2019 IEC 62304 - Medical Device Software Life Cycle Processes 6
S Relationship between IEC 62304 problem resolution and ISO 13485 IEC 62304 - Medical Device Software Life Cycle Processes 8
P IEC 62304:2006 A1:2015 - Software from the early 1990s IEC 62304 - Medical Device Software Life Cycle Processes 4
B IEC 62304:2015 vs IEC 62304:2006 + AMD1 IEC 62304 - Medical Device Software Life Cycle Processes 4
F IEC 62304 - Segregation and communication between software items IEC 62304 - Medical Device Software Life Cycle Processes 1
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
B IEC 62304 - Update Checklist IEC 62304 - Medical Device Software Life Cycle Processes 2
L Connection between IEC 62304 and Chapter 14 of IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Implementing EN 62304:2006 on existing and proven medical devices IEC 62304 - Medical Device Software Life Cycle Processes 6
M IEC 62304 - Develop an Architecture for the Interfaces of Software Items IEC 62304 - Medical Device Software Life Cycle Processes 8
S Does IEC 62304 require documenting unresolved anomalies for all safety classes? IEC 62304 - Medical Device Software Life Cycle Processes 4
JoshuaFroud Design and Development of Software under 13485:2016 or 62304? ISO 13485:2016 - Medical Device Quality Management Systems 6
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
T I need to make test reports according IEC 62304 & IEC 62366 IEC 62366 - Medical Device Usability Engineering 2
D Changing software classification via software - IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 3
D Software as risk control - Confused on one aspect of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 20
K Trying to figure out what satisfies a few aspects of IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 2
Similar threads
Top Bottom