7.6 Control of Monitoring and Measuring Equipment vs. 7.6.1 Measurement Systems Analy

G

GoKats78

#1
Please share your thoughts and insight concerning the following ISO/TS 16949:2009 situation.

Our TS Registration auditor says that ALL measuring and test equipment systems listed in a Process Control Plan must be calibrated and have Statistical studies conducted to analyze the variation present in the results because of what is said in 7.6.1.

7.6.1 Measurement systems analysis

Statistical studies shall be conducted to analyze the variation present in the results of each type of measuring and test equipment system. This requirement shall apply to measurement systems referenced in the control plan. The analytical methods and acceptance criteria used shall conform to those in customer reference manuals on measurement systems analysis. Other analytical methods and acceptance criteria may be used if approved by the customer.

Example: Pyrometer used to measure temperature, which is an internal Process requirement that could potentially affect a product requirement (Mechanical Properties) that is checked prior to shipment to the customer.

-------------------------------------------------------------------------------

I believe that this only applies to measurement devices that are used to provide evidence of conformity of product to determined requirements because of what is said in 7.6.

7.6 Control of monitoring and measuring equipment

The organization shall determine the monitoring and measurement to be undertaken and the monitoring and measuring equipment needed to provide evidence of conformity of product to determined requirements.

Example: Instron Robotic Mechanical Testing Machine used to measure Mechanical Properties (yield, tensile strength, etc.), which is a Product requirement of our customers.

Situations like this will only make people take things out of their Process Control Plans that should be in there for all the right reasons.

Please advise. Thank You.
 
Elsmar Forum Sponsor
#2
Re: 7.6 Control of Monitoring and Measuring Equipment vs. 7.6.1 Measurement Systems A

In a way we are talking apples and oranges... or perhaps apples and bushels of apples.

Calibration relates to specific measuring devices, whereas MSA relates to family or type of device. For example, if you use micrometers, you need to perform an analysis on micrometers, but not on each individual mic. On the other hand, each individual mic used will be required to be calibrated (or verified).
 
G

GoKats78

#3
Re: 7.6 Control of Monitoring and Measuring Equipment vs. 7.6.1 Measurement Systems A

That really doesn't address the concern..we have items (such as scales and thermo-couples) called out in the control plan..they are either calibrated and certified (as per 7.6) . The scales measure raw product ....we have never done any MSA on them as all product characteristics are checked later (via chemical and mechanical properties testing). we have never done any MSA on the thermocouples...but our auditor is stating that it is required..

This guy seems hung up on these two clauses and there interaction...
 

Howard Atkins

Forum Administrator
Staff member
Admin
#4
Re: 7.6 Control of Monitoring and Measuring Equipment vs. 7.6.1 Measurement Systems A

There are 2 separate requirements which are not dependent on one another.

7.6.1 is unquestionable "This requirement shall apply to measurement systems referenced in the control plan. " there is NO place for discussion

The requirements of 7.6
The organization shall determine the monitoring and measurement to be undertaken and the monitoring and measuring equipment needed to provide evidence of conformity of product to determined requirements.
The organization shall establish processes to ensure that monitoring and measurement can be carried out and are carried out in a manner that is consistent with the monitoring and measurement requirements.
Please note that there is also a requirement for "monitoring and measurement can be carried out"
Process control is a requirement of TS and a part of the control plan

Annex A (normative) Control plan

c) Process control ⎯ process parameters,
⎯ process-related special characteristics,
⎯ machines, jigs, fixtures, tools for manufacturing.
In regard to issues such as thermocouples they should be verified and linearity and bias studied as required according to their use.
MSA does not mean GRR specifically or at all.

The philosophy of TS is prevention not detection so the performance of thermoucouples correctly will prevent errors and save corrections
 
G

GoKats78

#7
Re: 7.6 Control of Monitoring and Measuring Equipment vs. 7.6.1 Measurement Systems A

Good point...that is what happens when you try to do things from memory....and for someone else!
 

qusys

Trusted Information Resource
#8
Re: 7.6 Control of Monitoring and Measuring Equipment vs. 7.6.1 Measurement Systems A

Please share your thoughts and insight concerning the following ISO/TS 16949:2009 situation.

Our TS Registration auditor says that ALL measuring and test equipment systems listed in a Process Control Plan must be calibrated and have Statistical studies conducted to analyze the variation present in the results because of what is said in 7.6.1.

7.6.1 Measurement systems analysis

Statistical studies shall be conducted to analyze the variation present in the results of each type of measuring and test equipment system. This requirement shall apply to measurement systems referenced in the control plan. The analytical methods and acceptance criteria used shall conform to those in customer reference manuals on measurement systems analysis. Other analytical methods and acceptance criteria may be used if approved by the customer.

Example: Pyrometer used to measure temperature, which is an internal Process requirement that could potentially affect a product requirement (Mechanical Properties) that is checked prior to shipment to the customer.

-------------------------------------------------------------------------------

I believe that this only applies to measurement devices that are used to provide evidence of conformity of product to determined requirements because of what is said in 7.6.

7.6 Control of monitoring and measuring equipment

The organization shall determine the monitoring and measurement to be undertaken and the monitoring and measuring equipment needed to provide evidence of conformity of product to determined requirements.

Example: Instron Robotic Mechanical Testing Machine used to measure Mechanical Properties (yield, tensile strength, etc.), which is a Product requirement of our customers.

Situations like this will only make people take things out of their Process Control Plans that should be in there for all the right reasons.

Please advise. Thank You.
Did you sign a PPAP with your customer?
If so, you should have agreed with it the reference manual for the analytical studies or alternative methods for each measurement or test systems recalled in the control plan ( form incoming to shipping through production):bigwave:
 

Howard Atkins

Forum Administrator
Staff member
Admin
#9
Re: 7.6 Control of Monitoring and Measuring Equipment vs. 7.6.1 Measurement Systems A

Did you sign a PPAP with your customer?
If so, you should have agreed with it the reference manual for the analytical studies or alternative methods for each measurement or test systems recalled in the control plan ( form incoming to shipping through production):bigwave:
PPAP is irrelevant this is a requirement of TS that the customer cannot waive.
There must be a MSA study- it does not have to be to a specific PPAP part
The PPAP handbook says
2.2.8 Measurement System Analysis Studies
The organization shall have applicable Measurement System Analysis studies, e.g., gage R&R, bias,
linearity, stability, for all new or modified gages, measurement, and test equipment. (see the Measurement Systems Analysis reference manual).
This is less than the standard!:agree1: which requires
"This requirement shall apply to measurement systems referenced in the control plan. "
This has no relevance to the requirements of PPAP
 
P

pldey42

#10
Re: 7.6 Control of Monitoring and Measuring Equipment vs. 7.6.1 Measurement Systems A

It seems to me that the answer lies in the requirement, and an understanding of what quality management means.

7.6.1 Measurement systems analysis

Statistical studies shall be conducted to analyze the variation present in the results of each type of measuring and test equipment system. This requirement shall apply to measurement systems referenced in the control plan.

Classical quality management principles include Deming's idea that variation is bad and should be reduced. While defects introduced by measurements system variation might (not will, might) be detected in product test and inspection, there's an expense in rejecting or reworking them – an expense that's passed on to the customer and makes the business less competitive. (Also, if product tests and inspections are performed on a sampling basis there's surely a risk of passing on some defective product to the customer.)

So it seems to me that the requirement means what it says – reduce or eliminate variation that might be introduced by all the measurements called up in the control plan using MSA. As Howard observes, the idea is to proactively prevent defective product from being made, because it's cheaper to do it right than to do it wrong and then scrap, rework or accidentally deliver it.

If that means that people will omit items from the control plan it seems to me they need an education in why they're doing control plans, and specifically the benefits in terms of reduced costs, scrap and rework of reducing measurement systems variation. And then an education in MSA ...

Hope this helps,
Pat
 
Thread starter Similar threads Forum Replies Date
E ISO 13485 7.6 Control of monitoring and measuring equipment - Assess the Validity ISO 13485:2016 - Medical Device Quality Management Systems 4
A Calibration Interval - AS9100 - Control of Monitoring and Measuring Equipment Calibration Frequency (Interval) 8
D Does 7.6 Control of Monitoring and Measuring Equipment align with ISO 17025 ISO 17025 related Discussions 4
G Control of Monitoring and Measuring Equipment for Servicing Photocopiers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
S ISO9001 Exclusion 7.6 Control of Monitoring and Measuring Equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
M 7.6 Control of Monitoring and Measuring Equipment - Monitoring vs. Measuring IATF 16949 - Automotive Quality Systems Standard 2
B Clarification on 7.6 Control of Monitoring and Measuring Devices Note AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
M Effective Control of Monitoring and Measuring Devices - Unused Devices AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
J Compliance with ISO 9001 Clause 7.6 Control of Monitoring & Measuring Devices ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
D Control of Measuring and Monitoring Instruments General Measurement Device and Calibration Topics 5
D Control of monitoring and measuring equipment - infrared thermometer for cross-check ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
R ISO 13485 Clause 7.6 Control of Monitoring and measuring devices - Requirements ISO 13485:2016 - Medical Device Quality Management Systems 10
W Computer Software - ISO 9001 Clause 7.6 Control of Monitoring and Measuring equipment ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Control of monitoring and measuring devices - Computer forensics company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
L Control of Monitoring and Measuring Devices - Clause 7.6 Exclusion? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
A Suggestions for: Control of Monitoring and Measuring Devices? (see 7.2.1) (1) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 33
3 Calibration Frequency - Control of Monitoring and Measuring Devices - Clause 7.6 Calibration Frequency (Interval) 14
R TS 16949 Clause 7.6 Control of Monitoring and Measuring Devices IATF 16949 - Automotive Quality Systems Standard 12
A Outsourcing Control of Monitoring and Measuring Devices ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
A Control of Monitoring & Measuring Devices ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
R 7.6 Control of Monitoring and Measuring Devices IATF 16949 - Automotive Quality Systems Standard 2
M Calibrate Computers - ISO 9001 Clause 7.6. Control of Monitoring & Measuring Devices ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 43
R 7.6 Control of Monitoring and Measuring Devices in the Service Industry Service Industry Specific Topics 12
E Control of Monitoring and Measuring Device Software General Measurement Device and Calibration Topics 4
C Monitoring and Control Instruments RoHS, REACH, ELV, IMDS and Restricted Substances 3
T IATF 16949 Control Plan and Process Monitoring Requirement(s) IATF 16949 - Automotive Quality Systems Standard 1
H ISO 9001:2008 Clause 7.6 Control of Monitoring and Measurement (Computer Software) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
M Monitoring and Control of Utilities and Supplies (7.5.1) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Q Monthly Data Review through Control Charts - Monitoring Complaints ISO 13485:2016 - Medical Device Quality Management Systems 3
N Medical Device External Standards Control and Monitoring Other US Medical Device Regulations 10
P Control of Monitoring and Measurement Devices - ISO 13485 Clause 7.6 says ... ISO 13485:2016 - Medical Device Quality Management Systems 3
R Pest Control and Monitoring Required? ISO 13485:2016 - Medical Device Quality Management Systems 7
R Control of Monitoring & Measurement Equipment - ISO 9001:2008, Clause 7.6 c) General Measurement Device and Calibration Topics 9
C Monitoring changes in ANSI standards - Control of External Documents Document Control Systems, Procedures, Forms and Templates 6
R Deciding what Processes need Control and Monitoring - 820.70 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
A Monitoring Out-of-control situations on X-bar and R chart Statistical Analysis Tools, Techniques and SPC 15
D 7.5.1.1 - Controls vs. Methods for Monitoring of Control IATF 16949 - Automotive Quality Systems Standard 2
T Control of Monitoring Device - Workers (employees) eyes - Visual Inspection ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
M Control of Monitoring and Measurment Manuals Miscellaneous Environmental Standards and EMS Related Discussions 1
E What are monitoring devices? ISO 9001 Clause 7.6 - Control of ... monitoring devices ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
I Document Control Software Document Control Systems, Procedures, Forms and Templates 2
T Controlling Expandable Forms in Paper-Based Document Control System Document Control Systems, Procedures, Forms and Templates 10
N ISO 13485 7.3.9 Change control in medical device software ISO 13485:2016 - Medical Device Quality Management Systems 6
C UDI Questions on Control Units Other US Medical Device Regulations 6
A Quality Control Datasheets Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
M Control-self assessment Internal Auditing 5
GreatNate Master Control QMS software Quality Tools, Improvement and Analysis 0
A Prototype control plan FMEA and Control Plans 2
J Records Control - Does each individual record need to be numbered? Records and Data - Quality, Legal and Other Evidence 2
shimonv Document Control Procedure Header Content Document Control Systems, Procedures, Forms and Templates 3