The whole "as appropriate" approach holds no water unless it's actually defined somewhere.
Thanks for the feedback. That's a good question regarding how "as appropriate" is defined, per ISO13485:
When a requirement is qualified by the phrase “as appropriate”, it is deemed to be appropriate unless the organization can justify otherwise. A requirement is considered appropriate if it is necessary for: — product to meet requirements; — compliance with applicable regulatory requirements; — the organization to carry out corrective action; — the organization to manage risks
As others have mentioned, this is required for traceability in the event of an issue with the equipment, so the "carry out correct actions" carries a lot of weight here. This is probably a losing battle trying to justify not recoding the equipment used for product acceptance testing.
I am interpreting the 'appropriateness' as deriving from the fact that a DMR required some measurement/test. If the use of a gauge is NOT required, and someone (during manufacture) decided to use one, then that would be IMO a case where there would be not be expected to find a record of the use of that gauge on the DHR.
That does make sense, thank you. So for final product acceptance we'd definitely need to record the equipment ID. For in process inspections that are indicators or repeated on the end product, there might be some ground to justify not including the tool ID?
Any thoughts on Incoming inspection? I guess that depends on the nature of what's being measured, and whether we could show that the measurements are redundant to other inspections we/our suppliers perform.
That might be more work than just recording the tool.