I have a Mitutoyo SV-622 profilometer that we currently calibrate with a standard 3 patch precision reference specimen. The lowest patch range is 16 micro inches. I’ve just been asked to acquire an 8 micro inch specimen for calibration purposes due to a blueprint callout on some manufactured parts. It is my understanding that it is not wise to use a specimen of such small nominal value due to the higher percentage of uncertainty it would generate.

The question I’m being asked by the end user is how would I know that the profilometer is giving accurate readings below 16 micro inches if it is not calibrated below that range.

Any help would be greatly appreciated.

-SAM
 
Elsmar Forum Sponsor
H

Hodgepodge

#2
I have a Mitutoyo SV-622 profilometer that we currently calibrate with a standard 3 patch precision reference specimen. The lowest patch range is 16 micro inches. I’ve just been asked to acquire an 8 micro inch specimen for calibration purposes due to a blueprint callout on some manufactured parts. It is my understanding that it is not wise to use a specimen of such small nominal value due to the higher percentage of uncertainty it would generate.

The question I’m being asked by the end user is how would I know that the profilometer is giving accurate readings below 16 micro inches if it is not calibrated below that range.

Any help would be greatly appreciated.

-SAM
Sam,
I also use a Mitutoyo profilometer, though not the same model as yours. The user manual includes a calibration procedure. In the procedure, only one reference standard is required for calibration. That would mean if there is no discrepancy between the reading and the known value of the roughness specimen, then the device will be accurate over the entire range of measurement.

Though this truly isn't adequate, it does say something for the confidence the manufacturer has in its equipment. What you will need is at least two roughness specimens with different roughness values (which you have) that have been calibrated by an accredited laboratory. Take a measurement on one of the specimens, adjust the gain if necessary. Then check the second specimen. If the profilometer returns the correct values on both specimens, then it is accurate over its operating range. You will be safe measuring well below an 8 micro.

On a related note: I once could not get our profilometer to accurately read both specimens, turned out the stylus on the detector had worn out and needed to be replaced.
 

Paul F. Jackson

Quite Involved in Discussions
#3
SAM135,

Measuring surface roughness measurements of very smooth surfaces is difficult. In order to get accurate results for surface finishes below 10 micro inches or 0.25 microns the stylus tip has to be in nearly brand new condition. If the stylus is slightly worn the profilometer measurements will underestimate the roughness of very smooth surfaces.

Many years ago I requested a deviation approval on behalf of a prototype supplier for the surface roughness of centerless ground pinion shafts. They measured 8 micro inches where 6 was required in the specification. The product engineer refused the SREA but allowed the supplier to polish to achieve the specified finish. I asked if current production shafts were polished and he said no. He stated that the production supplier was able to achieve the specified finish via grind processing only. He claimed he had data that confirmed that function deteriorated when initial finish exceeded 8.

As it happened I reviewed prototypes ordered from the production supplier not more than week later and armed with my brand new little Taylor Hobson portable Surftronic analyzer I discovered that regular production measuring 6 micro inches on the production profilometers actually measured 10 on mine. Needless to say my ah-ha caused a ruckus. We went to their lab to find calibration test blocks to settle the difference. While there I noticed a very high-end proficorder, they claimed was seldom used and that no one present had ever operated it. I assured them that I was familiar with it, we set it up, measured, and corroborated the 10 micro inch findings. All three devices including the production instrument with the worn stylus checked OK for value and linearity on the calibration test blocks.

The test blocks that come with the equipment that are used to calibrate gain and linearity have a lower value that is insensitive to slight stylus wear (I think approximately 25 micro inches) and an upper value sufficiently greater to set linearity without exceeding typically measured finishes (I think approximately 100 micro inches). There is standard that you can buy to check the condition of the stylus (I think approximately 10 micro inches) but to distinguish a 5 from a 7 the stylus has to be nearly new.

Paul
 
S

SAM135

#4
Thanks guys for the useful information.

I’m one of the guys that have no actual certified metrology training and must learn from books or asking questions as I go along. My employer won’t send me out for any training. Most of the calibration is cut and dry as I’m following written calibration instructions but some of it is vague and or leaves too many open ended questions.

I’m glad I found this site. I’ll be back for sure…
 
Thread starter Similar threads Forum Replies Date
L Micro-Vu InSpec Software Program Qualification and Validation (including 21 CFR Part 11) 6
S Tolerance limits for micro balance General Measurement Device and Calibration Topics 1
S IOQ protocol for a Micro-Vu measuring system - Is a PQ required? Qualification and Validation (including 21 CFR Part 11) 2
N How to Convert Micro Volts into Temperature Uncertainty Measurement Uncertainty (MU) 1
B Safety & EMC certificate updated when MCU (MIcro-Controller) is changed? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
B How to QC Small Micro-Fill Volumes in Amber Vials Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
J Micro Tubing Concentricity Checking using Crosshairs Manufacturing and Related Processes 1
E Micro-Vu Inspec Software User Manual Needed Quality Assurance and Compliance Software Tools and Solutions 1
B ISO 9001 for a small Micro Software Company Software Quality Assurance 18
A Baked Food Micro and Nutrional Value Testing Requirements Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Establishing Ranges for Weights for Calibration Verification of a Micro-Balance Calibration Frequency (Interval) 1
B Hand Swabbing Employees - Guidelines of micro limits and procedures and frequency Food Safety - ISO 22000, HACCP (21 CFR 120) 2
C Creating a Micro QMS - 4 geographical sites and 3 types of core activities ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Z What electronic height gage to buy? (micro-hite, fowler mestra, etc) General Measurement Device and Calibration Topics 6
M Calibration Requirements for MICRO 2 AIR COMPARATOR General Measurement Device and Calibration Topics 5
A Effectiveness of Training... micro or macro? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 29
P Micro Hardness Tester Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
Jen Kirley "Micro Combined-Heat-and-Power" System (micro-CHP) - The future of home energy? World News 10
Marc Advanced Micro Devices' new dual-core processors prices revealed After Work and Weekend Discussion Topics 2
Marc Micro direct methanol fuel cell system for IBM ThinkPad notebooks After Work and Weekend Discussion Topics 0
C What is the least acceptable TAR value? Mettler UMX2 micro balance - Class E2 weights General Measurement Device and Calibration Topics 4
I Microbiological Testing - Bioburden and Identification of Micro-Organisms ISO 13485:2016 - Medical Device Quality Management Systems 2
S Systems Approach or Process Approach? Micro level of managing ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 107
D SPC program called QC-Calc - Use with Micro-Vu equipment - Opinions please Quality Assurance and Compliance Software Tools and Solutions 2
V Micro Contamination - Emoderivates (producing albumin, etc from human plasma) ISO 13485:2016 - Medical Device Quality Management Systems 1
M Laser Micrometer Suggestion (if any) for diameters that are +/-.0001 inch General Measurement Device and Calibration Topics 1
C One Inch Micrometer Calibration Method General Measurement Device and Calibration Topics 2
C Uncertainty measurements for a 6 inch caliper Measurement Uncertainty (MU) 11
C Measurement Uncertainty fluctuates half a tick mark (20 millionths of an inch) Measurement Uncertainty (MU) 1
kedarg6500 What is a 5/8 inch NS HSG tap & drill size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
Marc Apple preparing to deliver iPad twins with 9.7-inch screens (rumor) After Work and Weekend Discussion Topics 0
Marc Three-Inch Device Detects 3,000 Types of Viruses and Bacteria World News 0
chris1price Sterile Pouch Peel Force of 1 pound/inch width Other Medical Device Related Standards 10
D How do I measure Pounds per Linear Inch (PLI) of a Continuous Wire? General Measurement Device and Calibration Topics 3
P A standard method of Dual Dimensioning drawings (Inch/Metric) Manufacturing and Related Processes 5
K Micron to Inch Conversion Misc. Quality Assurance and Business Systems Related Topics 10
Marc The world's largest LCD TV, a behemoth 108-inch liquid-crystal display After Work and Weekend Discussion Topics 2
D Micrometer that will read accurately to .00001 inch - Seeking General Measurement Device and Calibration Topics 11
J Maximum bend allowable tolerance on a three and a half inch truss head machine screw Manufacturing and Related Processes 4
R Measurement Uncertainty (MU) Budget for a 0-1 inch Micrometer Measurement Uncertainty (MU) 3
M SPC charts with Fixed Control Limits? Seam width and stitches per inch Statistical Analysis Tools, Techniques and SPC 11
B Need to know how to calibrate inch and metric gage blocks correctly. General Measurement Device and Calibration Topics 2
M SOPs - Reference Documents Section Document Control Systems, Procedures, Forms and Templates 3
B Reference to IEC 60950-1 in Amendment 2 IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Question on equipment - when to use reference only or research only stickers ISO 13485:2016 - Medical Device Quality Management Systems 5
J Marking "Distributed By" to product with distributors' reference-code EU Medical Device Regulations 3
Dazza 9001, 14001 and 45001 mandatory documents and records cross reference matrix Process Maps, Process Mapping and Turtle Diagrams 2
T CSE 5.01 reference on drawing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
N IEC 60601-1-1 - Stress test, reference voltage IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
CycleMike GD&T - Hole pattern - Print (attached) has a single Datum Reference Frame Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2

Similar threads

Top Bottom