820.120 UDI (Unique Device Identification) Labeling Verification Requirements

DCONN1

Starting to get Involved
#1
Hello all,

I am looking for guidance as to what exactly is required/mandated in terms of establishing barcode quality by the FDA UDI final rule. My knowledge on this topic is not comprehensive, and I certainly understand that using standards such as ISO/IEC 15415 and ISO/IEC 15416 could be of great use. However, I haven't been able to find any requirement mandating a specific barcode quality standard. The most relevant information I've encountered is:

§ 820.120 Device labeling

"(b)Labeling inspection.Labeling shall not be released for storage or use until a designated individual(s) has examined the labeling for accuracy[emphasis mine] including, where applicable, the correct unique device identifier (UDI) or universal product code (UPC), expiration date, control number, storage instructions, handling instructions, and any additional processing instructions. The release, including the date and signature of the individual(s) performing the examination, shall be documented in the DHR."

However, this section does not address quality; if you have a scanner which is able to read the barcode and show that the information is accurate, would this not satisfy this requirement?

Section (a) of § 820.120 remains unchanged by the UDI final rule and is relevant to the printed quality of the UDI:

"(a)Label integrity.Labels shall be printed and applied so as to remain legible and affixed during the customary conditions of processing, storage, handling, distribution, and where appropriate use."

"Remaining legible" in terms of the barcode (as well as human readable components) would certainly involve ensuring sustained readable barcode/human readable quality in some manner. However, since ISO/IEC 15415 and ISO/IEC 15416 are not referenced within the UDI final rule to my knowledge, is there a requirement to conform to these standards? Or is the method and scope of demonstrating compliance left up to the labeler?

My company uses GS1 as our accredited issuing agency. I am aware that GS1 has implementation guides for establishing verification programs for 1D and 2D barcodes which conform to /IEC 15415 and ISO/IEC 15416. However, is it necessary to follow these guidelines, or can a labeler establish a program as they see fit?

Again, I apologize for any ignorance on this topic and any help would be very much appreciated.
 
Elsmar Forum Sponsor

ScottK

Not out of the crisis
Staff member
Super Moderator
#3
I went with customer expectations on this issue.

Every one had an SOP that I had to review and they all required compliance to ANSI standards with an overall grade of C or better. As far as getting to that point - it was totally up to us.

While you don't need a bar code grader/verifier the expectation was that we would validate our printing system. And grader came in realty handy for that.

Most customers, in my experience, expect that you will own the grader/verifier to do regular checks.
 

Mark Meer

Trusted Information Resource
#5
GS1.org has quite a bit of information about UDI on the website.
They do, indeed.

...however, the question that I (and the OP, I presume) are more interested in is regarding the FDA's position and requirements/expectations. There is a lot of 3rd party guidance out there, but as far as anything official from the FDA...this is not clear.

...and, to be honest, it's not that clear in GS1 either. For example their technical overview of the Datamatrix states that print quality should be verified, and uses ISO/IEC 15415 as an example (but then go on to say it's "defacto" application standard...). They indicate throughout that verification needs to verify that (customer) requirements are met.

So back to the FDA, what does this mean? Is it acceptable to develop your own (simplified) tests for verifying print-quality with justifications such as risk, and redundant human-readable data?

...or, like many guidances seem to imply, is there an expectation that manufacturers need to invest in formal verification according to established standards (e.g. to ISO 15416, ANSI X3.182, ISO/IEC 15426-1, ISO/IEC TR 29158, ...etc.)?
 
Last edited:

DCONN1

Starting to get Involved
#6
Thanks for the replies all.

Yes, as Mark stated, I am most interested in determining the actual requirements of the UDI rule and what my company is mandated to do. I've continued to research this issue, but I've yet to come across anything resembling a clear mandate to use these standards. As such, if customers do not have relevant requirements pertaining to barcode quality testing, I would be hard pressed to justify the investment in expensive verification equipment, adding to production processes, and increasing training requirements.

21 CFR Part 830.20 states:

"A unique device identifier (UDI) must:
(a) Be issued under a system operated by FDA or an FDA-accredited issuing agency;
(b) Conform to each of the following international standards:
(1) ISO/IEC 15459-2, which is incorporated by reference at 830.10;
(2) ISO/IEC 15459-4, which is incorporated by reference at 830.10; and
(3) ISO/IEC 15459-6, which is incorporated by reference at 830.10.
(c) Use only characters and numbers from the invariant character set of ISO/IEC 646, which is incorporated by reference at 830.10."

While we will be sure to purchase and review these standards for compliance, does anyone know if ISO/IEC 15459-2, ISO/IEC 15459-4, ISO/IEC 15459-6 references any barcode quality standard such as ISO/IEC 15415 or ISO/IEC 15416?
 

Mark Meer

Trusted Information Resource
#7
Hi DCONN1,

I've decided to adopt a risk-based approach, and just document our decisions.

As was described in a previous thread, it can be said that the printed AIDC codes are:

1. Indelible (we test this internally);
2. Properly coded (we validate the software that does the encoding);
3. Scannable (we test); and
4. Redundant (the human-readable text is always provided).

(3) may be the only item which proper verification may be challenged, however given the intended use and risk of the device, and the redundancy described in (4), I just don't see any reasonable argument for doing extensive print-quality verification as per ISO/IEC standards.
 

DCONN1

Starting to get Involved
#8
Just as an update on this issue, the FDA released a draft guidance on July 26, 2016 titled "Unique Device Identification System: Form and Content of the Unique Device Identifier".

The paragraph beginning on line 264 addresses print quality with respect to ISO/IEC standards:

If a labeler choses a bar code form of AIDC, the bar code form of the UDI should be tested for print quality. Please refer to the most recent version of the following standards for more information on how to determine the print quality: ISO/IEC 15416 Information technology -- Automatic identification and data capture techniques -- Bar code print quality test specification -- Linear symbols; ISO/IEC 15415 Information technology -- Automatic identification and data capture techniques -- Bar code symbol print quality test specification – Two-dimensional symbols; and ISO/IEC TR 29158 Information technology -- Automatic identification and data capture techniques – Direct Part Mark (DPM) Quality Guideline. For linear and 2-D bar codes, labelers should consult the most recent version of the standards listed above, and the guidelines of their FDA-accredited issuing agency, to determine the minimum overall symbol grade based upon ISO/IEC verification processes. For purposes of this draft guidance, we define "overall symbol grade" as the arithmetic mean of the grades of multiple scans of the symbol. The minimum acceptable grade should be satisfied under the expected handling and use life of the device. Labelers should discuss print quality requirements with their FDA-accredited issuing agency.
To my knowledge this is the first time UDI print quality has been directly addressed by the FDA. It now seems as if companies which choose not to verify their barcodes with appropriate equipment according to these standards would not be in compliance.
 

Mark Meer

Trusted Information Resource
#9
Hmmm... this is always a frustrating element of regulatory creep: when "draft guidances" are treated (both by industry and the FDA) as de-facto mandatory.

Unfortunately, I suspect you are correct: that this draft guidance will become the standard expectation of compliance, and any deviation from it will raise issues.

(*Sigh*) I guess we'll all just have to budget for a few thousand dollars extra in additional equipment, training, and documentation maintenance just to comply (despite arguably little value to the company). :nope:
 

Sam Lazzara

Trusted Information Resource
#10
As a basic quality control measure I believe it is important to at least inspect each run of printed labels to make sure the the barcode can be correctly read using a simple "scanner" (USD $50 - $400) if you do not have the budget to use a real "verifier" (USD $3000-5000). My recommendation would be to either scan or verify at least 1 label printed for each unique production batch/lot.

Use a scanner is infinitely better than doing nothing. Using a verifier instead of a scanner does buy you assurance, but nowhere near the gain achieved when you go from nothing to a scanner.

And using a scanner is still a way to "check" print quality, albeit without the intense rigor of a real verifier, and without being able to claim conformance with the ISO print quality standards.

By the way, after doing quite a bit of research I landed on LVS verifiers as a good solution, for those with the verifier-bucks.
 
Last edited:
Thread starter Similar threads Forum Replies Date
S Suggested process to show compliance for this 820.120(a) for Label Integrity 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Implementation of 820.72 and 820.120 - Problem with Calibration Procedure 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
C QSR 820.120 labeling and packaging control 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R Sec. 820.120 Device Labeling & DHR - Count and Document All Labels 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
P Equipment 21 CFR 820.70(g) - User Requirements Document for Off the shelf equipment 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
D ISO 13485, FDA 21 CFR 820 and Auditing the Accounting Department ISO 13485:2016 - Medical Device Quality Management Systems 5
Ed Panek 21 CFR Part 820 - FDA Label Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
B Per 21CFR 820.198 (b) - Justification for no investigation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 22
Watchcat Does 820.30 include the manufacturing process? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
K 820.200 Servicing Procedure - Our system (because there is an API involved) is classified as Class 3 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
V Determining FDA 820 (registration) vs ISO 13485 - Supplier gives us the kit ISO 13485:2016 - Medical Device Quality Management Systems 1
R Addressing training requirements - 21 CFR Part 820.25 (1) & (2) Other US Medical Device Regulations 4
R How to improve a Validation program and procedures to FDA (21 CFR part 820) & ISO13485 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Marc Problem with 21 CFR Part 820 - US FDA Quality System Regulations (QSR) sub-forum link - 2 May 2019 Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 1
R 21 CFR Part 820 Contract Manufacturer of Medical Device Component 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Q Any good Checklists for ensuring SOPs cover ISO 13485:2016 and 21CFR 820? ISO 13485:2016 - Medical Device Quality Management Systems 3
B Classes/ Online Training on ISO 13485:2016 and FDA QSR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 5
C ISO 13485 certified as precursor to regulatory compliance to 21 CRF Part 820? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
N Change Control - Compliance with FDA 21 CFR Part 820 Document Control Systems, Procedures, Forms and Templates 3
D 21 CFR Part 820 (Subpart A) - Question about "Authority" 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
R RUO audits - CFR 820.22 and ISO 13485 8.2.4 Internal Auditing 3
A Training as well as certifications for ISO 13485:2016, 21 CFR 820... Training - Internal, External, Online and Distance Learning 5
K Production Unit Labeling - 21 CFR 820.184.e 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R 21 CFR Part 820.186 - Types of Quality System Records Document Control Systems, Procedures, Forms and Templates 1
M Does the Scope of 21 CFR Part 820.72 (Equipment) apply to Design? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
V 21 CFR 820 Compliant way to use Kaizen to address Nonconformances Nonconformance and Corrective Action 9
C 21 CFR Part 820.184 - Label Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
shimonv 21 CFR 820.20(d) says: Each manufacturer shall establish a Quality Plan 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
V Upgrading Systems from CFR 211 to CFR 820 (drug+device combination) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
S Looking for a checklist comparing ISO 13485:2016 and 21 CFR Part 820 ISO 13485:2016 - Medical Device Quality Management Systems 14
Q Internal Audit of Product Quality Complaint System (21 CFR Part 820) Customer Complaints 9
R Managing Employee Training Files - 21 CFR Part 820.25 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
M Release authorization before or after packaging per 21 CFR 820.80(d)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Need advice in streamlining antiquated 21 CFR 820 compliant QMS Quality Manager and Management Related Issues 2
J Minimum staff per 21 CFR Part 820 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Does a process map satisfy 21 CFR 820.20(d) (Quality Plan)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
R Calibration of processing equipment - 21 CFR 820.70 Production and process controls 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Issues with UDI additions to 21 CFR 820 requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
M Interpreting Process Controls - 21 CFR Part 820.70(a) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
A Mapping of IEC 62304 artefacts (SRS, SAD, etc) to the 820.30 phases IEC 62304 - Medical Device Software Life Cycle Processes 5
S How others have defined "reaudits"? 21 CFR Part 820.22 Internal Auditing 1
M FDA 21 CFR 820.250 - Does "valid statistical" always mean math? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
J Document Control Features (Document Content Requirements) and 21 CFR 820.40 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S 21 CFR Part 820.40(b) Clarification on Required Document Approvers 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R Where to find 21 CFR Part 820 Translated into Czech, Spanish, and German 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
N Certificate of Compliance from a Sister Company for FDA 21 CFR 820 Exemption 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
S Does FDA expect suppliers to be independently assessed to 21 CFR Part 820 ? US Food and Drug Administration (FDA) 4
A Effective Date (of Documents) issue outlined in 21 CFR Part 820.4 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
L Mobile Medical App - Understanding 21 CFR Part 820 Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
T "Special Process" Validation Requirements (21 CFR Part 820) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15

Similar threads

Top Bottom