Hello,

I have noticed that 510(k) A.I devices as below

Siemens.jpg

I think diagnostic softwares(Prostate Mr, Brain Mr) seem A.I analytic algorithm and they are operating as a module under image managing software(Engine).
I know that software alone is considered as medical device(SaMD) in FDA but I wonder A.I algorithm is also considered as a medical device.

My view is
A.I algorithm can't operate alone. It must be operating with image managing software.
Then it should be considered as component of image managing software.

Does A.I algorithm diagnostic software alone is considered as medical device?
 
Elsmar Forum Sponsor

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#2
Does this platform (AI) diagnose, detect, monitor, or suggest treatment or actions related to a health condition?
 

Brian K

Starting to get Involved
#3
Does this platform (AI) diagnose, detect, monitor, or suggest treatment or actions related to a health condition?
Normally, the platform detect, monitor the patient image. When it contains additional software(A.I diagnostic algorithm), it provides diagnose function. Kind of plug-in concept.
So my question was, A.I diagnostic algorithm is considered as medical device itself
 

yodon

Staff member
Super Moderator
#4
A.I algorithm can't operate alone.
Agreed. And based on that statement alone, the algorithm cannot be a medical device (software). As @Ed Panek points out, it has to have a medical purpose.

It must be operating with image managing software.
This, I think, is not totally accurate. Obviously there has to be input but it wouldn't have to be from image managing software.

Think of it like a database. You have a database back-end engine (e.g., SQL). The engine does nothing until an application is put on top of it. So you have to have a system that performs a medical purpose.
 

Brian K

Starting to get Involved
#5
Agreed. And based on that statement alone, the algorithm cannot be a medical device (software). As @Ed Panek points out, it has to have a medical purpose.



This, I think, is not totally accurate. Obviously there has to be input but it wouldn't have to be from image managing software.

Think of it like a database. You have a database back-end engine (e.g., SQL). The engine does nothing until an application is put on top of it. So you have to have a system that performs a medical purpose.

Then, you mean the devices captured in my question seem that they can perform a medical purpose not restricted to database right ?
 
Thread starter Similar threads Forum Replies Date
M ISO 14971 and Stand-Alone Diagnostic Software ISO 14971 - Medical Device Risk Management 4
D CE Mark advise sought for medical diagnostic Software CE Marking (Conformité Européene) / CB Scheme 4
G Importing Medical Device Software to India - Diagnostic Imaging Application Other Medical Device Regulations World-Wide 8
D FDA Approval for Diagnostic Software Clinical / Research US Food and Drug Administration (FDA) 4
W Software development company for Medical Device Diagnostic Equipment ISO 13485:2016 - Medical Device Quality Management Systems 5
R X-RAY Based Diagnostic Veterinary medical Devices Medical Device and FDA Regulations and Standards News 2
D Can a diagnostic device be class I? EU Medical Device Regulations 1
D IVD (in vitro diagnostic) Labelling Requirements - 98/79/EC CE Marking (Conformité Européene) / CB Scheme 2
Watchcat ISO 13485 for IVD (In-vitro Diagnostic Device) Manufacturers? ISO 13485:2016 - Medical Device Quality Management Systems 10
K Diagnostic X-ray devices - Applicability of Biocompatibility Testing per ISO 10993-1 Manufacturing and Related Processes 7
R Expected life span of Medical devices - X-Ray based diagnostic medical devices ISO 13485:2016 - Medical Device Quality Management Systems 6
M Informational ANVISA – CONSULTA PÚBLICA Nº 734, DE 22 DE OUTUBRO DE 2019 – Registro de dispositivos médicos para diagnóstico in vitro (in vitro diagnostic medical m Medical Device and FDA Regulations and Standards News 0
M Informational EU – EMA revised guidance – Questions & Answers on Implementation of the Medical Devices and In Vitro Diagnostic Medical Devices Regulations ((EU) 201 Medical Device and FDA Regulations and Standards News 2
M Informational EU – Flyer – Call for expression of interest for expert panels on medical devices and in vitro diagnostic medical devices (2019/C 323/05) Medical Device and FDA Regulations and Standards News 0
M Informational EU – Some documents related to Call for expression of interest for expert panels on medical devices and in vitro diagnostic medical devices Medical Device and FDA Regulations and Standards News 0
M Informational EU – Call for expression of interest for expert panels on medical devices and in vitro diagnostic medical devices Medical Device and FDA Regulations and Standards News 1
Watchcat Office of In Vitro Diagnostic and Radiological Health (OIR) De novos 2018 Other US Medical Device Regulations 1
M Informational EU – MDCG 2019-7 Guidance on Article 15 of the Medical Device Regulation (MDR) and in vitro Diagnostic Device Regulation (IVDR) regarding a “person re Medical Device and FDA Regulations and Standards News 7
M Informational Draft standardisation request as regards medical devices in support of Regulation (EU) 2017/745 and in vitro diagnostic medical devices in support of Medical Device and FDA Regulations and Standards News 2
M Informational US FDA final guidance – Marketing Clearance of Diagnostic Ultrasound Systems and Transducers Medical Device and FDA Regulations and Standards News 0
M Informational European Commission expert panels on medical devices and in vitro diagnostic devices Medical Device and FDA Regulations and Standards News 0
S IEC 60601-1-2 Radiated Immunity Test - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Informational USFDA Final Guidance – Medical Devices Containing Materials Derived from Animal Sources (Except for In Vitro Diagnostic Devices) Medical Device and FDA Regulations and Standards News 0
B MDSAP for OEM products in Canada for in vitro diagnostic test kit to include lancing device Other Medical Device Related Standards 0
S IEC 60601-1-2 immunity to proximity fields Test setup - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S IEC 60601-1-2 - RF Conducted Immunity Test Setup - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
S IEC 60601-1-2 Radiated E-Field Emission Testing - Diagnostic ECG device IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Medical Device News EU – Medical Devices Regulation (MDR) and In Vitro Diagnostic Medical Devices Regulation (IVDR) – infographics Medical Device and FDA Regulations and Standards News 0
M Medical Device News EU – Transition Timelines from the Directives to the Regulations – Medical Devices and in vitro Diagnostic Medical Devices Medical Device and FDA Regulations and Standards News 13
M FDA News USFDA Draft Guidance – Clarification of Radiation Control Regulations for Manufacturers of Diagnostic X-Ray Equipment Medical Device and FDA Regulations and Standards News 0
T RISK: Selling Diagnostic Devices on Amazon Manufacturing and Related Processes 0
S ISO13485 2016 - What the upshot is of failing to upgrade on a medical diagnostic device ISO 13485:2016 - Medical Device Quality Management Systems 1
T Question: Do we consider 60601-1-2 (4th ed) EMC testing a mode for a diagnostic port? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Re-Use of Diagnostic Reagent Solution Labels - Buffer Solutions for Internal Use ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
E EN 13640:2002 or EN23640:2013 For stability testing of in vitro diagnostic reagents ISO 13485:2016 - Medical Device Quality Management Systems 2
M IVD (in vitro diagnostic) Registration in Argentina Registrars and Notified Bodies 1
E IVD Diagnostic Kit Product Registration in Turkey - IVD Diagnostic Kits EU Medical Device Regulations 1
Q RUO (Research Use Only) vs. IVD (In Vitro Diagnostic) - Differences EU Medical Device Regulations 6
X Is in vitro diagnostic instruments not needed to be registered in India ? Other Medical Device Regulations World-Wide 5
X Morocco In Vitro Diagnostic Medical Device Registration Other Medical Device Regulations World-Wide 3
X In Vitro Diagnostic Medical Device in Palestine Registration and Importation Other Medical Device Regulations World-Wide 8
X In Vitro Diagnostic Medical Device in Turkey Registration Other Medical Device Regulations World-Wide 4
T What is Medical Device Classification for diagnostic equipment in ASEAN ? Other Medical Device Regulations World-Wide 2
Bev D Probability question - Components of Variation study on a single use diagnostic test Statistical Analysis Tools, Techniques and SPC 1
P Risk Management Files for IVD (Invitro Diagnostic Reagent) Kits ISO 14971 - Medical Device Risk Management 3
P IEC 60601-2-25:2011 Comparison Table (ECG, Diagnostic) IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
V Test Lab for JIS T 1501 (Performance of Ultrasonic Pulse-Echo Diagnostic Equipment) Japan Medical Device Regulations 1
L Hospital Laboratory IVD (In Vitro Diagnostic) Transportation Classification in Brazil Other Medical Device Regulations World-Wide 2
I IVD (in vitro diagnostic) Checklist for ISO 14971:2009 Annex H ISO 14971 - Medical Device Risk Management 3
S Veterinary Diagnostic Kits Regulations in South East Asia Other Medical Device Regulations World-Wide 8

Similar threads

Top Bottom