SBS - The best value in QMS software

A traditional 510K or Special 510K - Manufacturing facility and other changes

B

BigManP

#11
I'd previously communicated with an FDA rep about a docket related to our device but on a different subject. I'd mentioned that we plan to change labeling regarding MR compatibility, and she said that we would have to submit a 510(k) because the MR compatability data would have to be reviewed for the claim. That was all I got, so I was wondering if a Special 510(k) would be adequate. I'm still not sure of the answer myself, but we'll be working with a 510(k) consultant to help us with the process.
In the past we have submitted Special 510(k)s for minor design and/or material changes and as part of that added the MR Safe label and did not have any issues with it. Furthermore, I would always try a Special before a Traditional if I thought that we had a reasonable justification. They cost the same and the turn around on a Special is 30 days rather than 90. Worst case, they tell you that you that your submission is being changed to a Traditional, but you get credit for the 30 days already served. Good luck and watch out for those shady consultants! There are great ones out there, but not always easy to find.
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#12
In the past we have submitted Special 510(k)s for minor design and/or material changes and as part of that added the MR Safe label and did not have any issues with it. Furthermore, I would always try a Special before a Traditional if I thought that we had a reasonable justification. They cost the same and the turn around on a Special is 30 days rather than 90. Worst case, they tell you that you that your submission is being changed to a Traditional, but you get credit for the 30 days already served. Good luck and watch out for those shady consultants! There are great ones out there, but not always easy to find.
How do you tell a shady consultant from a great one, upfront?
 
B

BigManP

#13
If you can figure that out, you'll be ahead of the game for sure! Unfortunately, it is pretty difficult to tell if a regulatory/quality consultant is worth anything until after they have accomplished something. We worked with a consultant for 6 months and when he left all we had to show for it was a general framework of a system for a company 100 times our size and a filing cabinet full of labeled folders with no paperwork. Good luck!
 
T

tallu

#14
Could some one share good table of contents for Special 510k: device modification.

What sections are needed?
- cover letter
- Special 510k summary
- device description of modified device
- SE
- ...
 
B

BigManP

#15
The FDA provides guidance at http://www.fda.gov/MedicalDevices/D...remarketNotification510k/ucm142651.htm#link_5. I have successfully submitted several Special 510(k)s using the following format.

1.0 Medical Device User Fee Cover Sheet (Form FDA 3601)
2.0 CDRH Premarket Review Submission Cover Sheet
? Standards Data Report for 510(k)s
3.0 Special 510(k) Cover Letter
? Basis for Special 510(k) Submission
? Administrative Information
4.0 Truthful and Accurate Statement
5.0 Device Information
? Submitter Info
? Manufacturer Info
? Device Name/Class/Classification/Panel
? Predicate Device
? Labeling and Intended Use
6.0 Device Description and Comparison
? Currently Cleared Device Overview and Description
? Proposed Device Overview and Description
? Device Comparison
7.0 Substantial Equivalence Discussion
8.0 Summary of Design Control Activities
? Risk Assessment Method
? Declaration of Conformity
? Design Verification Results
9.0 Sterilization and Shelf Life
10.0 Biocompatibility
11.0 Class III Summary and Certification ? Not Applicable
12.0 Appendices
? Appendix A: 510(k) Summary
? Appendix B: Proposed Labeling
? Appendix C: Indications for Use Statement
? Appendix D: Proposed Engineering Blueprints
? Appendix E: Declaration of Conformity with Design Controls

Be sure to include a Design Control Activities Summary (DCAS) table in section 8. and don't include any of the actual test data. It is kind of messed up, but I used to include both and with my last they made me send a letter telling them to not review the data. However, two weeks later when they had questions I was asked to provide the data for further explanation. Typical FDA for you. I would also recommend adding the checklist that is described in the FDA link. I have run into issues where the FDA said I was missing content, but they just couldn't find it. The checklist will direct them to the satisfaction of each requirement.

Good luck!

Paul
 
T

tallu

#16
Thanks Paul!! :bigwave:

Its interesting that FDA doesn't want to see test data or evidence on device performance. They are mostly interested on dates and signs...

I had same kind of table content under work. I have included Biocompatibility and EMC sum-ups under Summary of Design Control Activities. What you think about that? Do I need to enclose biocompatibility certification/report as an exhibit, what about EMC test report.

With our previous cleared device we had some argue with the FDA on those issues.

Do you use basic form for the DCAS-table,
as example...

Modification/Feature Performed test Pass/Fail

Bluetooth Bluetooth pairing Pass
Data Logger Acquisition data recorded Pass
 
B

BigManP

#17
Yes, the FDA can be very confusing. I think that they often times make things much more difficult than they have to. What type of device modification are you proposing with your Special 510(k)? I would think that biocompatibility and EMC testing should probably have summaries providing a little more detail than the DCAS table would allow, but you can certainly be the judge of that. For example, if the material and processes were identical for the existing and the proposed devices, I would address it in the DCAS table but I would also provide a separate summary that highlights the points from the original submission. In the DCAS table, the reviewer that I have had did want to see the actual test results, but didn't want to see the entire protocol. So, depending on your device pass/fail might be acceptable, but I typically will be provided physical test data (strengths, forces, etc.). For my tables, I use device modification/potential risk/verification method/acceptance criteria/results and or comments. This was the template that was provided to me from the FDA and in their example they listed physical test data as well.

Good luck!

Paul
 
Thread starter Similar threads Forum Replies Date
P Letter to file or special 510k or traditional 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S The key differences in 510k types - Traditional vs. Abbreviated vs. Special US Food and Drug Administration (FDA) 0
N Traditional 510K Submission-Financial Certification or Disclosure Statement 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
D What is a reasonable consulting fee for preparation of a complete traditional 510k? Consultants and Consulting 11
Y OTS (Off The Shelf) Software Validation for 510k Traditional 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
R Charge (Fair Prices) by a Consultant for a traditional 510k Review Other US Medical Device Regulations 2
S Traditional 510k Filing SOP (Standard Operating Procedure) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
Watchcat Advantages to pursuing a Special 510(k) instead of a Traditional 510(k) Other US Medical Device Regulations 5
Aymaneh Traditional 510(k) Declaration of Conformity Other Medical Device Regulations World-Wide 3
A Several sections of traditional 510(K) have the same or similar requirement of content. US Food and Drug Administration (FDA) 4
A Some questions about voluntary consensus standards in traditional 510(k) submission US Food and Drug Administration (FDA) 6
A Some doubts about traditional 510(k) submission US Food and Drug Administration (FDA) 5
K 510(K) Submission (Traditional) for Medical Device US Food and Drug Administration (FDA) 17
F Replacement Bone Screws - Traditional 510(k) or Special 510(k)? Other US Medical Device Regulations 7
T Executive summary for Traditional 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
L Traditional 510(K) - Software - Clinical testing? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
Q Do we really need a traditional Quality Manual? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
Q Special 510(k) or New Traditional - Currently Inactive Products Other US Medical Device Regulations 2
G Need help on Device Description in the Acceptance Checklist for Traditional 510(k) Other US Medical Device Regulations 2
S What are the differences between Traditional 510(k) and a PMA to meet 21 CFR 814.20 ? Other US Medical Device Regulations 3
L What is the difference between Traditional and Abbreviated 510(K) application ? Other Medical Device and Orthopedic Related Topics 3
Q Sterilisation ISO 11135 EtO-in-a-Bag (Novel Non-Traditional EtO Sterilization) Other Medical Device Related Standards 12
Le Chiffre Listing standards in a traditional 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13
BradM Do you Hulu? Moving from "traditional" television After Work and Weekend Discussion Topics 21
M "Non-traditional" Corrective Action Procedure Nonconformance and Corrective Action 11
E Do I need submit a traditional 510(k) or Abbreviated 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
V Traditional Quality Costing vs. Activity Based Costing (ABC Quality Costing) Quality Tools, Improvement and Analysis 3
A Use of Declaration of Conformity in Traditional 510(k)??? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
L DOE Integrated Methodology - Traditional DOE vs. Taguchi Methods vs. Shainin Techniqu Quality Tools, Improvement and Analysis 7
Y 510k requirement for bringing in gloves Medical Device and FDA Regulations and Standards News 0
G Special 510K possible? Only the packaging has changed Other US Medical Device Regulations 0
K 510k FDA review, will they accept Biocompatibility result generated using feasibility product lots? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
H 510k with Contract Manufacturer Other US Medical Device Regulations 2
K FDA 510k electrosurgery 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
J Leveraging FDA 510k Clearance for International Registrations Other Medical Device Regulations World-Wide 2
K 510K new manufacturer but same components Other US Medical Device Regulations 6
S Two or more predicates suitable? 510K submission US Food and Drug Administration (FDA) 5
S 510k: What to include if your own device is the predicate US Food and Drug Administration (FDA) 5
L 510K - Trade Name Change 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
Ed Panek Adverse Event Clinical Trial using a 510K approved Device Other US Medical Device Regulations 6
R Countries that accept 510k approvals - Class II nonsterile medical device Other Medical Device Regulations World-Wide 2
M Timing of 510k submission Other US Medical Device Regulations 9
S Examples of software changes that required a 510k US Food and Drug Administration (FDA) 2
D 510K and Changes to Verification and Validation US Food and Drug Administration (FDA) 2
Ronen E Can anyone please explain this? (510k submission requirement) Other US Medical Device Regulations 14
H New 510K VS Documents about design change. Other US Medical Device Regulations 2
P Medical Device not on commercial distribution but installed in the establishment. 510k? Other US Medical Device Regulations 10
V Presubmission FDA for 510k US Food and Drug Administration (FDA) 4
A How to deal with changed shared components in 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Can we take a part from 510k cleared medical device and use it in class I device? Other US Medical Device Regulations 3

Similar threads

Top Bottom