Accelerated Ageing Test for Orthopedic Implants -- ASTM F 1980 or ISO 11607 or both are applicable ?

mpfizer

Involved In Discussions
#1
Hi,

What are the ASTM or ISO standards to be followed for the above ... ASTM F 1980 or ISO 11607 OR both ?

Has the stability studies to be done at higher temperature AND Humidity OR only higher temperature.

what should the ambient temperature be taken as ... it can vary from say 15 to 40 degree ( centigrade ) ... or do we need to do study at 15 and 40 deg. temperature

Any help would be appreciated.

Thanks

michelle
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
Are you studying the shelf life of the sterile (or otherwise) packaging, the shelf life of the implant or the use life of the implants?

Both standards you mention apply to the (sterile) packaging. For the implants, you need a careful engineering analysis or a device-specific standard. Otherwise you risk GIGO.
 

QM_123

Starting to get Involved
#3
Hi,

What are the ASTM or ISO standards to be followed for the above ... ASTM F 1980 or ISO 11607 OR both ?

Has the stability studies to be done at higher temperature AND Humidity OR only higher temperature.

what should the ambient temperature be taken as ... it can vary from say 15 to 40 degree ( centigrade ) ... or do we need to do study at 15 and 40 deg. temperature

Any help would be appreciated.

Thanks

michelle
As Ronen says, these standards for sterile packaging. We perform this to be sure for shelf life of the device (according to ISO 11607 you shall start real-time shelf life determination study within 3 months after AAT) . Regarding your question related with temperature: You know there is a calculation, if you prefer high temperature, your testing time will decrease or if you prefer lower temperature your test may last more than you expect. I suggest you to consider you packaging material specifications to determine the temperature and humidity. I assume you are asking this for shelf life and answer is related with that. If your concern is device aging it is very complicated.
 
Thread starter Similar threads Forum Replies Date
Z Industrial Standard on Accelerated Ageing test for plastic carrier tape Reliability Analysis - Predictions, Testing and Standards 8
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
E When to perform accelerated aging studies and transport test(OQ or PQ) Qualification and Validation (including 21 CFR Part 11) 4
R Accelerated Aging - Creating test samples - Implantable medical device Question Other Medical Device Related Standards 4
G Accelerated stability of medical devices Other Medical Device Related Standards 1
S Stability accelerated vs. real-time Other Medical Device Related Standards 3
B Class III Medical Device Accelerated conditions Other Medical Device Related Standards 1
shrutisancheti Predicate device accelerated aging - 1 year of shelf life is already over Other Medical Device Related Standards 1
A Accelerated Reliability Demonstration Test Reliability Analysis - Predictions, Testing and Standards 4
A Accelerated aging and real time study EU Medical Device Regulations 2
shimonv Storage Conditions of 5-40 Celsius - Accelerated and real time aging - Dilemma EU Medical Device Regulations 3
W Shelf life and accelerated aging ISO 13485:2016 - Medical Device Quality Management Systems 2
A Accelerated Stability Studies - Class III Medical Gloves Other Medical Device and Orthopedic Related Topics 1
V Accelerated Learning and Simulations / Scenarios - Statistics Statistical Analysis Tools, Techniques and SPC 0
M Accelerated Service Life Test for Medical Devices Other Medical Device and Orthopedic Related Topics 1
N Accelerated Aging Testing Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
U Accelerated Aging for Shelf Life - Beginning Point Difference CE Marking (Conformité Européene) / CB Scheme 2
C Is there a standard for HALT (High Accelerated Life Testing)? Reliability Analysis - Predictions, Testing and Standards 15
N Accelerated Aging Test for Disposable Protective Garment for Incontinence Other US Medical Device Regulations 1
6 Accelerated Shelf Life Testing Protocol for Juice Containing Beverages Food Safety - ISO 22000, HACCP (21 CFR 120) 1
Ron Rompen Seeking Test Lab which can do 1440 Hour Accelerated Weathering Characteristics Misc. Quality Assurance and Business Systems Related Topics 1
B Test Procedure or Standard for Accelerated Aging for AIMD Various Other Specifications, Standards, and related Requirements 2
D Stability (Accelerated Life) Testing of Medical Devices for Home Use ISO 13485:2016 - Medical Device Quality Management Systems 3
M Shelf-Life - Accelerated Age Test ISO 13485:2016 - Medical Device Quality Management Systems 8
M Accelerated data enough to claim 3 years and be on the Market? ISO 13485:2016 - Medical Device Quality Management Systems 13
D Anyone familiar with HALT (Highly Accelerated Life Test) to predict normal life Reliability Analysis - Predictions, Testing and Standards 10
R Anyone familiar with HAST (Highly Accelerated Stress Test)? Reliability Analysis - Predictions, Testing and Standards 10
A Format of Stent Stability Testing (Accelerated and Long Term Aging) Other Medical Device and Orthopedic Related Topics 8
C Thermal Accelerated Aging Test(s) for Epoxy Reliability Analysis - Predictions, Testing and Standards 6
R Accelerated Stability Studies - Duration and Conditions - Class III Medical Device Other Medical Device and Orthopedic Related Topics 4
R Long Term vs. Accelerated Life Testing FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
S Validation of Sterilization Expiry Date - Accelerated Aging, etc ISO 13485:2016 - Medical Device Quality Management Systems 6
V Accelerated Stability Studies - In vitro Diagnostic Devices ISO 13485:2016 - Medical Device Quality Management Systems 2
J Customer Prototype Support - 4.4.10 - Accelerated life tests QS-9000 - American Automotive Manufacturers Standard 2
Moumen H Variations between ASTM A29 Standard for steel bars and Mill test certificates specs Manufacturing and Related Processes 1
D Tip test 10° -> 5° IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Sponsor name presented in the test report EU Medical Device Regulations 4
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Non parametric test for semi-quantitative data. Statistical Analysis Tools, Techniques and SPC 5
Ed Panek Disaster Recovery Test IT (Information Technology) Service Management 3
A How much does a complete biocompatibility test package cost? Other ISO and International Standards and European Regulations 1

Similar threads

Top Bottom