Dear fellow colleagues in the medical device field,

We have a skin product ready and classification will be IIb.
At this moment we only have accelerated stability data, only performed on 1 batch, results are good.
Can we already commercialize the product and have it on the Market?
Can we claim a 2 or 3 year expiry date shelf life for the product and be on the Market with only results from the accelerated study.
We commit ourselves to start up stability according ICH, long-term, 3 batches, etc.

Are there any regulations pointing at this in de ISO13485 or other guidelines that tell what is necessary to be on the Market regarding shelf life/aging data?

Kind regards
 
Elsmar Forum Sponsor
R

Rob Udo

#2
Hi,

I'm not aware of any specific standards. You don't give information whether the accelerated aging was done for 1,2, or 3 years simulated aging.

Common practice is to claim e.g. one year based on your accelerated aging results. It should be part of your risk assessment whether you feel comfortable with the performed testing in order to justify your claim. Make sure that you feel comfortable with a sample size of one. Also considere shelf life claims of similar products in the market and use that information to justify your decision. In that case you should be able to demonstrate equivalence of the product (materials used), sterilization method (if applicable), packaging configuration.

Based on real time aging results for 2 year, you can extend the shellf life to 2 years. You can start a 3 year study simultaneously. Make sure that you document your shelf life study in an aging plan, and stick to it.

Be aware that it is not only about stability testing. Also consider packaging integrity.
 
#4
From all the standards that are available the best one to use is:
ASTM F1980-07, Standard Guide for Accelerated Aging of Sterile Systems for Medical Devices.
It is not perfectly-matching your case but it is the best in our market segment and it is a consensus standard.
 

Michael Malis

Quite Involved in Discussions
#5
Having just an accelerated data on 1 batch to claim 3 years for Class IIb product is not enough!

You must have at min. 3 batches with accepted Functional and Package integrity tests of accelerated data to claim the appropriate shelf life for your product. Also, the accelerated test mustbe followed with the Real Time Data. If you don't have a real time data, - you can have a real problem with the product because accelerated data is still is only simulation of the real outcome!
 

Sam Lazzara

Trusted Information Resource
#6
Hello Michael,
Can you please let me know the standard/guideline that has the 3 lot/batch requirement you refer to (for medical devices)?
I did see a 3 lot/batch guideline on the ICH website for pharmaceutical stability testing, but have not yet found such a written guideline for devices.

Thank you.
 

v9991

Trusted Information Resource
#7
First :- ICH Q1E specifies that maximum shelf assigned can be either 2X or x+12 whichever is lower (X- no. of months of stability in real time). You might need to check its applicability to medical devices.

Second - Reference of 3 batches of stability, from the pharma perspective, these stability batches are part of validation program; where these validation batches are charged for stability; I think these details are covered in CTD format of regulatory submissions as well.

hope that helps.
valiveti.
 
G

gholland

#8
FDA is starting to back away from the 3 lot thing, they are much more interested in the variability of the product/process being tested. If you can justify that you capture all of the variability in the product and process with one lot... not likely but you never know... then you can go with one lot. You have to have a solid justification for your sample size.

Our company sets shelf life on accelerated aging results but also completes real time aging results as confirmation of the aging tests. We run at risk assuming that accelrated aging will be much rougher on the packaging and product than sitting on a shelf for 3 years.

For a 3 year shelf life claim you need 3 year accelrated aging data, your original post wasn't clear what kind of data you have. You can't claim 3 year shelf life on 1 or 2 year aging data.

:2cents:
 
W

wendyzeng

#9
FDA is starting to back away from the 3 lot thing, they are much more interested in the variability of the product/process being tested. If you can justify that you capture all of the variability in the product and process with one lot... not likely but you never know... then you can go with one lot. You have to have a solid justification for your sample size.

Our company sets shelf life on accelerated aging results but also completes real time aging results as confirmation of the aging tests. We run at risk assuming that accelrated aging will be much rougher on the packaging and product than sitting on a shelf for 3 years.

For a 3 year shelf life claim you need 3 year accelrated aging data, your original post wasn't clear what kind of data you have. You can't claim 3 year shelf life on 1 or 2 year aging data.

:2cents:
Thanks for your information!
 

v9991

Trusted Information Resource
#10
I am certainly NOT into devices; hence following queries.

1) on FDA moving away from the 3 batches-->''focus on variability...";
can you pl provide any reference for your note with specific reference to stability data? ; btw, what are the regulatory guidance/references for carrying out stability studies of medical devices?

I have found similar note on waiving the process validation requirements... that is more specific and to 'devices'and perspective of "Quality by Design" and modern manufacturing and analytical technologies.this can be found in CPG...
http://www.fda.gov/Drugs/GuidanceComplianceRegulatoryInformation/Guidances/ucm124720.htm

IX. ACCEPTABILITY OF PRODUCT TESTING
In some cases, a drug product or medical device may be manufactured individually or on a one-time basis. The concept of prospective or retrospective validation as it relates to those situations may have limited applicability, and data obtained during the manufacturing and assembly process may be used in conjunction with product testing to demonstrate that the instant run yielded a finished product meeting all of its specifications and quality characteristics. Such evaluation of data and product testing would be expected to be much more extensive than the usual situation where more reliance would be placed on prospective validation.


2) ; but isn't it obvious that evaluating the device on stability implies that it should not be just on shelf but it must be actually operational as it is intended to be used( atleast a simulated environment of operation). And why do we treat that as accelerated condition.
(if i can mention a simple-equivalent example here; if we need to prove the shelf life of TV, then its obvious that you would subject it to 'usage' for specified time for the stated shelf life time; and accelerated condition is something like operating in various temperature/power/... fluctuation conditions)

thanks in advance for clarification.
valiveti.
 
Thread starter Similar threads Forum Replies Date
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
E When to perform accelerated aging studies and transport test(OQ or PQ) Qualification and Validation (including 21 CFR Part 11) 4
R Accelerated Aging - Creating test samples - Implantable medical device Question Other Medical Device Related Standards 4
G Accelerated stability of medical devices Other Medical Device Related Standards 1
S Stability accelerated vs. real-time Other Medical Device Related Standards 3
B Class III Medical Device Accelerated conditions Other Medical Device Related Standards 1
shrutisancheti Predicate device accelerated aging - 1 year of shelf life is already over Other Medical Device Related Standards 1
M Accelerated Ageing Test for Orthopedic Implants -- ASTM F 1980 or ISO 11607 or both are applicable ? Other Medical Device and Orthopedic Related Topics 2
A Accelerated Reliability Demonstration Test Reliability Analysis - Predictions, Testing and Standards 4
A Accelerated aging and real time study EU Medical Device Regulations 2
shimonv Storage Conditions of 5-40 Celsius - Accelerated and real time aging - Dilemma EU Medical Device Regulations 3
W Shelf life and accelerated aging ISO 13485:2016 - Medical Device Quality Management Systems 2
A Accelerated Stability Studies - Class III Medical Gloves Other Medical Device and Orthopedic Related Topics 1
V Accelerated Learning and Simulations / Scenarios - Statistics Statistical Analysis Tools, Techniques and SPC 0
M Accelerated Service Life Test for Medical Devices Other Medical Device and Orthopedic Related Topics 1
N Accelerated Aging Testing Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
U Accelerated Aging for Shelf Life - Beginning Point Difference CE Marking (Conformité Européene) / CB Scheme 2
C Is there a standard for HALT (High Accelerated Life Testing)? Reliability Analysis - Predictions, Testing and Standards 15
N Accelerated Aging Test for Disposable Protective Garment for Incontinence Other US Medical Device Regulations 1
6 Accelerated Shelf Life Testing Protocol for Juice Containing Beverages Food Safety - ISO 22000, HACCP (21 CFR 120) 1
Ron Rompen Seeking Test Lab which can do 1440 Hour Accelerated Weathering Characteristics Misc. Quality Assurance and Business Systems Related Topics 1
B Test Procedure or Standard for Accelerated Aging for AIMD Various Other Specifications, Standards, and related Requirements 2
D Stability (Accelerated Life) Testing of Medical Devices for Home Use ISO 13485:2016 - Medical Device Quality Management Systems 3
M Shelf-Life - Accelerated Age Test ISO 13485:2016 - Medical Device Quality Management Systems 8
D Anyone familiar with HALT (Highly Accelerated Life Test) to predict normal life Reliability Analysis - Predictions, Testing and Standards 10
R Anyone familiar with HAST (Highly Accelerated Stress Test)? Reliability Analysis - Predictions, Testing and Standards 10
A Format of Stent Stability Testing (Accelerated and Long Term Aging) Other Medical Device and Orthopedic Related Topics 8
C Thermal Accelerated Aging Test(s) for Epoxy Reliability Analysis - Predictions, Testing and Standards 6
Z Industrial Standard on Accelerated Ageing test for plastic carrier tape Reliability Analysis - Predictions, Testing and Standards 8
R Accelerated Stability Studies - Duration and Conditions - Class III Medical Device Other Medical Device and Orthopedic Related Topics 4
R Long Term vs. Accelerated Life Testing FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
S Validation of Sterilization Expiry Date - Accelerated Aging, etc ISO 13485:2016 - Medical Device Quality Management Systems 6
V Accelerated Stability Studies - In vitro Diagnostic Devices ISO 13485:2016 - Medical Device Quality Management Systems 2
J Customer Prototype Support - 4.4.10 - Accelerated life tests QS-9000 - American Automotive Manufacturers Standard 2
J Need Help with FPY Data in Assembly Process Manufacturing and Related Processes 7
Q AMS 2750 E or F Continuous Furnace TUS Data Collection AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M Reduce occurrence rating based on the PMS data and customer complaint data ISO 14971 - Medical Device Risk Management 2
J Customer Complaint & SCAR, false data Nonconformance and Corrective Action 14
Brizilla Employee Data Privacy Policy - ISO 9001:2015 requirement(s)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D Safety data sheets software REACH and RoHS Conversations 2
M Data Protection and Privacy Policy - looking for a template/example EU Medical Device Regulations 1
S Non parametric test for semi-quantitative data. Statistical Analysis Tools, Techniques and SPC 5
M Disabling measurement data during fault conditions IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C EU MDR - Annex II 6.1 Pre-clinical and clinical data EU Medical Device Regulations 4
P Ppk results shown as asterisk after the transformation of Non-normal data Using Minitab Software 4
lanley liao How to correctly understand the bullet list d) of 6.3 Analysis of Data for API Spec Q1 Oil and Gas Industry Standards and Regulations 7
Steve Prevette Informational I am presenting a webinar Thursday - "Data Driven Decision Making" - 19 November 2020 Statistical Analysis Tools, Techniques and SPC 5
qualprod Best practice to ensure inputting of data in production Lean in Manufacturing and Service Industries 19
D Preservation of Electronic Data / Information Technology ISO 13485:2016 - Medical Device Quality Management Systems 5
M Comparing data from destructive testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7

Similar threads

Top Bottom