Accessible metal parts connected to secondary ground
Talsh
Hi everyone.
My system include a BF applied part connected to a Panel PC
The system is driven by a medical power supplier with output of 18 VDC .
The panel PC has a metallic accessible part (VESA mount and connectors) which are connected to the panel PC functional ground.
The accessible metallic part is not protective earthed - no electrical connection between the accessible part and mains ground.
The medical power supplier provides 2 MOPP, and the BF applied part is isolated with 2 MOPP form the panel PC.
My questions are:
  1. Is this design acceptable?
  2. Can the metal accessible parts be left not connected to protective earth ?
  3. If so, does this system can be considered as class II ME ?
Thanks in advance,
Peter Selvey
Yes, the design is acceptable, and the parts can be left earth free or functionally earthed (e.g. connected to earthed 0V for EMC/noise purpose, but not earthed for electric shock protection).
The classification (Class I, Class II) is related to the mains connection, so it is really only applicable to the power supply, which in the diagram seems to be Class I. The main unit does not in principle have a class (or would be Class III in other standards). IEC 60601-1 is a bit vague about this, which is historically based, bit of a long story. Many people (and test labs) still call the main unit Class II, and put the double box symbol on. However, if the standard is read carefully, all the clauses that are relevant for Class II, including marking of the double box symbol, are not applicable as the Clause title or text limits the application to mains parts.
J Patient connection and accessible parts - enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A APPLIED vs ACCESSIBLE PART IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E How to test 8.7.4.7 d) ? What "not protectively earthed accessible part" means? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
aleexander Grounding of metallic accessible parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
B Accessible parts per 8.4.2 b) - Touch currents in combination with MOOP or MOPP? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C IATF 16949:2016 Section 8.5.1.2 item d) accessible for use at the designated work area(s) IATF 16949 - Automotive Quality Systems Standard 5
Ahang Separation between secondary circuit and accessible parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
V 2MOP of mains voltage is required between 2nd Circuity and Accessible parts? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
M Barrier Free vs. Wheelchair Accessible - Differences Occupational Health & Safety Management Standards 1
A 61010-1 Accessible Parts - Limit Values and Crepage and Air Clearance Distances Other Medical Device Related Standards 1
A Insulation between Secondary Accessible Parts and Earth IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
A Ground bond test when no ground is accessible IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Should Documented Procedures be accessible to all people/employee? Document Control Systems, Procedures, Forms and Templates 20
T Section 7.5.1.2 - Instructions SHALL be accessible for use at the workstation IATF 16949 - Automotive Quality Systems Standard 18
Q Work Instructions 4.9.1 - Work instructions must be readily accessible QS-9000 - American Automotive Manufacturers Standard 8
C Mistake Proofing Winding Direction Stamped Metal Parts Manufacturing and Related Processes 7
O New metal detector, ccp -help! Food Safety - ISO 22000, HACCP (21 CFR 120) 2
A Critical Characteristics - Sheet metal stamping APQP and PPAP 9
M Metal parts bonding/grounding on Class II, internally powered with functional earth? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J ISO 13485 for Metal Finishing Medical Device and FDA Regulations and Standards News 5
A Metal Enclosure Connected to PCB Ground for ESD Reasons IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Lighting Standard for Metal Processing Inspection Manufacturing and Related Processes 4
K Quality Control/Assurance in the Metal Industry Quality Tools, Improvement and Analysis 10
M 'Active' device? Software/laptop with attached camera 'looking' at passive metal probe EU Medical Device Regulations 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Raw material certificates - CC - Safety products - Sheet metal stamping IATF 16949 - Automotive Quality Systems Standard 1
A Literature review/HACCP validation of metal detection Food Safety - ISO 22000, HACCP (21 CFR 120) 0
M Informational US FDA paper – Biological Responses to Metal Implants Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Final guidance – Metal Expandable Biliary Stents – Premarket Notification (510(k)) Submissions Medical Device and FDA Regulations and Standards News 0
Q IATF rule for single site - Ingots from scrap metal recycling company IATF 16949 - Automotive Quality Systems Standard 0
C AMS2750E - Noble metal as a Load sensor AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
V Thinning percentage in sheet metal forming per SES (Suzuki Engineering Standard) Manufacturing and Related Processes 1
F AS9100 D - Verification of purchased product - Sheet metal and aluminum extrusion stock AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
F MOPP in battery powered device with metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Error proofing ideas sheet metal punches Manufacturing and Related Processes 8
D IEC 60601 Defibrillation protection test - Metal foil IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Ron Rompen Crack testing of powdered metal parts General Measurement Device and Calibration Topics 4
A Test method for presence or absence of hexavalent chromium on a metal coating RoHS, REACH, ELV, IMDS and Restricted Substances 4
R Statutory/Regulatory for NACE Code 25.9 (Manufacture of fabricated metal products) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Industry Standards - Sheet Metal a Machine Shop Manufacturing and Related Processes 1
J Measuring an Untrue Radius - Metal Stamped and Formed Parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
S Metal Radiographic and Ultrasonic Testing - Special Processes AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 11
J Receiving Inspection for Metal Strips & Wires Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
V Root Cause help needed - Sheet metal breaking at press Problem Solving, Root Cause Fault and Failure Analysis 9
Ron Rompen Help with Corrosion (Humidity) Testing Procedure - Sintered Metal Steel Parts Manufacturing and Related Processes 9
C DFAR 252.225-7008 - Is aluminum, say T6061, considered specialty metal (i) steel? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Ron Rompen Sintered Metal Machined Parts Rust Removal Methods Manufacturing and Related Processes 8
J Internal PPM's in a medium sized metal stamping plant Quality Manager and Management Related Issues 6
G Metal Scale (SIP-Societe Genevoise) change in size over time? General Measurement Device and Calibration Topics 8