Acclimation time prior to calibration

CaliperJim

Involved In Discussions
Hi. We calibrate micrometers, calipers, and protractors in house. The lowest device measurement tolerance we have is 0.0005. The shop floor where the devices are being used is uncontrolled. How long do I need to allow the devices to sit in the calibration room before calibrating? Thank you very much.
 

Dazzur

Involved In Discussions
If you're talking about ISO 9001:2015 here, you don't. Just calibrate them, it doesn't call for specifics, just that you have a suitable environment, what that suitable environment is up to you.
Not every organisation is going to have the resources for Calibration Labs.
 

Michael_M

Trusted Information Resource
I am not familiar with the specifics of ISO 17025 (Calibration) however, I believe it is 24 hours. That being said, with a tolerance of .0005 (inches I assume) it would take some drastic heat difference to show growth or shrinkage.

When you say "tolerance we have is .0005" do you mean the device can only read to .0005 or that is the out of calibration failure? If it's the out of calibration failure, I would assume your calipers fail frequently and your micrometers rarely fail.

Unless you have specific requirements from customers that state how you have to do something, I would not worry about it.
 

CaliperJim

Involved In Discussions
When you say "tolerance we have is .0005" do you mean the device can only read to .0005 or that is the out of calibration failure? If it's the out of calibration failure, I would assume your calipers fail frequently and your micrometers rarely fail.
We have micrometers that read to .0001 but our tolerance is .0005. Our tolerance for calipers is .001.
 

CaliperJim

Involved In Discussions
If you're talking about ISO 9001:2015 here, you don't. Just calibrate them, it doesn't call for specifics, just that you have a suitable environment, what that suitable environment is up to you.
Not every organisation is going to have the resources for Calibration Labs.
Our customers require us to adhere to MIL-STD-45662A and we are ISO 9001:2015 certified.
 

dwperron

Trusted Information Resource
One way you can figure out thermal expansion of the tools is using the formula
▲L = L * ▲T * CE
Where ▲L is the thermal expansion, L is the length of the tool / part being measured, ▲T is the temperature difference in the measurement, and CE is the coefficient of expansion for tool steel. The CE varies for different steel alloys, but a value of 6.5 ppm/°F is typically used
So suppose you have a 6 in caliper, and the factory floor temperature is 80 °F and your calibration space is 70 °F (a 10 °F difference that needs to be accounted for). Plugging in those numbers you get:
L = (6 in)(10°F)(6.5 ppm/°F) = 390 µin.
As you can see the number is small, but it can get significant. Suppose that you have a 24 in caliper and you bring it from the 80 °F factory to your 70 °F lab and you don't let the temperature of the caliper acclimate. then the thermal expansion that needs to be accounted for is:
L = (24 in)(10°F)(6.5 ppm/°F) = 1560 µin, or 0.0015 in. As you can see, that is now a considerable error.
 

CaliperJim

Involved In Discussions
One way you can figure out thermal expansion of the tools is using the formula
▲L = L * ▲T * CE
Where ▲L is the thermal expansion, L is the length of the tool / part being measured, ▲T is the temperature difference in the measurement, and CE is the coefficient of expansion for tool steel. The CE varies for different steel alloys, but a value of 6.5 ppm/°F is typically used
So suppose you have a 6 in caliper, and the factory floor temperature is 80 °F and your calibration space is 70 °F (a 10 °F difference that needs to be accounted for). Plugging in those numbers you get:
L = (6 in)(10°F)(6.5 ppm/°F) = 390 µin.
As you can see the number is small, but it can get significant. Suppose that you have a 24 in caliper and you bring it from the 80 °F factory to your 70 °F lab and you don't let the temperature of the caliper acclimate. then the thermal expansion that needs to be accounted for is:
L = (24 in)(10°F)(6.5 ppm/°F) = 1560 µin, or 0.0015 in. As you can see, that is now a considerable error.
How can I determine the adequate soak time for the tools? Our calipers range from 0-6 to 0-24, our micrometers range from 0-1 to 8-12, our height gauges are 0-24.
 

Miner

Forum Moderator
Leader
Admin
If I were to design a study to determine the adequate soak time, I would use a repeated measures design as follows.
  1. Note the ambient temperature for the measurement device on the shop floor.
  2. Bring the device into the lab and immediately measure the calibration standard(s) (3-5 different sizes).
  3. Repeat at defined time intervals (e.g., hourly) for 24 hours (the typical recommended soak time).
  4. Repeat steps 1-3 over the course of several days.
  5. Analyze as a repeated measures design and determine at what time interval the measurement results have stabilized (this will probably vary depending on the shop floor temperature (i.e., larger delta T = longer soak time)).
 
Thread starter Similar threads Forum Replies Date
J Determination of software safety class (62304) prior to software risk analysis ISO 14971 - Medical Device Risk Management 3
D How to address the content deviation of 'cannot apply criteria of risk acceptability prior to...' ISO 14971 - Medical Device Risk Management 1
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
A Single-Use accessory disinfected prior to use Other Medical Device Related Standards 6
R Disruptions that happen prior to surveillance audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
B Is 14971 Annex C checklist now in ISO/TR 24971 required to complete prior to 510k filing? ISO 14971 - Medical Device Risk Management 3
M How to apply IEC 62304 Problem Resolution Process for bugs discovered prior to release i.e. during initial development? IEC 62304 - Medical Device Software Life Cycle Processes 2
blackholequasar ISO 13485 certification prior to Medical Device Manufacturing... worth it? ISO 13485:2016 - Medical Device Quality Management Systems 4
S Inventory prior to AS9120 certification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Quality Objective for QMS prior to Certification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 10
P Writing a presubmission to the FDA prior to the De novo submission US Food and Drug Administration (FDA) 4
A Is it mandatory to obtain ISO13485 certificate prior to CE marking? EU Medical Device Regulations 8
V Supplier wants to perform pre-qualifications prior to every run instead of locking in parameters. Supplier Quality Assurance and other Supplier Issues 7
L Unacceptable risk for basic safety prior to or after mitigation? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
JoshuaFroud CE Marking and Brexit - Product shipped to a distributor prior to the March deadline EU Medical Device Regulations 2
S Adding Value prior to signing EASA Form 1 EASA and JAA Aviation Standards and Requirements 5
S IMS Manual Review and Revision Prior to Audit ISO 14001:2015 Specific Discussions 4
A ISO 9001:2015 registrar Auditor requesting copies of procedure prior to audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
Q Top Management Change Prior to Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
H Do you have to perform an full internal prior to IATF 16949 audit? IATF 16949 - Automotive Quality Systems Standard 21
K Markets with No Prior Approvals? Without Country of Origin or other Prior Approval Other Medical Device Regulations World-Wide 5
R AS9100 FAI - First Article Inspection to a prior Customers PO AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
A Internal Audit - Findings - Recent Internet Audit (Prior to Certification) IEC 27001 - Information Security Management Systems (ISMS) 7
S Soak Time to Acclimate prior to Calibration General Measurement Device and Calibration Topics 1
O Process Mapping prior to Validation and Software to use to Map Process Maps, Process Mapping and Turtle Diagrams 12
JoCam Choose Medical Device Distributors prior to Device Registration? Other Medical Device Regulations World-Wide 6
E Is Design Transfer to Production be complete prior to Release of first production lot 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Q ISO 9001:2008 requirement for Design Organization Approval prior to "Repair" ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
Q AS9100 Repairs - Approval from a non-aerospace customer prior to conducting a repair AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
L Who can perform ISO 9001:2008 Internal Audit prior to our Registration Audit? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
C How to Clean Gage Blocks and Gage Pins prior to Calibration General Measurement Device and Calibration Topics 4
L How to add a New Factory under a Prior 510K Establishment US Food and Drug Administration (FDA) 2
D Question... Record retention Prior to FDA NDA application and approval Records and Data - Quality, Legal and Other Evidence 3
L Regarding CDRH Cover Sheet and prior related Submissions Other Medical Device and Orthopedic Related Topics 3
P Should I send survey questionnaire prior to Supplier EMS Audit? Miscellaneous Environmental Standards and EMS Related Discussions 14
C Re-Verification of Nonconforming Product prior to closing our DMR AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
C New decree in France: Authorization required prior advertising Medical Devices EU Medical Device Regulations 1
J SFDA Labeling Requirements prior to import to any Chinese territory China Medical Device Regulations 2
S Charging Consulting Fees prior to 510(k) Approval Other US Medical Device Regulations 3
Q Company acquisition prior to QMS and ISO Audits ISO 13485:2016 - Medical Device Quality Management Systems 1
Roberticus Responsibility for Outsourced Process, ALL in-house Mfg/QC done (prior FG PN) ISO 13485:2016 - Medical Device Quality Management Systems 5
M Internal Audit of our ISO 9001 Quality Systems prior to the Registration Audit Internal Auditing 9
J Product Designed prior to Implementation of AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
L Sigma & the Distributor - Rejected pieces from prior shipments to current measures Six Sigma 7
L Method for Poka-Yoke (Mistake Proofing) Validation prior to SOP APQP and PPAP 6
C How do you Validate Solderability of Components Prior to Soldering Operations Manufacturing and Related Processes 5
B Samples for Customer Evaluation prior to Full Release 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
K When to do Final Release if Medical Device goes off site prior to shipping? ISO 13485:2016 - Medical Device Quality Management Systems 5

Similar threads

Top Bottom