Hello everyone, I am developing a class IIb laser / robotics device for the US market (through 510k) and was wondering if using an accredited lab for our test reports (60601-1, 60601-1-2, 60601-2-22) is a requirement by the FDA? The plan is to write these reports ourselves from the data that we get from a non-accredited lab here in Denmark. I understand if the FDA prefers seeing reports written by accredited test labs, but if it isn't strictly necessary, then I would like to give the in-house report writing a shot.
Thank you
Elsmar Forum Sponsor
yodon
Staff member
Super Moderator
#2
Don't know, never tried, probably wouldn't want to try! :) I'm sure the first question would be to ask if the data are reliable. So while it may be possible, it's likely to get scrutinized quite heavily.
On a bit of a side note, FDA released guidance last year on an Accreditation Scheme for Conformity Assessment (ASCA) Pilot Program. So it would seem they are wanting more than even being accredited; now they want the lab to undergo their accreditation scheme. I believe I saw that UL actually achieved this.
Peter Selvey
Staff member
Super Moderator
#3
I would suggest adding more information in the report.
At one extreme the evidence could just be a list of clauses in a table with simply "Pass" or "N/A" written, which I have sometimes seen as an MDD auditor. A report like that should fail a regulatory review.
The other extreme would be detailed report showing where necessary a reasoned assessment, links to supporting material and test data, and where tests are done, sample identification, raw data, calculations, equipment used, calibration status, test date, location, environment, notes etc, plus keep on file the qualifications of persons actually doing the tests and actual calibration records.
That's the kind of stuff an accredited laboratory has to keep but they may not put it all in the report. That's the benefit of accreditation.
One option would be to ask your non-accredited lab to bundle all these background records either in the report, or as part of package, which is then reviewed for completeness by your company. Another option would be to audit them (as a critical subcontractor) and prepare an agreement that they will keep these records on your behalf for the period required by regulations, and keep records of the audit/agreement in your supplier assessments.
For IEC 60601-1-2 the non-accredited status could be more difficult to defend as the actual tests and equipment are highly specialised and it's unlikely someone in your company will have the ability to audit them or review the records (the auditor also needs to be qualified in the field). I think it's better to use an accredited lab for EMC if possible.
IEC 60601-1 tests are not that complicated, so it should be OK. The main problem with IEC 60601-1 is just that it's a huge standard to manage.
For IEC 60601-2-22, chances are your company will have more experience than the lab, so that should be fine. You could even prepare your own report in-house, again making sure to keep all the necessary background records.
Thread starter Similar threads Forum Replies Date
D IATF 16949 SI 10, External non-accredited lab IATF 16949 - Automotive Quality Systems Standard 4
E Accredited vs. non-accredited labs for 60601 compliance in the US Other Medical Device Related Standards 4
W ISO standards /Social Audit Franchise (Accredited & Non Accredited) Service Industry Specific Topics 7
S CB accredited by non-aligned AB - industry views? ISO 13485:2016 - Medical Device Quality Management Systems 1
N Accredited Tests and Non-Accredited Tests - Differences General Measurement Device and Calibration Topics 4
R ISO registered company list _ those certified by non-accredited auditing firms ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
B Non UKAS / EA / IAF accredited certification bodies - Certification company QMS Registrars and Notified Bodies 12
Sidney Vianna Accredited registrars issuing non accredited certificates Registrars and Notified Bodies 38
C Are accredited calibration laboratories more profitable than non-accredited General Measurement Device and Calibration Topics 4
L Acceptable use of a non accredited external ISO/IEC 17025 laboratory ISO 17025 related Discussions 5
L A2LA - Sending testing to a non-accredited lab if customer doesn't require A2LA? General Measurement Device and Calibration Topics 2
G Compliance with ISO 9001-2015 for ISO 17025 Accredited Labs? ISO 17025 related Discussions 5
AllTheThings ISO-17025 accredited TEST lab performing internal calibrations (Yes: Here is how) ISO 17025 related Discussions 6
M Must a company be 17025 accredited to perform internal calibrations? ISO 17025 related Discussions 7
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
M Calibration Certificate Result issued by an accredited external laboratory General Measurement Device and Calibration Topics 9
N Database that lists labs accredited to ASTM / ANSI standards Various Other Specifications, Standards, and related Requirements 6
G What to do with correction factors overlooked by an accredited standard (6.4.11)? ISO 17025 related Discussions 5
B Directory of Largest Global Certification Bodies (Accredited) ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 3
S Internal calibrations - Part of an ISO 17025 accredited testing laboratory (Automotive) ISO 17025 related Discussions 16
B Accredited laboratory or not under MDD 93/42/EEC IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
D Who are the accredited Certified Bodies for CE Mark for Medical Devices? CE Marking (Conformité Européene) / CB Scheme 7
Jen Kirley Is your registrar (CB) accredited? ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 9
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate Qualification and Validation (including 21 CFR Part 11) 3
M Accredited Calibration Sevice Provider using computerized system to issue calibration certificate General Measurement Device and Calibration Topics 2
Sidney Vianna Interesting Discussion Legal compliance as part of ISO 45001 accredited certification. Major OSHA penalties in the USA. Occupational Health & Safety Management Standards 15
Q ISO 14001:2015 accredited but import most product - CB Raised minor NC on Life Cycle Perspective ISO 14001:2015 Specific Discussions 2
I IATF 16949, accredited logos, calibration capabilities, etc. IATF 16949 - Automotive Quality Systems Standard 17
I ISO 9001 and VWAD (Verified-Accredited Wholesale Distributors certification) Coffee Break and Water Cooler Discussions 1
B AS9100 - Are special process suppliers required to be NADCAP accredited? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
I FAR Part 43 and Application to 17025 Accredited Cal Lab Federal Aviation Administration (FAA) Standards and Requirements 2
S Is it mandatory for a Calibration Lab to be ISO 17025 accredited? General Measurement Device and Calibration Topics 7
Sidney Vianna IAF Endorsed Accredited 13485 Certification helping Control of Medical Distributors ISO 13485:2016 - Medical Device Quality Management Systems 0
M BS 10008:2014 Registration - UKAS accredited company in the UK to assess our company IEC 27001 - Information Security Management Systems (ISMS) 4
JoCam Use of ISO Logos if company is not certified by an accredited certification body ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 20
P ISO 9001 Requirements for ISO 17025 Accredited Calibration Laboratory ISO 17025 related Discussions 5
C CB Accreditaton - CB that was accredited by IAS Registrars and Notified Bodies 5
Crusader Calibration Source not ISO 17025 accredited ISO 17025 related Discussions 6
S Do Microbiology Testing Labs need to use ISO 17025 accredited vendors? ISO 17025 related Discussions 2
L What are Accredited laboratories? (Medical Device Class IIb Related) EU Medical Device Regulations 1
Sidney Vianna Interesting Discussion IAF CertSearch Database - Repository of "properly" Accredited Management System Certs ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 80
E Should companies that perform calibration services be ISO 17025 accredited? ISO 17025 related Discussions 13
DietCokeofEvil How is it that gage manufacturers are not required to be accredited? General Measurement Device and Calibration Topics 9
A How many laboratories are accredited to ISO 17025 in the USA? ISO 17025 related Discussions 5
B ISO 17025 Accredited Certificate Requirements ISO 17025 related Discussions 2
Ajit Basrur ICCBBA Is Now an FDA-Accredited Issuing Agency for Unique Device Identifiers (UDI) Other US Medical Device Regulations 1
U Once upon a QMS - Is a Statement of Uncertainty a must for ISO 17025 accredited labs? ISO 17025 related Discussions 8
N How Do I Make a Business Plan - 17025 Accredited Laboratory ISO 17025 related Discussions 3
N Internal Calibration vs. Accredited Calibration for Balance - Differences Calibration Frequency (Interval) 3
T To establish ISO 9001:2008 in a ISO 17025 accredited Company ISO 17025 related Discussions 4
Similar threads
Top Bottom