Definition Accuracy and Resolution in a Measurement System: What are the differences?

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#11
Wait a minute! Is the micrometer actually used in service to measure dimensions with the smaller tolerance?
Good point, if the "tolerance" specified on the NC is what they are trying to measure on the part, then the gage is useless. If the tolerance is the gage tolerance, then there really is no basis for complaint (it just looks like a standard gage spec, see this) - the NC needs to identify the relationship to the part tolerance to be of an issue.
 
Elsmar Forum Sponsor

msbettyhunt

Starting to get Involved
#12
Hi! what I found the main difference between accuracy and resolution in measurement is that accuracy can be defined as the amount of uncertainty in measurement with respect to an absolute standard while on the other hand resolution refers to the range of the scale divided by the display readout. You could also say that it is the capacity divided by readability. High readability delivers a great resolution, but it doesn’t always result in the best accuracy.
 

bobdoering

Stop X-bar/R Madness!!
Trusted Information Resource
#13
My favorite example of this is the 1/100" steel rule. Accuracy would be how well the rule is marked to the standard. Resolution is 1/100" but there ends up being a lot of error in readability trying to see or count the lines - especially without magnification (an opportunity to increase readability. The gage "system" includes the person that had to read it. I had a gage R&R for a hand gage that failed because one of the operators had taken Excedrin and was shaking enough to affect the repeatability!
 
Thread starter Similar threads Forum Replies Date
P Measurement Equipment - 10:1, 4:1 - Is it resolution or accuracy General Measurement Device and Calibration Topics 1
C Gage Resolution / Measurement Accuracy - Laser scan micrometer has a 6 digit display General Measurement Device and Calibration Topics 4
K Micrometer Calibration - Resolution, Accuracy and Range General Measurement Device and Calibration Topics 16
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Analog Pressure Gauge with no EN837 accuracy class information General Measurement Device and Calibration Topics 5
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Warning about Deltronic Gage Pins - The "Certification of Accuracy" document General Measurement Device and Calibration Topics 13
G Posting Measuring Equipment Accuracy for User Information General Measurement Device and Calibration Topics 4
S Gagemaker MT-3000 accuracy for calibration Calibration and Metrology Software and Hardware 2
J CMM Accuracy Formula General Measurement Device and Calibration Topics 3
L Calibration Accuracy In Relation To Reporting General Measurement Device and Calibration Topics 1
Q Acceptable calibration accuracy of a 60" linear measuring device General Measurement Device and Calibration Topics 16
M Calibration certificates avowing accuracy beyond the uncertainty limits? Measurement Uncertainty (MU) 7
S Jewelry vs. Normal Laboratory Balances - Accuracy and calibration General Measurement Device and Calibration Topics 2
T Tape Measure, the right device for 2 or 3 decimals accuracy on prints General Measurement Device and Calibration Topics 8
W Equipment for Oven Calibration, Uniformity and System Accuracy Tests AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
S Opinions on the "Best" CMM for High Accuracy Machining General Measurement Device and Calibration Topics 9
J AMS 2750E System Accuracy Test (SAT) Frequency AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
A Process to Validate the Accuracy of Test Reports (AS9100) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 38
D AMS 2750E - Digital Instrument Calibration Accuracy Requirements General Measurement Device and Calibration Topics 1
shimonv Soldering Station Accuracy out of specification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Help trying to explain accuracy procedure General Measurement Device and Calibration Topics 1
D Long length measurement of 18" up to 100" with an accuracy of .005 min General Measurement Device and Calibration Topics 8
N Cry for help - FDA calibration requirement for accuracy & precision General Measurement Device and Calibration Topics 2
C Accuracy Equation - Linear gage with a stated accuracy of (1.5+L/50) ?m L= mm General Measurement Device and Calibration Topics 5
W Dimensional Measurement of 10% accuracy and 25% uncertainty standard General Measurement Device and Calibration Topics 6
I UUT Accuracy as part of Uncertainty Budget Measurement Uncertainty (MU) 6
A Uncertainty Measurement - What this accuracy specification "1% + 10^5": means? General Measurement Device and Calibration Topics 2
L Inspection Equipment Accuracy - Measuring the diameters of small holes Measurement Uncertainty (MU) 5
E How does Non-Linear UUT affect TAR (Test Accuracy Ratio) Calculation? Measurement Uncertainty (MU) 3
G Digital Autoclave Accuracy Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
S AMS 2750E Heat Treat Furnace System Accuracy Test SAT Report question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
A Accuracy for electrosurgical analyzer General Measurement Device and Calibration Topics 2
A Calculating Accuracy for a Multimeter General Measurement Device and Calibration Topics 1
D Verification of weighing scale 150kg capacity 0.05g accuracy General Measurement Device and Calibration Topics 9
A Accuracy Calculation of Leakage Current for an Electrical Safety Analyzer General Measurement Device and Calibration Topics 5
G Laser Precision Linear Measurement - Inherrent Accuracy of 10 -6 - Nice PDF Paper General Measurement Device and Calibration Topics 1
S Stated Accuracy of Portable CMM Arm and Real World Use Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
A Intrinsic Uncertainty and Accuracy of a Medical Electrical Safety Analyzer General Measurement Device and Calibration Topics 1
V CMM plane construction method types and accuracy differences for measuring Pos Tol Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
J Weighing Balance Accuracy for Plastic Moulded Parts variation study Manufacturing and Related Processes 3
G Best X-Y Microscope Stage for Submicron Accuracy Capability, Accuracy and Stability - Processes, Machines, etc. 2
K Wilson Hardness Tester Model 525 Accuracy/Repeatability issues help General Measurement Device and Calibration Topics 4
P Measurement Uncertainty and Equipment Accuracy Measurement Uncertainty (MU) 2
C Basic Internal Calibration Procedure to check the Accuracy of a Balance Calibration Frequency (Interval) 6
C Accuracy Problem on the Calibrator - Two Labs with different Results General Measurement Device and Calibration Topics 7
R How do I extend Correction factor of Calibration cert to compute instrument accuracy Measurement Uncertainty (MU) 1
S Accuracy of a Measurement Device Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
C Simpson 260 Series 8 Equipment Accuracy General Measurement Device and Calibration Topics 5

Similar threads