Accuracy Calculation of Leakage Current for an Electrical Safety Analyzer

A

akmal122g

#1
Hello. I am trying to calculate the accuracy of leakage current for an electrical safety analyzer. The accuracy for leakage current is

?5% of reading ?1mAmp DC and 48 Hz to 100 kHz

The nominal value for calibration is 0.100 mAmp. Can someone help me how to calculate the accuracy?
 
Elsmar Forum Sponsor

Hershal

Metrologist-Auditor
Staff member
Super Moderator
#2
Based on the numbers, looks like the reading desired is 0.1mA, and the unit accuracy comes out to 0.1005mA, and that would be across the listed frequency range, and also DC.
 
N

notadog

#3
Good Afternoon, This is my first attempt at answering someones question. I suggest downloading from a reliable website, one that doesn't have any viruses, "TolCal4". I have used it for a number of years now. It has its problems, but I can't do without it.
 

dgriffith

Quite Involved in Discussions
#4
Hello. I am trying to calculate the accuracy of leakage current for an electrical safety analyzer. The accuracy for leakage current is

?5% of reading ?1mAmp DC and 48 Hz to 100 kHz

The nominal value for calibration is 0.100 mAmp. Can someone help me how to calculate the accuracy?
I was going to try my hand at this but hesitated because it doesn't make sense to me, else I'm missing something. I'd like to see this explained further.
Is this an AC spec within the listed frequencies, plus or minus some allowed DC offset?


My take is ?5% of reading ?1mAdc; the reading in question is the calibration input, 100 ?A (0.100 mAmp).
So, 5% of 100 ?A is 5 ?A, therefore: 100?A ?5?A
I don't know what to do with ?1mAdc. . . need more info.
 
N

notadog

#5
The calculator wants to know what is the resolution of UUT? After looking a little further, I suspect that the resolution is .001. If that is the case, then the calculation is 1.105 mA.
 
Last edited by a moderator:

dgriffith

Quite Involved in Discussions
#6
Hello. I am trying to calculate the accuracy of leakage current for an electrical safety analyzer. The accuracy for leakage current is

?5% of reading ?1mAmp DC and 48 Hz to 100 kHz

The nominal value for calibration is 0.100 mAmp. Can someone help me how to calculate the accuracy?
I think Hershal had it right-- the spec should read (?5% reading ?1mAmp) at DC and from 48 Hz to 100 kHz. Makes more sense.

zakmal122g, What is this the specification of? If it's the instrument you are making the 100?A measurement with, then you have the wrong instrument. We are just estimating based on the information you gave: we're filling in the blanks, so to speak.
I wouldn't be making a 100?A measurement with something whose accuracy is 10 times greater at 1mA, no matter how many zeros/decimal places the display can indicate. It isn't good metrology.
 
Last edited:
Thread starter Similar threads Forum Replies Date
E How does Non-Linear UUT affect TAR (Test Accuracy Ratio) Calculation? Measurement Uncertainty (MU) 3
J Setting accuracy limit for a read out General Measurement Device and Calibration Topics 4
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Analog Pressure Gauge with no EN837 accuracy class information General Measurement Device and Calibration Topics 5
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Warning about Deltronic Gage Pins - The "Certification of Accuracy" document General Measurement Device and Calibration Topics 13
G Posting Measuring Equipment Accuracy for User Information General Measurement Device and Calibration Topics 4
S Gagemaker MT-3000 accuracy for calibration Calibration and Metrology Software and Hardware 2
J CMM Accuracy Formula General Measurement Device and Calibration Topics 3
L Calibration Accuracy In Relation To Reporting General Measurement Device and Calibration Topics 1
Q Acceptable calibration accuracy of a 60" linear measuring device General Measurement Device and Calibration Topics 16
M Calibration certificates avowing accuracy beyond the uncertainty limits? Measurement Uncertainty (MU) 7
S Jewelry vs. Normal Laboratory Balances - Accuracy and calibration General Measurement Device and Calibration Topics 2
T Tape Measure, the right device for 2 or 3 decimals accuracy on prints General Measurement Device and Calibration Topics 8
W Equipment for Oven Calibration, Uniformity and System Accuracy Tests AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
S Opinions on the "Best" CMM for High Accuracy Machining General Measurement Device and Calibration Topics 9
Q AMS 2750E System Accuracy Test (SAT) Frequency AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
A Process to Validate the Accuracy of Test Reports (AS9100) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 38
D AMS 2750E - Digital Instrument Calibration Accuracy Requirements General Measurement Device and Calibration Topics 1
shimonv Soldering Station Accuracy out of specification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Help trying to explain accuracy procedure General Measurement Device and Calibration Topics 1
D Long length measurement of 18" up to 100" with an accuracy of .005 min General Measurement Device and Calibration Topics 8
N Cry for help - FDA calibration requirement for accuracy & precision General Measurement Device and Calibration Topics 2
C Accuracy Equation - Linear gage with a stated accuracy of (1.5+L/50) ?m L= mm General Measurement Device and Calibration Topics 5
W Dimensional Measurement of 10% accuracy and 25% uncertainty standard General Measurement Device and Calibration Topics 6
I UUT Accuracy as part of Uncertainty Budget Measurement Uncertainty (MU) 6
A Uncertainty Measurement - What this accuracy specification "1% + 10^5": means? General Measurement Device and Calibration Topics 2
L Inspection Equipment Accuracy - Measuring the diameters of small holes Measurement Uncertainty (MU) 5
G Digital Autoclave Accuracy Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
S AMS 2750E Heat Treat Furnace System Accuracy Test SAT Report question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
A Accuracy for electrosurgical analyzer General Measurement Device and Calibration Topics 2
A Calculating Accuracy for a Multimeter General Measurement Device and Calibration Topics 1
D Verification of weighing scale 150kg capacity 0.05g accuracy General Measurement Device and Calibration Topics 9
G Laser Precision Linear Measurement - Inherrent Accuracy of 10 -6 - Nice PDF Paper General Measurement Device and Calibration Topics 1
S Stated Accuracy of Portable CMM Arm and Real World Use Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
A Intrinsic Uncertainty and Accuracy of a Medical Electrical Safety Analyzer General Measurement Device and Calibration Topics 1
V CMM plane construction method types and accuracy differences for measuring Pos Tol Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
J Weighing Balance Accuracy for Plastic Moulded Parts variation study Manufacturing and Related Processes 3
G Best X-Y Microscope Stage for Submicron Accuracy Capability, Accuracy and Stability - Processes, Machines, etc. 2
K Wilson Hardness Tester Model 525 Accuracy/Repeatability issues help General Measurement Device and Calibration Topics 4
P Measurement Uncertainty and Equipment Accuracy Measurement Uncertainty (MU) 2
C Basic Internal Calibration Procedure to check the Accuracy of a Balance Calibration Frequency (Interval) 6
C Accuracy Problem on the Calibrator - Two Labs with different Results General Measurement Device and Calibration Topics 7
R How do I extend Correction factor of Calibration cert to compute instrument accuracy Measurement Uncertainty (MU) 1
P Measurement Equipment - 10:1, 4:1 - Is it resolution or accuracy General Measurement Device and Calibration Topics 1
S Accuracy of a Measurement Device Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
C Simpson 260 Series 8 Equipment Accuracy General Measurement Device and Calibration Topics 5
H Accuracy and Precision of Test Equipment required? General Measurement Device and Calibration Topics 2

Similar threads

Top Bottom