Accuracy for electrosurgical analyzer
akmal122g
Hello. I am trying to calibrate an analyzer but do not understand the accuracy of the equipment.
Technical specification
Main test load range: 50 ohm to 750 ohm
Accuracy (DC to 500 kHz): ?4% of selected value measured at calibration
to: ?1% (across the entire operating temperature
range)
Can someone help me with this?
Thank you.
Hershal
Metrologist-Auditor
range)
This appears to be DC and AC Voltage and Resistance. The statement above suggests it electrical output will be within the 4% of selected value regardless of what point in its operating range it happens to be at. But based on the above, appears that the total accuracy, including resistance and voltage cannot exceed 1%.
I would likely start with the voltage, DC, take several steps to 500kHz, then go back down for linearity. Two or three times to also assure the repeatability. Measure the resistance at DC across several values, but then you need to measurethe resistance in an AC manner also.
Then calculate your Type A and Type B uncertainties and your expanded uncertainties. Hopefully that will be much less than the stated accuracy levels.
Hope this helps.
J Setting accuracy limit for a read out General Measurement Device and Calibration Topics 4
R Clinical accuracy and repeatability of IR(infrared) thermometer, no maximum error criteria is recommended in ISO Other Medical Device Related Standards 11
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
H Analog Pressure Gauge with no EN837 accuracy class information General Measurement Device and Calibration Topics 5
shimonv IEC 60601-1 Essential Performance - Is the signal accuracy specification an essential requirement? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
G Warning about Deltronic Gage Pins - The "Certification of Accuracy" document General Measurement Device and Calibration Topics 13
G Posting Measuring Equipment Accuracy for User Information General Measurement Device and Calibration Topics 4
S Gagemaker MT-3000 accuracy for calibration Calibration and Metrology Software and Hardware 2
J CMM Accuracy Formula General Measurement Device and Calibration Topics 3
L Calibration Accuracy In Relation To Reporting General Measurement Device and Calibration Topics 1
Q Acceptable calibration accuracy of a 60" linear measuring device General Measurement Device and Calibration Topics 16
M Calibration certificates avowing accuracy beyond the uncertainty limits? Measurement Uncertainty (MU) 7
S Jewelry vs. Normal Laboratory Balances - Accuracy and calibration General Measurement Device and Calibration Topics 2
T Tape Measure, the right device for 2 or 3 decimals accuracy on prints General Measurement Device and Calibration Topics 8
W Equipment for Oven Calibration, Uniformity and System Accuracy Tests AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
S Opinions on the "Best" CMM for High Accuracy Machining General Measurement Device and Calibration Topics 9
Q AMS 2750E System Accuracy Test (SAT) Frequency AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
A Process to Validate the Accuracy of Test Reports (AS9100) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 38
D AMS 2750E - Digital Instrument Calibration Accuracy Requirements General Measurement Device and Calibration Topics 1
shimonv Soldering Station Accuracy out of specification 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Help trying to explain accuracy procedure General Measurement Device and Calibration Topics 1
D Long length measurement of 18" up to 100" with an accuracy of .005 min General Measurement Device and Calibration Topics 8
N Cry for help - FDA calibration requirement for accuracy & precision General Measurement Device and Calibration Topics 2
C Accuracy Equation - Linear gage with a stated accuracy of (1.5+L/50) ?m L= mm General Measurement Device and Calibration Topics 5
I UUT Accuracy as part of Uncertainty Budget Measurement Uncertainty (MU) 6
A Uncertainty Measurement - What this accuracy specification "1% + 10^5": means? General Measurement Device and Calibration Topics 2
L Inspection Equipment Accuracy - Measuring the diameters of small holes Measurement Uncertainty (MU) 5
E How does Non-Linear UUT affect TAR (Test Accuracy Ratio) Calculation? Measurement Uncertainty (MU) 3
G Digital Autoclave Accuracy Requirements ISO 13485:2016 - Medical Device Quality Management Systems 2
S AMS 2750E Heat Treat Furnace System Accuracy Test SAT Report question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
A Calculating Accuracy for a Multimeter General Measurement Device and Calibration Topics 1
D Verification of weighing scale 150kg capacity 0.05g accuracy General Measurement Device and Calibration Topics 9
A Accuracy Calculation of Leakage Current for an Electrical Safety Analyzer General Measurement Device and Calibration Topics 5
G Laser Precision Linear Measurement - Inherrent Accuracy of 10 -6 - Nice PDF Paper General Measurement Device and Calibration Topics 1
S Stated Accuracy of Portable CMM Arm and Real World Use Inspection, Prints (Drawings), Testing, Sampling and Related Topics 9
A Intrinsic Uncertainty and Accuracy of a Medical Electrical Safety Analyzer General Measurement Device and Calibration Topics 1
V CMM plane construction method types and accuracy differences for measuring Pos Tol Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
J Weighing Balance Accuracy for Plastic Moulded Parts variation study Manufacturing and Related Processes 3
G Best X-Y Microscope Stage for Submicron Accuracy Capability, Accuracy and Stability - Processes, Machines, etc. 2
K Wilson Hardness Tester Model 525 Accuracy/Repeatability issues help General Measurement Device and Calibration Topics 4
P Measurement Uncertainty and Equipment Accuracy Measurement Uncertainty (MU) 2
C Basic Internal Calibration Procedure to check the Accuracy of a Balance Calibration Frequency (Interval) 6
C Accuracy Problem on the Calibrator - Two Labs with different Results General Measurement Device and Calibration Topics 7
R How do I extend Correction factor of Calibration cert to compute instrument accuracy Measurement Uncertainty (MU) 1
P Measurement Equipment - 10:1, 4:1 - Is it resolution or accuracy General Measurement Device and Calibration Topics 1
S Accuracy of a Measurement Device Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
C Simpson 260 Series 8 Equipment Accuracy General Measurement Device and Calibration Topics 5
H Accuracy and Precision of Test Equipment required? General Measurement Device and Calibration Topics 2