Addition of IFU on local language by distributor - Article 16?
MDRResearcher
Hello everybody!
Could you please help us with a case study:
Distributor adds instructions for use on local language in addition to IFU on English added by manufacturer (inside the secondary packaging box of medical device).
Does this action fall under article 16.2.a) and requires certification and notification?
Is there a difference if the manufacturer supplies the IFU on local language and the distributor only adds it to the box (in relation to the translation beeing performed by distributor)?
Kind regards
htcoztrk
Hello MDRResearcher,
I do not know much about the translation article yet but I found a white paper which also covers generally about the translation that I am hoping to be helpful to you.
If you ask for my opinion, I believe according to MDR Art. 16, point 3, whether distributor translates the IFU or only put the it into the box, requires these activities to be mentioned on the device/ its packaging or accompanying document: ‘A distributor or importer that carries out any of the activities mentioned in points (a - includes translation) and (b - includes repackaging) of paragraph 2 shall indicate on the device or, where that is impracticable, on its packaging or in a document accompanying the device, the activity carried out together with its name, registered trade name or registered trade mark, registered place of business and the address at which it can be contacted, so that its location can be established.' If I know correctly there are recommended symbols to be used for translation and repacking by third parties other than legal manufacturer. You can have a look also to MedTech Europe's guide.
Best Regards,
Hatice
Q Considering addition of mass calibration to ISO17025 scope ISO 17025 related Discussions 6
D Distribution Center mandatory addition to Certificate? IATF 16949 - Automotive Quality Systems Standard 3
dgrainger New decision by MHRA on disclaimers - addition of "manner of use as perceived by the consumer" UK Medical Device Regulations 1
O In addition to the standard, what other ISO 13485 sources do people recommend? ISO 13485:2016 - Medical Device Quality Management Systems 5
shrutisancheti Addition of flat panel detector EU Medical Device Regulations 1
I Importer Address in addition to Manufacturer details on label EU Medical Device Regulations 13
S Is Lot Number required in addition to UDI serial number? Other Medical Device Regulations World-Wide 7
C Addition of Design & Manufacture - CE Mark Reissue? EU Medical Device Regulations 1
K Addition of a test reader to existing CE marked product. Other ISO and International Standards and European Regulations 2
M CO2 Incubators - Equipment Qualifications in addition to Calibration Qualification and Validation (including 21 CFR Part 11) 5
Q Actions Taken - Updating a FMEA for the addition of a downstream poke yoke FMEA and Control Plans 3
E Addition of Label by Initial Importer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S Addition in the Scope of Quality Manual ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
S Is TGA approval required for a ARTG listed IVD for a sample type addition? Other Medical Device Regulations World-Wide 11
WCHorn Informational How the addition of "Risk" will affect ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 103
G Document Control - Are hard copies required in addition to soft copies maintained? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
S MDD Technical File CE Mark - Addition of a Device to our Current CE Marked Family EU Medical Device Regulations 4
D Packaging Addition - Non-sterile barrier to protect the whole package 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 12
P Addition of Testing to Our ISO 17025 Scope - How? ISO 17025 related Discussions 1
E Addition of French Language (translation) - New 510(k) required? Other US Medical Device Regulations 1
somashekar ISO 13485 & ISO 9001 certified with addition of site. Covegratulations 10
N AS9100 Certification in addition to TS-16949 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
J Addition of a Class 1 (Sterile) Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 3
L CEDAC (Cause and Effect with Addition of Cards) - Sharing Experiences and Form(s) Training - Internal, External, Online and Distance Learning 15
K Differences - ISO17025:1999 vs. ISO17025:2005 - The addition of "Improvement" in 4.10 ISO 17025 related Discussions 16
S Value addition to product ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A Validation of Processes in addition to Customer Requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
J Differences - ISO 13485 - Requirements in Addition to ISO 9001:2000 ISO 13485:2016 - Medical Device Quality Management Systems 25
C Is it necessary to have a documented 3 R recycling plan in addition to measures? Miscellaneous Environmental Standards and EMS Related Discussions 10
A ISO 80601-2-30 - the rated range of cuff pressure in the IFU Other Medical Device Related Standards 0
K 2021/2226 on Electronic Instructions for Use (IFU) for medical devices Article 22 Packs EU Medical Device Regulations 1
B X-Ray Sensor Device Symbols for IFU Medical Device and FDA Regulations and Standards News 1
D Basic udi-di to not appear on IFU? EU Medical Device Regulations 4
M Is there a way to obtain redacted copies of IFU from a different manufacturer? EU Medical Device Regulations 2
D Labelling and IFU deadlines EU Medical Device Regulations 3
K Different Indications for Use in Different Countries -- how to handle IFU? Other Medical Device Regulations World-Wide 4
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
M MDR - eIFU - When you can supply an electronic IFU instead of a paper IFU EU Medical Device Regulations 6
J Warnings/Cautions in Medical Device IFU Medical Device and FDA Regulations and Standards News 4
Ed Panek IFU Contact Requirements - FDA and MDD/MDR US Food and Drug Administration (FDA) 1
C Does an accessory need an IFU if it use is discussed in the Parent device IFU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
R MDR, Annex I, 23.1 Interpretation - IFU on the website EU Medical Device Regulations 5
J EU 207/2012 E-IFU Service Video EU Medical Device Regulations 1
A Which guidance do you use for guiding the requirements of IFU for 510(k) submission? Medical Device and FDA Regulations and Standards News 5
U Intended Users in IFU (Instructions for Use) EU Medical Device Regulations 5
B Do IFU designs have to be document controlled under ISO 13485? Document Control Systems, Procedures, Forms and Templates 2
K No IFU, where the requirements would go? EU Medical Device Regulations 5
P IFU Translation Requirements for EU EU Medical Device Regulations 2
V Preparing the IFU in compliance with MDR 745, Chapter III EU Medical Device Regulations 2
E Commission Regulation (EU) No. 207/2012 E-IFU Applicability EU Medical Device Regulations 5