Additional Isolation Required? Doppler Probe Desktop Device

F

Fred1977

Hi All,
I'm working on a medical device and am a bit lost at sea as to whether isolation is required or not. The device consists of a small desktop unit with TFT display, buttons and USB socket for PC connect. The desktop unit is battery powered (4x1.5V), or can get 5V from USB. A doppler ultrasound probe (huntleigh) is plugged into the desktop unit via a cable. Typically the operator holds the probe against patients neck for about 30 seconds at a time to capture a signal.

The device presently contains no isolation. I gather the probe can be classified as a type B applied part, and in 60601-1 2006 I read this:

8.5.2.2 Type B applied parts:
...Type B part that is NOT protectively earthed shall be seperated by one MOPP from metal accessible parts that are nont protecticely earthed, unless:
- Metal accesible part is physically contiguous with applied part, or
- the risk that the metal accesible part will make contact with a source of voltage or leakage current above permitted limits is acceptably low.

Does the PC connection mean quite high leakage current is possible, and thus neither exclusion paragraph applies and I need that one MOPP isolation? From some reading I gather the USB port already offers one MOPP, in which case I would need no further isolation... but I am not certain that the 'PC USB port already offers one MOPP' statement is true or not.

Secondly, if I read further into the 'components and wiring' section, I find this:
8.10.4.1 Limitation of operating voltages:
Cord-connected hand held control devices of ME equipment and their associated cords shall contain only conductors operating at voltages =< 60Vdc, in circuits isolated from MAINS part by two MOPPs.

In the case of the above, does this mean I need two MOPPS? I would think that the doppler probe is not a 'hand held control', or is it? If this paragraph does apply to me, it just leads to more questions. The doppler probe itself allegedly includes 1 MOPP of seperation by virtue of its plastic casing design (I am awaiting confirmation from Huntleigh), in which case if the probe has a MOPP, and the PC USB port already offers a MOPP, I might be safe.. however the above paragraph says the conductors within the cord are also subjkect to the 2 MOPPs isolation, which means I need to isolate the doppler inside my desktop case housing.

Any guidance will be most appreciated, I fail to see any risk in the product's current un-isolated configuration, and would hate to have to add layers of expensive isolation if I don't need to.

Many thanks!
 

Peter Selvey

Leader
Super Moderator
Generally for ultrasound probes the plastic cover (acoustic coupling lens with waterproofing) is considered the isolating part which provides the Type BF isolation (1500Vrms) and also protects against the internal voltages. Type BF is not required but most manufacturers use Type BF and it makes sense because you also have to isolate against the internal transducer drive voltages.

These drive voltages can be up to 200Vp, although the rms value is low at around 5Vrms. Some might argue that presence of 200V means that 4000Vrms isolation but this is not correct due to the low rms voltage. It's a bit complicated, but the 1500Vrms for the Type BF is far in excess of what is reasonable for the situation given the low rms.

The Type BF isolation also provides additional isolation for use with a PC. The PC provides 2MOOP but only 1 MOPP. With the probe insulation 2MOPP is provided to the patient.

The clause mentioned (8.10.4.1) only requires 2 MOP (not 2 MOPP). It practice this could means 2 MOOP or 2 MOPP depending on who uses the control (operator or patient). In your case, it would be 2 MOOP, which the PC already provides, so no further action required.

I think it all fits with common sense and the standard.
 
F

Fred1977

Peter,
Thank you for a crystal clear reply and one which I am very pleased to hear.
Regards,
Fred
 

Roland chung

Trusted Information Resource
Generally for ultrasound probes the plastic cover (acoustic coupling lens with waterproofing) is considered the isolating part which provides the Type BF isolation (1500Vrms)

Thermal cycling test specified in clause 8.9.3 applies to the ultrasound probe. However, I am a little bit confused about the sub-clause 8.9.3.2 and 8.9.3.3. Which one do you think we should carry out?

8.9.3.2 Insulating compound forming solid insulation between conductive parts
8.9.3.3 Insulating compound forming a cemented joint with other insulating parts

Thank you,
Roland
 
Last edited by a moderator:

Peter Selvey

Leader
Super Moderator
Yes, it's confusing. For a previous situation (also an ultrasound) we concluded 8.9.3.3 was appropriate, but it was not clear. Perhaps someone with an IEC 60950 background might be able to help (the tests seem to be borrowed from 950).

Of interest, we found that multiplying the voltages by 1.6 was too much for a typical medical ultrasound, especially for endoscopic applications. Already in the 1500Vrms there is a number of worst case assumptions which don't exist in the real world, it's overkill but it's possible to comply. However, with the 1.6 factor the voltage can be too much.
 

Roland chung

Trusted Information Resource
Yes, the thermal cycling test in IEC 60601-1 is derived from IEC 60950-1. Both standards treat the cemented joints in the same way. But for insulating compound used as solid insulation, 950 does not require the electric strength test voltage to be multiplied by 1.6.

Why do you think subclause 8.9.3.3 is appropriate for the ultrasound probes?
 

Peter Selvey

Leader
Super Moderator
Again, not really sure on this one ... it seems like 8.9.3.2 is for cases where an for example epoxy is used to fill a gap with the test to make sure there are no bubbles etc, whereas 8.9.3.3 is for cases where an adhesive is used in a joint between two pieces of solid insulation, to make sure the seal is effective in the long term.

I could imagine the case where an epoxy is used on the surface of a PCB where the spacings are way too small (less than clearance limits). In that case both clauses could apply: 8.9.3.2 to make sure that the epoxy has no bubbles, and 8.9.3.3 to make sure the seal on the surface of the PCB is OK.

But, if the original spacing is OK for clearance, and only fails creepage, then only 8.9.3.3 is required (bubbles are not a problem, but the seal needs to be checked).
 
R

raymond02302

The ultrasound probe generally consist of different materials, plastic cover/housing and ultrasound transmitting/receiving window. They are cementeded by chemical reagent such as epoxy and other kinds of reagent. The gap filled with such reagent makes a cemented joint. This is why clause 8.9.3.3 is applicable. The thermal cycling test is to evaluate reliability of this cemented joint.
 
Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
B AS9100 Certification Audit Accounting for Additional Customer QMS Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 15
S Additional registration requirements for different EU-states on top of standard CE-registration? EU Medical Device Regulations 3
C Performing Additional Inspections IATF 16949 - Automotive Quality Systems Standard 10
V Certified Auditor - Need of additional certification specific to industry ( GMPs) ASQ vs ECA vs others Professional Certifications and Degrees 1
L Impact on 510(k) to use an additional manufacture US Food and Drug Administration (FDA) 0
I Additional National requirements for placing MD in the EU countries EU Medical Device Regulations 7
B GM GP-12 Exit Criteria - Additional Safe Launch Inspection - (Also see GM1927-28 ) Customer and Company Specific Requirements 2
A Medical device Reporting : Good Faith Effort for Additional information Other US Medical Device Regulations 2
J CE Certificate and additional processing EU Medical Device Regulations 4
dgrainger Informational Additional Guidance Regarding the Vigilance System as outlined in MEDDEV 2.12-1 rev. 8 EU Medical Device Regulations 1
V Additional Controls for processing the raw data on different computers (processing system) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 5
A Selling CE marked self-test in Malta - any additional requirements? EU Medical Device Regulations 0
K Letter to file vs Special 510 for additional accessory in orthopaedic kit Other US Medical Device Regulations 1
W FAIR to AS9102 additional documentation Document Control Systems, Procedures, Forms and Templates 5
C Additional Training for CCT (Certified Calibration Technician) Recertification Training - Internal, External, Online and Distance Learning 5
P Adding an additional distributor to a current Mexico Medical Device Registration Other Medical Device Regulations World-Wide 0
Q Additional KPI to Top Management beside Quality Objectives ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
Q Risk Management - Additional Process in ISO 9001:2015? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Changes in MDSAP regarding to CMDCAS - Are there any new additional requirements? Canada Medical Device Regulations 1
L How document an additional software of my main device software?? IEC 62304 - Medical Device Software Life Cycle Processes 3
B Additional requirements for selling Class I Medical Device in Poland EU Medical Device Regulations 3
A Special Characteristics -> Any specific additional requirements to be met? IATF 16949 - Automotive Quality Systems Standard 7
J Example of additional Audits of Remote Supporting Functions IATF 16949 - Automotive Quality Systems Standard 2
B Airbus & Boeing Standard and Additional Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
Z Getting Charged for AS9104/1 Additional Audit Time General Auditing Discussions 6
C AS 9102 Form 0 "Additional Information for High Risk Characteristics" AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
M Additional Procedures required for QA Department Document Control Systems, Procedures, Forms and Templates 10
P Does a 'recommended' additional Audit need to be conducted? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
A Additional Control Measures for Outsourced Processes under ISO 9001:2008 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
P ISO13485 Clause 7.2.1 (d) Additional Requirements - Need advise ISO 13485:2016 - Medical Device Quality Management Systems 2
G MDLA - Extensions allowed for 60 day Request for Additional Information? Canada Medical Device Regulations 4
G Additional Intended Use Predicate 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
T Shipping Additional Product ("Bundle") with a Class 2 Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
B CMDCAS versus ISO 13485 - What are the additional requirements Canada Medical Device Regulations 21
A TS 16949 5.6.2.1 - Input to Management Review - Additional Management Review Meetings and related Processes 7
B Additional Handset added to System - Additional 510(k) Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Adding an additional site to our current TS 16949 registration scope IATF 16949 - Automotive Quality Systems Standard 5
M Refurbishing a Medical Device - Any additional information needed 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
P CMDCAS Checklist with the additional requirements for Health Canada Canada Medical Device Regulations 4
B Additional site to be included in TS Cert. (Common owners & some functions) IATF 16949 - Automotive Quality Systems Standard 5
J Restructuring and Additional Responsibilities - Need New Quality Title Quality Manager and Management Related Issues 15
C Allowance for Slip Fit - Additional Allowance applied to GO NO GO Plug Gage General Measurement Device and Calibration Topics 26
Govind Additional discount on certification fee Professional Certifications and Degrees 4
S EC Official Journal Harmonized Standards - Additional requirements? ISO 13485:2016 - Medical Device Quality Management Systems 3
S Number of copies for an FDA additional information request? ISO 13485:2016 - Medical Device Quality Management Systems 3
J RoHS Compliance - Steel - Customer wants additional samples tested RoHS, REACH, ELV, IMDS and Restricted Substances 32
G Sample Report to justify additional Document Control workforce Document Control Systems, Procedures, Forms and Templates 3
W Additional Regulatory Requirements for Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 4
K Lead Internal Auditor Training - Does the lead auditor require additional training? Internal Auditing 6
H Additional sources for ASQ CQT exam sample questions Professional Certifications and Degrees 6

Similar threads

Top Bottom