Adjustment of Calibration Intervals (Shortening or Lengthen)

#1
Thank you all in advance for the help.

Basically, I am trying to change the culture toward the calibration standards and methods at my work - we have a lot of local procedures in place for processes and responsibilities for each engineer.

Currently there is an instruction in place that states if an out of Tolerance is found during a calibration effort (this goes for all equipment, from Oscilloscopes, to calipers, to flow modules.. etc) then the asset will have the interval cut in half; to lengthen the interval back to its original interval, then said asset must have 3 calibration cycles without any out of Tolerance finding.

So now we have assets with a wide range of intervals, nothing is consistent, and we are wasting time, money, downtime, and everything else under the sun.

We had an asset with a 3 year cycle, an out of Tolerance was found, but it left in tolerance, now it’s at a 18 month cycle, then another out of tolerance was found and now 9 months - we are currently at this asset being at. 1.5 month cycle due to so many out of Tolerance findings.

I am working on a proposal to follow Establishment and Adjustment of Calibration Intervals RP1, that way we can save money, have more time with equipment etc. But each time an asset has an out of tolerance , we provide an impact report to cover the time between the calibration where everything was in tolerance, until the out of Tolerance finding.

This all is weird to me, as long as the asset was re adjusted back into tolerance, we should not be consistently reducing intervals. I want to apply a A3 model to create a better system. All the assets that we reduce intervals on always come back “ as left in tolerance “

I feel like all the frequent calibrations put extra stress on the equipment, because now it’s subject to whatever happens when it leaves our facility.

Can anyone help me add to my point that this practice is not value added, perhaps more specific information to assist with my current issue? Thank you!
 
Elsmar Forum Sponsor

GunLake

Involved In Discussions
#2
I can't help you prove this does not add value because i think it does, in a way I do the same except i rarely ever go below 6 months. At that point i will replace it or find out why it's going out of tolerance. Reason for this is, You have a gauge that is out of tolerance at 3 years. So you go to 1.5 years and its out of tolerance again, So 1 year. Then 6 months, Then you switch it back to 3 years because you're sick of checking it so often, Well you know for sure 2 years of that gauge being used is out of tolerance. Normally if i find an out of tolerance and i decide to change it from lets say 1 year to 6 months, In 6 months i will change it back to 1 year if i find it within tolerance.

It's a bit easier for me, Since we calibrate almost everything in house. About the only thing i would change IMO is instead of cutting it in half do .5. So 6 months, 1 year, 1.5 years, 2 years, etc. To me it makes it easier and more consistent. and the 3 calibration cycles, I would just do it at 1 cycle.

I'm sure other people have different opinions but this is how I've done it for the last 7 years.
 
Last edited:

John Predmore

Trusted Information Resource
#3
I can think of three failure modes when a device goes out of tolerance - wear (or similar phenomena which cause drift in one direction over time), wear-out or other deterioration (an increase in uncertainty which may be labeled noise, which appears as random bidirectional fluctuation), or a bump (or similar incident which causes a sudden change). Setting a proper calibration interval attempts to monitor and correct for the occurrence of wear-like phenomena. Periodic re-calibration can detect wear-out as it progresses (if somebody looks for it), and trend charts may predict when deterioration will exceed allowable limits, then at some point the device must be refurbished or replaced. A bump is the hardest failure mode to predict. A bump could happen one year after a device is calibrated or one hour after. Reducing the calibration interval is no assurance against a bump. Shortening the interval will lead to devices often found in-tolerance, which is what you reported.

Based on my theory, I have an idea what may be causing your problem, but no specifics. With this theory, you might investigate and find evidence, you might identify what is occurring to a particular device, you might devise a solution. Or, I might be wrong. What you do with this theory is up to you. Good luck.
 

Mike S.

Happy to be Alive
Trusted Information Resource
#4
Something to think about....

When possible and practical I like to have a "check standard" for pre-use verification of important measuring equipment. As John said, a device can be calibrated today and go out of calibration tomorrow, and you may use it who knows how long until you find that out. Now you have a potential mess -- determining the impact of the OOT condition.
 
#5
The process does add value because it is more costly to recall products. Reducing the interval when out of tolerance condition is found is the proper method. If you are down to a 3-month interval from a 3-year interval you should be investigating what is causing the out of tolerance condition.
 

Tagin

Trusted Information Resource
#6
The process does add value because it is more costly to recall products. Reducing the interval when out of tolerance condition is found is the proper method. If you are down to a 3-month interval from a 3-year interval you should be investigating what is causing the out of tolerance condition.
^^ This.

The existing policy is at best a passive risk mitigation of symptoms. In a proper risk mitigation, you would do some kind of investigation for root cause (e.g., your A3) and thus mitigate the cause.
 

dwperron

Trusted Information Resource
#7
"Can anyone help me add to my point that this practice is not value added, perhaps more specific information to assist with my current issue?"

Not really, because you are wrong.

Calibration interval adjustment is designed to reduce risk in making an out of tolerance measurement.
Out of tolerance tools lead to risk of product recalls, that is where you really lose money and time.
I think your current algorithm is too aggressive at reducing intervals and too slow in raising them, but that is based on my experience, not with knowledge of your situation.

If you are adjusting instruments and their interval is shortening that is an indication of a problem with the equipment, not the calibration program. Just because your instruments were adjusted back into tolerance does not guarantee that they will now hold that accuracy, which makes your thought of not reducing intervals on these make no sense.
If you have an instrument that is on a short interval "due to so many out of Tolerance findings" then you should not be using the instrument! It is not providing reliable results, and puts your process at risk.

Not changing the intervals probably will mean that you will continue to have out of tolerances. That leads to another "impact report to cover the time between the calibration where everything was in tolerance, until the out of Tolerance finding." How much time is spent on these exercises? Keeping your instruments in tolerance will reduce the impact reports.

My experience in the calibration world shows that most calibrated instruments hold their accuracy. I normally see a calibration failure rate of about 5%. If you have so many out of tolerance events that they have become burdensome to your organization you probably have an equipment / use issue, not a calibration issue.
 
Thread starter Similar threads Forum Replies Date
Jerry Eldred Adjustment of Calibration Intervals (Calibration Frequency) Calibration Frequency (Interval) 13
T Calibration Certificate Readings After Adjustment General Measurement Device and Calibration Topics 5
WEAVER Calibration of Measuring Instrument by Adjustment of its Scales IATF 16949 - Automotive Quality Systems Standard 2
R Definition Safeguarded from Adjustment - Definition - Calibration - 7.6 d) Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 9
L Difference between Calibration, Verification and Adjust (Adjustment) General Measurement Device and Calibration Topics 6
S Gas Spectrometers - Calibration vs. Zero Adjustment General Measurement Device and Calibration Topics 4
J In-house calibration - Resource for calibration, adjustment of calipers, indicators General Measurement Device and Calibration Topics 2
A Calibration Interval (Frequency) Adjustment using Hilliard type 2 algorithm Calibration Frequency (Interval) 11
Marc Customer Audit of Guide 25 Calibration - Full pre and post adjustment data? General Measurement Device and Calibration Topics 1
A Depth micrometer rod adjustment? General Measurement Device and Calibration Topics 8
K Proto 6106 and 6106A Adjustment procedures General Measurement Device and Calibration Topics 0
A Product Impact When Thread Rings Require Adjustment General Measurement Device and Calibration Topics 9
K Vision System Camera Adjustment - Change Verification? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
B In-Process Sampling Plan for a Secondary Package Adjustment Process Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
R Temperature Adjustment when ME EQUIPMENT is rated for used at Higher Altitudes IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
G Assessing Process Capability on Variation (Hardware Adjustment Mean Shift) Capability, Accuracy and Stability - Processes, Machines, etc. 4
R FLUKE 1653B Adjustment - Instructions or Service Manual wanted General Measurement Device and Calibration Topics 4
M Minitab Pareto Chart Text Adjustment Using Minitab Software 10
D TS 16949 - 8.1.2 - Over adjustment should be understood throughout the organization IATF 16949 - Automotive Quality Systems Standard 11
F Definition Process Capacity and Over Adjustment - Definitions Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 12
D Pre-Control Chart to guide in-process inspection / adjustment decisions Inspection, Prints (Drawings), Testing, Sampling and Related Topics 43
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 0
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 5
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3

Similar threads

Top Bottom