Aging rig in medical device industry - To qualify or not

Fylgia

Registered
Hi,
I need some help in my thinking and to do some bench marking.

I'm working with medical device and in R&D we do component verification.
Part of component verification is to do stress tests / aging tests where the components are exposed to cold/heat/chemicals/pressure etc using a purpose built test rig. The rig is built upon calibrated sensors and a control software.

If the test rig would gather any measurements I gladly classify the rig as GxP and thus be subject to qualification (validation) according to our procedures.
However, the rig does not take any measurements or gather any results. It purpose is only to expose the components during a predetermined time.


  1. If you have similar equipment - do you qualify (IQ/OQ) or anything else?
  2. What did you base your decision on for how to treat the rig?
Thank you!
 

Ronen E

Problem Solver
Moderator
How can you be sure that the rig did what you intended it to do? I suppose no one sits there to supervise it while it works.
 

JJ_FDA

Involved In Discussions
I'm with Ronen. You need to qualify it - demonstrate that it does what you think it does.
 

yodon

Leader
Super Moderator
I'll "third" Ronen's response and agree qualification is needed. You're making quality decisions based on the use of the equipment so you need to have confidence that the equipment is doing what's intended.

Further, since it has control software, the software itself should be validated!
 

Fylgia

Registered
Thank you all!

I actually think I can draw a parallel to the use of a climate chamber, which I would always qualify.
The chamber itself do not measure anything on the component, but I want to be sure it works as expected.

(Comment - the people voting for non-qualification will claim that it is not GxP and it can be controlled by calibrating sensors before and after test run to reassure it has performed as expected)

Cheers!
 

JJ_FDA

Involved In Discussions
the people voting for non-qualification will claim that it is not GxP and it can be controlled by calibrating sensors before and after test run to reassure it has performed as expected

But does that not qualify :bonk: as a qualification test?
 
Thread starter Similar threads Forum Replies Date
Ed Panek USA Aging Population Coffee Break and Water Cooler Discussions 5
William55401 21 CFR 11 Password Aging - What does "periodically" revised mean in your org? Qualification and Validation (including 21 CFR Part 11) 3
S Sterile Disposable Cardiac Guidewire 3yr Real Time Sterility Failure When Confirming Accelerated Aging / Labeling US Medical Device Regulations 3
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
E When to perform accelerated aging studies and transport test(OQ or PQ) Qualification and Validation (including 21 CFR Part 11) 4
R Accelerated Aging - Creating test samples - Implantable medical device Question Other Medical Device Related Standards 4
shrutisancheti Predicate device accelerated aging - 1 year of shelf life is already over Other Medical Device Related Standards 1
A Accelerated aging and real time study EU Medical Device Regulations 2
M Real-time-aging Validation - ISO / ASTM Standard - Class 111 Orthopedic Implant Other Medical Device and Orthopedic Related Topics 1
Sam Lazzara Time Zero testing in med device aging studies Qualification and Validation (including 21 CFR Part 11) 3
shimonv Storage Conditions of 5-40 Celsius - Accelerated and real time aging - Dilemma EU Medical Device Regulations 3
W Shelf life and accelerated aging ISO 13485:2016 - Medical Device Quality Management Systems 2
N Accelerated Aging Testing Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
U Accelerated Aging for Shelf Life - Beginning Point Difference CE Marking (Conformité Européene) / CB Scheme 2
N Accelerated Aging Test for Disposable Protective Garment for Incontinence Other US Medical Device Regulations 1
S Medical Device Aging & Safety Questions by FDA Other US Medical Device Regulations 6
B Test Procedure or Standard for Accelerated Aging for AIMD Various Other Specifications, Standards, and related Requirements 2
E Aging/Shelf-Life Testing Responsibility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
P Regression Analysis for Aging Study in Minitab Statistical Analysis Tools, Techniques and SPC 10
Marc Aging in Mice Reversed - Immortality, anyone? World News 2
S Audit Summary Report - Nonconformance Aging Analysis Nonconformance and Corrective Action 10
A Medical Device (Gloves) Stability (Aging) testing and Shelf Life Other Medical Device and Orthopedic Related Topics 15
A Format of Stent Stability Testing (Accelerated and Long Term Aging) Other Medical Device and Orthopedic Related Topics 8
C Thermal Accelerated Aging Test(s) for Epoxy Reliability Analysis - Predictions, Testing and Standards 6
P Can I submit a 510(k) without aging data? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
T Aging Buckets for Metrics - Redefining a full set of metrics Six Sigma 3
Stijloor Aging Gracefully Funny Stuff - Jokes and Humour 1
S Validation of Sterilization Expiry Date - Accelerated Aging, etc ISO 13485:2016 - Medical Device Quality Management Systems 6
B Maintenance FMEA's - Aging Plant Equipment FMEA and Control Plans 6
J Applicable Great Britain regulations for a custom-built test rig or evaluation kit used for research purposes only. Design and Development of Products and Processes 0
E Uncertainty of Mean - Test rig that measures water temperatures Measurement Uncertainty (MU) 3
S RIG testing and BCR list - Heard of these acronyms? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
B MSA Study on Test Rig Constraints - Destructive, Long Duration Tests Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Ajit Basrur Random Insult Generator (RIG) Coffee Break and Water Cooler Discussions 2
N Regulations for Medical Device in Jamaica Other US Medical Device Regulations 0
M Classification of device according to China medical device regulations China Medical Device Regulations 4
M Looking for kindly help... If there is any prohibits from contract manufacturing in US? About SR medical devices US Medical Device Regulations 0
adztesla Design change and Process change Class 3 medical device Medical Device and FDA Regulations and Standards News 2
K New set of medical devices - samples EU Medical Device Regulations 3
M Importing a custom component for our medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
FRA 2 FDA Reusable Vs Single Use Medical Devices in EU Other Medical Device and Orthopedic Related Topics 1
adztesla Adding new supplier/ Medical device design change ISO 13485:2016 - Medical Device Quality Management Systems 2
adztesla Medical device Drawing related change justification Other Medical Device Related Standards 0
O Medical Device With 2 Power Cords IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
Stoic Which medical device companies are using Recombinant Factor C (rFC) instead of Limulus Amebocyte Lysate (LAL) for endotoxin testing? Sustainability, Green Initiatives and Ecology 0
I Commercializing a non medical device software Medical Information Technology, Medical Software and Health Informatics 10
I Non-medical device software Preventive Action and Continuous Improvement 0
C California Prop 65 for medical device labels US Medical Device Regulations 3
adztesla GDP rules- redline drawings in medical device Medical Device and FDA Regulations and Standards News 2
R Instructions for Use supplied with Medical Device Components Medical Device and FDA Regulations and Standards News 4

Similar threads

Top Bottom