Alternatives to Metal Foil for touch current measurement of enclosures made of insulating material.

Kenny03

Registered
Hello,

first time posting here and excited to be a part of this community, so we've got two parts (a power adapter and an applied part) in our system with enclosures made of insulating material. So, as part of production line safety tests, we wrap these parts in aluminum foil to measure touch current for the power adapter and patient leakage current for the applied part.

So I got a request from the production team asking "Is there a better way to do this?". We'd prefer not to use aluminum foil. is it possible to simplify this to use a metal plate or box? Would appreciate any tips and recommendations here.
 

Peter Selvey

Leader
Super Moderator
As part of design tests (type tests) it's good to do the foil method, but not in production. Production tests should be tailored to suit the design. For example, the secondary circuit of the power adaptor should pass the touch current test for operator. If so, foil around the adaptor is a waste of time, since it will obviously comply as well being one step removed. So you could do the test direct to the secondary rather than to the foil, which is much easier to implement and far more reliable, and should yield consistent results that can be analysed (e.g. if leakage changes >10%, you can tell something is changed and ask adaptor manufacturer what is going on). Or the adaptor might be fully certified with it's own production tests monitored by a third party, and this includes secondary touch current, if so you can skip this in your own production line tests.

Patient leakage can also be simplified based on the design, and it can even be eliminated if it is blindingly obvious by design that it can never fail.

There are always special cases for example a medical ultrasound probe might rely on a sealed joint to protect against internal voltages/currents reaching the patient. In this case, it's reasonable to do patient leakage current tests with the probe immersed in 0.9% saline. This is better than using foil since it's hard to ensure the foil always contacts all the critical points.
 

Kenny03

Registered
Hi peter,

thanks for the reply.

Ah yes, i see your rationale, as touch current limits are intended to protect against the fault case where the secondary side hot line is shorted to chassis. And so if we directly connect to the secondary side hot line for our touch current measurement and we pass, then we certainly will pass with the foil wrapped adapter touch current. Have i got that right?

As for the applied part, yeah i saw that alternative method mentioned in the standard but that really only works for seal/waterproof parts and unfortunately, ours is not. :( It seems like we're stuck with the foil wrapped option in this case.
 
Thread starter Similar threads Forum Replies Date
S Question: Best alternatives to audits for ensuring compliance Quality Tools, Improvement and Analysis 5
G Alternatives to time apps (mobile applications)? Coffee Break and Water Cooler Discussions 3
C Alternatives to ISO-9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
D Alternatives to Customer Feedback Survey in ISO 9001 Quality Manager and Management Related Issues 4
J0anne Alternatives to 'Latex Free' Labelling Other Medical Device Related Standards 9
L New Website from the FDA concerning UDI Exceptions, Alternatives and Time Extensions Other US Medical Device Regulations 0
B Alternatives to Google Reader - is Feedly a good option? After Work and Weekend Discussion Topics 2
A Thailand - Certificate of Free Sale Alternatives Other Medical Device Regulations World-Wide 18
G ISO 9001 Clause 7.6 Tamper Seals - Alternatives to Tamper-Proof Seals ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
P Web based file sharing i.e. Dropbox for Doc Control - Comments/Alternatives? Document Control Systems, Procedures, Forms and Templates 6
S Medical Device Class IIa Nonsterile ISO 13485 Alternatives ISO 13485:2016 - Medical Device Quality Management Systems 3
K Alternatives to GR&R (Gage R&R) - Only 4 sample parts and one operator Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 22
B "Controlled" stamp for Controlled Documents - Any alternatives? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 17
Ron Rompen Alternatives to welding two pieces of aluminum extrusion together at a miter joint Manufacturing and Related Processes 2
S Power Outage During Sample Conditioning - What Alternatives are Available? General Measurement Device and Calibration Topics 1
8 Gage Track Calibration Software Alternatives Quality Assurance and Compliance Software Tools and Solutions 8
R Looking for alternatives to ISO 9001 accreditation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 41
A Shuttle Foam Question - Alternatives? Coffee Break and Water Cooler Discussions 13
S Controlling external processes - Alternatives ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
P Gage R&R and non-normal distribution - Alternatives to a standard gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
R Alternatives to the Title of 'QUALITY' Career and Occupation Discussions 51
E Insulation class II device with metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
C Mistake Proofing Winding Direction Stamped Metal Parts Manufacturing and Related Processes 8
O New metal detector, ccp -help! Food Safety - ISO 22000, HACCP (21 CFR 120) 2
A Critical Characteristics - Sheet metal stamping APQP and PPAP 9
T Accessible metal parts connected to secondary ground IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Metal parts bonding/grounding on Class II, internally powered with functional earth? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J ISO 13485 for Metal Finishing Medical Device and FDA Regulations and Standards News 5
A Metal Enclosure Connected to PCB Ground for ESD Reasons IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Lighting Standard for Metal Processing Inspection Manufacturing and Related Processes 4
K Quality Control/Assurance in the Metal Industry Quality Tools, Improvement and Analysis 10
B SOP for CNC turret punching machine for sheet metal process Manufacturing and Related Processes 0
M 'Active' device? Software/laptop with attached camera 'looking' at passive metal probe EU Medical Device Regulations 3
E IEC 60601-1 - Unearthed Medical Device Metal Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Raw material certificates - CC - Safety products - Sheet metal stamping IATF 16949 - Automotive Quality Systems Standard 1
A Literature review/HACCP validation of metal detection Food Safety - ISO 22000, HACCP (21 CFR 120) 0
M Informational US FDA paper – Biological Responses to Metal Implants Medical Device and FDA Regulations and Standards News 0
M Informational US FDA Final guidance – Metal Expandable Biliary Stents – Premarket Notification (510(k)) Submissions Medical Device and FDA Regulations and Standards News 0
Q IATF rule for single site - Ingots from scrap metal recycling company IATF 16949 - Automotive Quality Systems Standard 0
C AMS2750E - Noble metal as a Load sensor AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
V Thinning percentage in sheet metal forming per SES (Suzuki Engineering Standard) Manufacturing and Related Processes 1
F AS9100 D - Verification of purchased product - Sheet metal and aluminum extrusion stock AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
F MOPP in battery powered device with metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
D Error proofing ideas sheet metal punches Manufacturing and Related Processes 8
D IEC 60601 Defibrillation protection test - Metal foil IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Ron Rompen Crack testing of powdered metal parts General Measurement Device and Calibration Topics 4
A Test method for presence or absence of hexavalent chromium on a metal coating RoHS, REACH, ELV, IMDS and Restricted Substances 4
R Statutory/Regulatory for NACE Code 25.9 (Manufacture of fabricated metal products) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Industry Standards - Sheet Metal a Machine Shop Manufacturing and Related Processes 1
J Measuring an Untrue Radius - Metal Stamped and Formed Parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8

Similar threads

Top Bottom