AMS 2750 F Instrument Calibration Results and Recording

#1
AMS 2750F Section 3.2.5.2 g states that the "Error" shall be documented on the calibration report. We have always had the Correction Factor reported and never had it questioned by Nadcap until this year. Before having my calibration source complete redo the report to include the Error, I wanted to see if there were other options. Has anyone run into this in AMS 2750 journey? We only use the CF so I'm trying to eliminate confusion for the operators by adding another value on the cert. Is a comment on the cert saying something to the affect "Thermocouple readings provided in "correction factor" format. The Error is the opposite value of the Correction Factor."
 
Elsmar Forum Sponsor

darkopsghost

Involved In Discussions
#2
It's been a long time since I've been audited by NADCAP for Composit Bonding and had to do Uniformity Surveys (AMS2750D), but I listed everything on a cover page including offsets. Never had issue with a single Survey, but I do know a few years ago there was a change in the mindset and I think it was all driven from temperatures far exceeding what we used. I do know there is a lot of confusion as to what is actually required and some if not all of that comes from those selling a service. A sister company ran into that after I left and they were lead down a primrose path only to find out none of what they were being told was even close to the requirements of 2750. Sometimes, the same holds true with NADCAP Auditors.
 

Randall Beck

Involved In Discussions
#3
By the book answer is the Error needs to be reported according to rev F. However, as long as it is CLEARLY defined as the correction factor or the Uncorrected error I cannot recall as single auditor over 14 years that would write this as long as it is clear the error is the inverse of the reported correction factor on your calibration report as you stated above.

A much larger issue for me was trying to get my calibration companies to properly document the modification and correction offsets on the report and account for the modification offsets. I got overly frustrated with 3 calibration companies that could not or did not understand the new requirements, I bought a secondary standard calibrator and now perform all internal calibrations myself. I made my own calibration report to include both error and correction factors for my field test instruments for our convenience. Problems solved.....
 
Thread starter Similar threads Forum Replies Date
B AMS 2750 Furnace Instrument Calibration Operating Range AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
K Pyrometry Question - Furnace and Oven Instrument Calibration IAW AMS 2750 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
Q AMS 2750 E or F Continuous Furnace TUS Data Collection AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
J AMS 2750 E Clause3.5.17.13 TUS Pass/Fail Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
A Salt Bath Probe Style SAT requirement Waiver per AMS 2750 Rev E Section 3.4.7 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
J Could Fluke thermometer be used in SAT test in accordance to AMS 2750 ? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
B Urgent Help With Calibration Issue AMS 2750 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
B Inhouse Calibration Plus TUS in accordance with AMS 2750 rev E? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 31
S AMS 2750 - Instrumentation Calibration Requirements Manufacturing and Related Processes 12
L Calibration Uniformity Procedure for AMS 2750 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
S AMS 2750 rev E - Changes Detailed - Work In Progress Manufacturing and Related Processes 13
S SAE AMS 2750 rev E - Changes Matrix? Manufacturing and Related Processes 13
C Nadcap AMS 2750 Data Storage - I'm looking for pre-packaged software AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
M How are ISO-17025 and AMS-2750 related? ISO 17025 related Discussions 2
C Furnace Classes in relation to AMS 2750 to determine Calibration Interval Calibration Frequency (Interval) 2
G AMS 2750 - Tell me what troubles, specifically, you have had with AMS 2750 Various Other Specifications, Standards, and related Requirements 4
Q Need New AMS 2750 Compliant Temperature Data Logger Manufacturing and Related Processes 8
S AMS 2769 - weekly frequency test AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
B Determining SAT Offsets vs TUS Offsets per SAE AMS 2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 0
R DFARS Compliance - Purchasing Aluminum Alloy 7075 -T73511 per AMS-QQ-A-200/11 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
Q AMS 2750E Sensitivity Requirements Field Test Instrument AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
Q AMS 2750E System Accuracy Test (SAT) Frequency AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
D AMS 2750E - Digital Instrument Calibration Accuracy Requirements General Measurement Device and Calibration Topics 1
M SAE-QQ-A-250/11A has been superseded by SAE-AMS-4027N AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E AMS-H-6875 compared to MIL-H-6875 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
M SAE AMS 2750E Relocation Hot or Cold Recording Sensor Manufacturing and Related Processes 2
S AMS 2750E Heat Treat Furnace System Accuracy Test SAT Report question AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
Douglas E. Purdy Relation of Aluminum 6061-0 to AMS 4025 or AMS 4027 Various Other Specifications, Standards, and related Requirements 3
L Determination of AMS 2750D Furnace Classifications AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
N AMS 2750D Process Audit Check Sheet ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M What other standards meet AMS 6415? Various Other Specifications, Standards, and related Requirements 1
P Standards Subscription - Online web site that provides AMS, NAS, SAE etc, standards Various Other Specifications, Standards, and related Requirements 1
S SAE AMS 2175 Radiographic Report of Casting Other ISO and International Standards and European Regulations 8
G S30100 steel bar spec - related specs, ASTM or AMS? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
J AMS 4218 Spec, Equivalent AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
G Understanding AMS 2750D compliant calibration report - items under 3.2.6.2 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
D SAE AMS 2750D - Pyrometry - Heat Treating procedure wanted Reliability Analysis - Predictions, Testing and Standards 12
A AMS Specification QQ-N-290 - What does the class 1, Type IV (QS) mean? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 6
B Ams 7875(pwa53-5) Various Other Specifications, Standards, and related Requirements 5
D CQI-9 vs. AMS 2750D - Are there discrepancies? Various Other Specifications, Standards, and related Requirements 1
D SAE AMS 2750D - Pyrometry - Heat Treating - Does anyone know of a good class? Training - Internal, External, Online and Distance Learning 5
D AMS 2750C - Guidance under AMS 2750C for when a furnace fails a uniformity survey? Various Other Specifications, Standards, and related Requirements 2
M SMED - 2750 Ton Injection Molding Machine Benchmark Benchmarking 3
D Comparing inline Conductivity instrument with offline test results General Measurement Device and Calibration Topics 0
R Clinical evaluation for instrument of a navigation system EU Medical Device Regulations 1
WEAVER Can you determine exact date an instrument went out of tolerance? General Measurement Device and Calibration Topics 10
R Suggest modern equipment/instrument for plastic bottle inspection? Quality Tools, Improvement and Analysis 1
H California Instrument AC Load - 222 Error Message Manufacturing and Related Processes 0
N 60601-2-2 Ed 6 Clause 201.8.8.3.102 - HF instrument leakage IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 4

Similar threads

Top Bottom