An example of risk analysis of class I MD
Tidge
Trusted Information Resource
#11
Presumably the item in question is not an ointment or cream, but some sort of device. Personally I'm having trouble imagining what sort of 'cosmetic device' suddenly now has a medical purpose. I jumped to decorative contact lenses only because I'm aware of some past FDA crackdowns in that area. Maybe a foot spa? There are likely bigger issues with the new marketing plan than trying to comply with 14971, but a rigorous application the standard would be a benefit to the manufacturer, even if they only see having a RMP as some sort of paperwork exercise.
The fact that a product has been on the market as a cosmetic is not necessarily relevant.
The sole point of relevance I see as being potentially useful, for the specific circumstance described here: If the device has already been on the market, presumably the manufacturer is aware of the harms that arise from the use of the device because they have been taking complaints and/or collecting data about their market. If they are trying to start an assessment of risks from scratch, they absolutely need to have some basic understanding of the harms, whether they intend to 'calculate' acceptability or not.
Elsmar Forum Sponsor
Watchcat
#12
Many cosmetic devices walk a fine line between cosmetic and medical devices. Manufacturers often try to stay on the cosmetic side of the line, because of the lower regulatory burden. They do this by choosing their words very carefully in their labeling and advertising.
It is not uncommon for these manufacturers to identify a medical market that they decide is lucrative enough to warrant the cost of the higher regulatory scrutiny. If so, all they need to do to convert the device from cosmetic to medical is change the claims.
Depending on regulatory definitions, it is also possible for a product to be a medical device in one jurisdiction and a cosmetic device in another, just like it may be one class of device in one jurisdiction and a different class in another. In that case, all you need to do to suddenly convert a product from a cosmetic device to a medical device, is to decide to market it in a jurisdiction where it is regulated as a medical device.
Ronen E
Problem Solver
Staff member
Moderator
#13
Welcome to the wonderful world of formulated medical devices (a sort of devices) - some of which are certainly creams and ointments, others are solutions, powders etc. To qualify as medical devices they need to function chiefly not through a metabolic, pharmacological or immunological mechanism (each of which has quite a precise definition) a chemical reaction in or on the body or (specifically) through being metabolised. For example, an ointment may serve as a physical barrier, a powder may serve as a packing agent etc. I suggest not jumping to conclusions whilst we weren't told what the product is and what it's intended for (before and after the "change").
Added in edit: I initially related to the EU medical devices definition (see struckthrough text), and later recalled that that discussion evolved in an FDA context so replaced it with the relevant wording from the FDA's definition. The argument is not really affected by that, though the FDA's definition is not as specific.
I think you presume a little too much.
The nature of any received complaints (and their occurrence/reporting rate) may well be related to the manufacturer's claims, the distribution channels etc., i.e. as what the product has been marketed, and for what. They may not have gotten many (or any) complaints related to the medical use because such use didn't occur yet (because no one has intended it and promoted it, yet).
Second (as already noted), they might have not been collecting data about "their market" (Q: what market - medical or cosmetic?), not because they're a dodgy company or a lazy/sloppy team, but simply because it's not required or expected for cosmetics at the same level that's expected for medical devices (in most jurisdictions). Proactive data collection costs money.
Exactly my (bigger) point: Understanding the harms (and their realisation mechanisms) is paramount for assessing and evaluating risks, which is both what the standard requires (in contrast with calculation) and what is practical in the vast majority of cases in which a new product is brought to the market (new use = new product, in medical devices terms).
Watchcat
#14
Color me skeptical, but the point may be moot, as I have yet to see a company rigorously apply anything it saw as some sort of paperwork exercise.
Ronen E
Problem Solver
Staff member
Moderator
#17
Here you go:
http://demo.onyx-healthcare.com/product/112/certification/RM report_1511051.pdf
Would be interested to hear Tidge's and Ronen E's high-level assessment (gut reaction?) as to whether this is likely to serve as a useful template.
It's too long for any kind of assessment, but my gut feeling is that it's worth investing the time going through if one has no other template (and likes working by someone else's). I have my own template, which is about the same length, and is actually just a walk through ISO 14971 (taken quite literally).
Watchcat
#18
I'm not big on using other people's documents as templates, but I think looking at an example or two if you've never prepared that type of document before can be helpful, especially in situation where someone is likely to be reviewing your document, and that someone has seen a lot of them, and therefore may have certain expectations about format and content.
For that purpose, I'd be inclined to want to see more than one example. I found this one easily by googling "risk management report" content OR format OR outline. I expect there are more to be found.
Ronen E
Problem Solver
Staff member
Moderator
#19
Not so much from an auditor or official reviewer standpoint, but I have seen "a lot of these", and I can tell you that this one seems like one of the better ones. Most of the ones I've seen (including many that have passed regulatory scrutiny with flying colours) were quite poorly implemented and reported, or in plain French - c**p. If I was to judge by what the average auditor is likely to expect / look for (based on my not-unlimited experience), they don't really look for something like this (fine) example you provided. If I'm, on the other hand, to judge by my own professional compass, this example is quite good (caveat: haven't thoroughly reviewed it yet).
Not sure it matters to Google (only Spinoza's God knows what's in its algorithm), but to me a "risk management report" is something more limited, and that example is more along the lines of a completed "risk management file".
Tidge
#20
It definitely appears that in this example the first 38 pages are meant to cover multiple elements of a risk management file, and not everything is something that I would include within a risk management plan/report. I generally prefer to keep multiple documents, where the individual documents 'belong to' the most appropriate group of people. My bias: my experiences have been with manufacturers of diverse product lines with individual products within product lines, so maintaining separate policy documents outside of individual risk files is what I am most familiar with.
The major area of concern I have in the approach of this example is (scant? ambiguous? quasi-boilerplate?) in the area of post-production monitoring and how it feeds back into the RM process. If this specific example is intended to cover all of 14971, I don't see how it is satisfying the clauses of section 10 (of 14971:2019, section 9 of 14971:2012).
Here is a more targeted issue, but I believe it is a comment somewhat general to implementation of 14971: Specific to this example, I feel that there may be some ambiguity about who came up with the severity of harm ratings... my strict reading of the file implies that it was the Engineering R&D associate. Engineering associates are typically not qualified to make a determination as what types of injury require professional medical intervention. (Feel free to view this as a potential audit finding against 13485 - Human Resources). This is the sort of assessment that I'd recommend be 'owned' by a different group of people. As with my earlier comment about dividing up the information, including the actual rating systems within a RM Plan/Report (as opposed to a more general policy document that is referenced by the RMP/RMR) implies that different RM Plans treat severity (of similar harms) differently (possibly depending on the device/product line).
Thread starter Similar threads Forum Replies Date
J Does anyone have an example of Risk-Benefit Analysis per ISO 14971? Other ISO and International Standards and European Regulations 2
A An Example of a Risk Analysis Report for a Class II Medical Device (Oxygen Mask) ISO 14971 - Medical Device Risk Management 29
B Section 7.2.2.2 Organization Manufacturing Feasibility - Risk Analysis example? IATF 16949 - Automotive Quality Systems Standard 1
O Example Risk Analysis Procedure or Flow Chart - TS 16949 Document Control Systems, Procedures, Forms and Templates 2
Q Example of the Risk Template Document Control Systems, Procedures, Forms and Templates 1
M Risk managment report of Surgical Mask Example ISO 14971 - Medical Device Risk Management 14
A 5 x 5 Risk Matrix - Looking for a good example Manufacturing and Related Processes 2
D Validation of existing equipment - Risk based approach example ISO 13485:2016 - Medical Device Quality Management Systems 3
M Example ISO 14971 policy and risk criteria ISO 14971 - Medical Device Risk Management 0
O Example of Turtle Diagram with Risk/Opportunities General Auditing Discussions 1
B Process Risk Assessment Example for a Manufacturing Company Quality Tools, Improvement and Analysis 2
P Example Risk Assessment for CAPA's Document Control Systems, Procedures, Forms and Templates 5
M Example of a Risk Assessement File for Medical Device Other ISO and International Standards and European Regulations 5
S Aspect vs. Impact and Hazard vs. Risk - Short/clear explanation & example Miscellaneous Environmental Standards and EMS Related Discussions 11
D AS9100C Risk Management Procedure Example AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 111
K Risk Management Policy example and advice wanted ISO 14971 - Medical Device Risk Management 3
O Supplier Risk Assessment template or example needed Supplier Quality Assurance and other Supplier Issues 5
D Risk Assessment - Problem Management Process - Seeking Model/Example Software Quality Assurance 1
J Example of a defined procedure for carrying out Material Review Board (MRB) Manufacturing and Related Processes 0
S IATF 16949 Internal Audit Example IATF 16949 - Automotive Quality Systems Standard 7
N Example for design and development planning,input,output,review,verification,validation and transfer Misc. Quality Assurance and Business Systems Related Topics 4
optomist1 Paynter Chart Example APQP and PPAP 1
M Data Protection and Privacy Policy - looking for a template/example EU Medical Device Regulations 1
A Verification protocol example Design and Development of Products and Processes 2
Q Example Process orientation to the process leadership, management, goals (tasks of the top management) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
D Test summary report example for design validation wanted - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 1
M Example of statement for procedure pack MDR Article 22? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
N Post Market Surveillance Package Example Wanted ISO 13485:2016 - Medical Device Quality Management Systems 1
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
D Redacted 510(k)s - Example wanted Document Control Systems, Procedures, Forms and Templates 17
F Hi friends, can anyone show me an example of a procedure for ISO 13485 6.4.1 Work Environment? ISO 13485:2016 - Medical Device Quality Management Systems 2
DuncanGibbons Looking for example aerospace part CAD files to be used for a case study Career and Occupation Discussions 2
Jimmy123 Example of a P-Diagram for Process FMEA - Uncontrollable noises FMEA and Control Plans 39
D Monitoring and measuring resources - Example of how section ISO 9001 7.1.5 would apply to a manufacture of software ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
A ISO/TS 22163 Item 5.2.4 Safety policy - Example wanted Other ISO and International Standards and European Regulations 2
F Internal Audit - Procedure example Internal Auditing 5
I Functional Safety Analysis/Report Example IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
H Man Hour Plan 145.A.30 example needed EASA and JAA Aviation Standards and Requirements 0
R Real Example of a currently marketed Procedure Pack CE Marking (Conformité Européene) / CB Scheme 1
D Example for Accessories of "Annex XVI non-medical devices" and how to classify it now with respect ot EU MDR Other Medical Device Regulations World-Wide 1
MichaelDRoach Oganization's Self Declaration of Conformance to ISO 28000 wording example wanted Supply Chain Security Management Systems 1
A Australia TGA Medical Device Sponsor Agreement - Example or Template Other Medical Device Regulations World-Wide 18
J ISO 13485 8.4 Analysis of Data - Procedure example ISO 13485:2016 - Medical Device Quality Management Systems 1
S Question about a basic additive variance/tolerance example Quality Tools, Improvement and Analysis 2
N ISO 9001: 2015 - Example standards policies, procedures, acknowledgements and checklists ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M AS9100d (section 7.1.3) example of compliance in machine shop environment. AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
supadrai How does the FDA count violations of, for example, misbranding? (Medical Devices) Other US Medical Device Regulations 1
E Labware cleaning validation protocol example Qualification and Validation (including 21 CFR Part 11) 1
D Laboratory Manual ISO/IEC 17025 Example wanted ISO 17025 related Discussions 2
P Template or example Excel list of ISO certificates issued to clients ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
Similar threads
Top Bottom