So I finally got curious and took a look. I found this document useless from a regulatory perspective, because, unless I missed something, it doesn't identify the harms associated with the risks.

Engineering associates are typically not qualified to make a determination as what types of injury require professional medical intervention.
Oh good, my second opportunity to rant about this today. IMO, anyone really qualified to make this determination would tell you that trying to quantify the severity associated with a clinical harm makes no clinical sense whatsoever In this case, it's just plain weird, because you don't even know what harms are being rated. It's like, well, if some kind of harm occurred, then whatever kind of harm it might be, it would have a severity of 2.

(Feel free to view this as a potential audit finding against 13485 - Human Resources)
Don't tease me, Seymour! Have you ever actually known an ISO 13485 auditor to cite this as a finding?
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#22
Engineering associates are typically not qualified to make a determination as what types of injury require professional medical intervention.
As a boilerplate critique, I accept its validity (especially with the "typically" emergency exit included). But in terms of adding value (or of process effectiveness and process efficiency), and as a generalised approach, I think that it's overreaching. It all depends on the general risk realm the specific subject device is in, and the specific harms. Sometimes it's quite obvious and one doesn't need to be a MD or RN to tell. For borderline cases, consulting the right professionals is definitely warranted.
 

Tidge

Involved In Discussions
#23
For the example provided I have strong opinions where it could be improved, but I tried to restrict my comments to where I felt it was 'obviously' deficient. Watchcat's valid point about the way harms are (not) addressed is on the same spectrum as my complaint about who is capable of assessing the severity of the harms. My experience has been that a successful approach is to have an experienced medical professional take a comprehensive list of harms and then rank them according to a scale (typically 1 to 5) that allows their use in making risk assessments. If an organization has such a "Master Harms" list, it removes this element of subjectivity from individual risk management plans.

Not every harm from a general "Master Harms" list is going to be relevant for every device, but if the organization suddenly gets of complaint that includes a patient or user harm that wasn't originally considered in the risk file an organization with a general master harms list is more likely to be able to respond in a more mature (less reactive) manner than one that didn't even consider the possibility of that harm. For example, it might seem absurd to consider "choke hazards" for a corrective contact lens, but that doesn't mean a lens manufacturer could ignore a complaint that someone choked on one of their lenses.

People can quibble about the (possibly perceived as arbitrary) numerical value associated of such an assessment, but this is the only practical way I know of to address risks to apply a methodology that allows for non-serious harms to be less acceptable risks than those from more serious harms (S x O). An organization may have ranked "Finger cut that doesn't require stitches" as a S=2 and "Kidney necrosis" as an S=5, but if the users is cutting their finger every time they touch the device, this is almost certainly an unacceptable risk, independent of the risks associated with tissue necrosis.
 

Ronen E

Problem Solver
Staff member
Moderator
#24
My experience has been that a successful approach is to have an experienced medical professional take a comprehensive list of harms and then rank them according to a scale (typically 1 to 5) that allows their use in making risk assessments. If an organization has such a "Master Harms" list, it removes this element of subjectivity from individual risk management plans.
I don't disagree - my experience points in the same general direction. However, this falls (IMO) under the "if we lived in a perfect world... however..." advice category. Not every little design change process warrants engaging a (relevant) clinician, and one is many times not available within the org. When they aren't, it would usually cost not a few $$. the more complicated/risky the application is, the harder it will be to locate a relevant individual, and that individual's service is likely to cost more. Please don't forget that the clinician is taking a serious responsibility in doing so, especially if their output is "Masterised" and subsequently used in instances and contexts that the clinician can't predict. So aside from them jacking up the price (high risk warrants high return, that's Economics 101), they are also likely to jack up their severity assessments. An employee of the company (e.g. an engineering associate) is not personally exposed in that manner and thus likely to be less concerned.

BTW, the fact that an individual is a clinician and not an engineer doesn't mean that their judgement is not subjective. Less invested maybe (if they're external to the project), but not less subjective. The way I know to deal with subjective judgement is to obtain multiple opinions, but that makes the exercise even more costly (in some case it's justified).
this is the only practical way I know of to address risks to apply a methodology that allows for non-serious harms to be less acceptable risks than those from more serious harms (S x O).
I assume you meant "this is the only practical way I know of to address risks to apply a methodology that allows for non-serious harms to be more acceptable risks than those from more serious harms (S x O)."
People can quibble about the (possibly perceived as arbitrary) numerical value associated of such an assessment, but this is the only practical way I know of to address risks to apply a methodology that allows for non-serious harms to be less more acceptable risks than those from more serious harms (S x O).
I think that not many argue against risk ranking in this manner. I don't - as long as everyone is aware that the ranking is relative and subjective in most practical cases (where a lot of hard data is already available, that's great, but it's not so common in device D&D). I was arguing against getting overly wrapped up in "precise" calculations based in most cases on wishful thinking and a lot of coffee.
 

Tidge

Involved In Discussions
#25
I assume you meant "this is the only practical way I know of to address risks to apply a methodology that allows for non-serious harms to be more acceptable risks than those from more serious harms (S x O)."
I actually meant what I originally wrote, but perhaps I ought to have qualified the sentiment as "before the application of risk controls." It can be unacceptable to allow a near-constant (high occurrence) amount low-level harm (i.e. non-serious, low severity harms such as minor cuts, burns, etc.) just as it can be acceptable to have an extremely low occurrence rate of some very serious harms. This is of course, modulo discussions centered on "all risks shall be reduced as much as possible".
 

Ronen E

Problem Solver
Staff member
Moderator
#27
I actually meant what I originally wrote, but perhaps I ought to have qualified the sentiment as "before the application of risk controls." It can be unacceptable to allow a near-constant (high occurrence) amount low-level harm (i.e. non-serious, low severity harms such as minor cuts, burns, etc.) just as it can be acceptable to have an extremely low occurrence rate of some very serious harms. This is of course, modulo discussions centered on "all risks shall be reduced as much as possible".
Thanks for clarifying, I think I understand now.
I agree that harm severity should not alone drive the acceptability determination. I didn't notice the subtlety you referred to because you specifically wrote "S x O". Both severity and likelihood should be taken into account when considering a risk.
Again, I'm all in for rating and ranking, as long as we recognise our own limitations and don't delude ourselves that its perfectly rational / objective / watertight or the likes. There is a lot of potential value in the process, but it's mostly in taking the time and attention to ponder and to look for a better understanding, not in the mechanistic application of a mathematical algorithm.
 

Watchcat

Trusted Information Resource
#29
Good example. How would you rate the severity of the clinical harms that are potentially associated with a paper cut, prior to application of risk controls?
 
Last edited:

Ronen E

Problem Solver
Staff member
Moderator
#30
Good example. How would you rate the severity of the clinical harms that are potentially associated with a paper cut, prior to application of risk controls?
If you recall, I was referring to making a determination as what types of injury require professional medical intervention. Not the severity of the clinical harms in general.
 
Thread starter Similar threads Forum Replies Date
J Does anyone have an example of Risk-Benefit Analysis per ISO 14971? Other ISO and International Standards and European Regulations 2
A An Example of a Risk Analysis Report for a Class II Medical Device (Oxygen Mask) ISO 14971 - Medical Device Risk Management 29
B Section 7.2.2.2 Organization Manufacturing Feasibility - Risk Analysis example? IATF 16949 - Automotive Quality Systems Standard 1
O Example Risk Analysis Procedure or Flow Chart - TS 16949 Document Control Systems, Procedures, Forms and Templates 2
A 5 x 5 Risk Matrix - Looking for a good example Manufacturing and Related Processes 2
D Validation of existing equipment - Risk based approach example ISO 13485:2016 - Medical Device Quality Management Systems 3
M Example ISO 14971 policy and risk criteria ISO 14971 - Medical Device Risk Management 0
O Example of Turtle Diagram with Risk/Opportunities General Auditing Discussions 1
B Process Risk Assessment Example for a Manufacturing Company Quality Tools, Improvement and Analysis 2
P Example Risk Assessment for CAPA's Document Control Systems, Procedures, Forms and Templates 5
M Example of a Risk Assessement File for Medical Device Other ISO and International Standards and European Regulations 2
S Aspect vs. Impact and Hazard vs. Risk - Short/clear explanation & example Miscellaneous Environmental Standards and EMS Related Discussions 11
D AS9100C Risk Management Procedure Example AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 111
K Risk Management Policy example and advice wanted ISO 14971 - Medical Device Risk Management 3
O Supplier Risk Assessment template or example needed Supplier Quality Assurance and other Supplier Issues 5
D Risk Assessment - Problem Management Process - Seeking Model/Example Software Quality Assurance 1
M Example of statement for procedure pack MDR Article 22? Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
N Post Market Surveillance Package Example Wanted ISO 13485:2016 - Medical Device Quality Management Systems 1
D Calibration Process Flow Map Example Wanted General Measurement Device and Calibration Topics 3
D Redacted 510(k)s - Example wanted Document Control Systems, Procedures, Forms and Templates 14
F Hi friends, can anyone show me an example of a procedure for ISO 13485 6.4.1 Work Environment? ISO 13485:2016 - Medical Device Quality Management Systems 2
DuncanGibbons Looking for example aerospace part CAD files to be used for a case study Career and Occupation Discussions 2
Jimmy123 Example of a P-Diagram for Process FMEA - Uncontrollable noises FMEA and Control Plans 38
D Monitoring and measuring resources - Example of how section ISO 9001 7.1.5 would apply to a manufacture of software ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
A ISO/TS 22163 Item 5.2.4 Safety policy - Example wanted Other ISO and International Standards and European Regulations 2
F Internal Audit - Procedure example Internal Auditing 5
I Functional Safety Analysis/Report Example IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
H Man Hour Plan 145.A.30 example needed EASA and JAA Aviation Standards and Requirements 0
R Real Example of a currently marketed Procedure Pack CE Marking (Conformité Européene) / CB Scheme 1
D Example for Accessories of "Annex XVI non-medical devices" and how to classify it now with respect ot EU MDR Other Medical Device Regulations World-Wide 1
MichaelDRoach Oganization's Self Declaration of Conformance to ISO 28000 wording example wanted Supply Chain Security Management Systems 1
A Australia TGA Medical Device Sponsor Agreement - Example or Template Other Medical Device Regulations World-Wide 11
J ISO 13485 8.4 Analysis of Data - Procedure example ISO 13485:2016 - Medical Device Quality Management Systems 1
S Question about a basic additive variance/tolerance example Quality Tools, Improvement and Analysis 2
N ISO 9001: 2015 - Example standards policies, procedures, acknowledgements and checklists ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M AS9100d (section 7.1.3) example of compliance in machine shop environment. AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 3
supadrai How does the FDA count violations of, for example, misbranding? (Medical Devices) Other US Medical Device Regulations 1
E Labware cleaning validation protocol example Qualification and Validation (including 21 CFR Part 11) 1
D Laboratory Manual ISO/IEC 17025 Example wanted ISO 17025 related Discussions 2
P Template or example Excel list of ISO certificates issued to clients ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
T Example wanted - Template for internal audit IATF 16949 IATF 16949 - Automotive Quality Systems Standard 1
D System Level FMEA example wanted FMEA and Control Plans 2
E Example of 9001:2015 audit report with non-conformances ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
A IATF 16949 Cl. 8.4.1.2 Supplier Selection Process - assessment example IATF 16949 - Automotive Quality Systems Standard 11
D Where I can find an ISO 13485:2016 Audit Schedule example? ISO 13485:2016 - Medical Device Quality Management Systems 4
R Clinical evaluation Meddev (2.7.1 revision 4) - Example or Template EU Medical Device Regulations 4
A IATF 16949 - 9.2.2.1 Internal Audit Programme - looking for example IATF 16949 - Automotive Quality Systems Standard 5
S Looking for an example of a medical device file ISO 13485:2016 - Medical Device Quality Management Systems 1
C PFMEA example with machining (lathe+mill) FMEA and Control Plans 1
L Simple Example for PFMEA / Control Plan Training Training - Internal, External, Online and Distance Learning 1
Similar threads


















































Top Bottom