An indication that the device is a medical device (MDR, Annex 23.2q) - applicable for accessories?

RA_QA_Expert

Starting to get Involved
#1
Hi all,

MDR (Annex I, point 23.2q Information on the label) requires an indication that the device is a medical device.
Is that applicable also for accessories? Based on definition ‘accessory for a medical device’ means an article which, whilst NOT being itself a medical device.

Thank you in advance for clarification.
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
Accessory is not a medical device per Article 2(2) and so it should not bear the indication for a medical device per Annex I, section 23.2(q).
 

James

Involved In Discussions
#3
Accessory is not a medical device per Article 2(2) and so it should not bear the indication for a medical device per Annex I, section 23.2(q).
For added clarity, all of chapter 3 applies to accessories (the accessory being 'the device' in all of the other clauses) apart from 23.2(q) which is explicit in stating "medical device"?

Thanks

James
 

L_O_B

Involved In Discussions
#4
That is correct.
'Devices' include medical devices, accessories and products listed in annex XVI (see article 1(4)).
'Accessory for a medical device' is by definition (see article 2(2)) not a medical device.
 

kunal9218

Starting to get Involved
#5
That is correct.
'Devices' include medical devices, accessories and products listed in annex XVI (see article 1(4)).
'Accessory for a medical device' is by definition (see article 2(2)) not a medical device.
But what if the manufacturer maintains the DHF as a device and not just as an accessory?
 

shimonv

Trusted Information Resource
#6
KP,
The way the information is organised, whether as one or in two separate DHF's does not change the definition of what an accessory is.
 
Thread starter Similar threads Forum Replies Date
V Changing Indication of Use for a Class IIa Medical Device EU Medical Device Regulations 2
R Class II Medical Device - Marketing vs. Indication For Use (IFU) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
K 510(k) expanding the indication for use for a cleared device Other US Medical Device Regulations 9
S Indication of Inoperable ECG test result IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
V Colour of indication for ALARM SIGNAL inactivation states IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
E IEC 60601-2-57 item 201.12.4.2 - Indication of parameters relevant to safety IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Indications for Use - Is it possible to have a different indication for use EU Medical Device Regulations 1
I Adverse Event Reported during Non-Approved Indication Use EU Medical Device Regulations 4
R Coexisting Investigational Use and CE-Mark - Adding a new Indication EU Medical Device Regulations 4
T "Indication for Use" Identifying Specific Procedures US Food and Drug Administration (FDA) 2
V Old 510(k) with no Indication Statement 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
K Indication for Use/Intended Use/New Use - Different look but functionally the same 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
S CE Label having the indication for the applied part EU Medical Device Regulations 7
C Requirements or Guidance for Pediatric Indication for Use (IVD) EU Medical Device Regulations 4
Mohammed Gouda What is the indication of "adjusted scale correction" to calibration of liq thermom.? General Measurement Device and Calibration Topics 5
T "Indication for Use" used in 510(k) - Indications for Use vs. Intended Use ISO 13485:2016 - Medical Device Quality Management Systems 7
P Verifying improvement results - Indication that controls have worked Statistical Analysis Tools, Techniques and SPC 5
P Temperature indication In Vitro Medical Devices USA 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Q Storing and developing SAMD (Software as a Medical Device) in the Cloud IEC 62304 - Medical Device Software Life Cycle Processes 0
I ISO 2233:2000 Question - Medical Device Shipping/Transportation Validation Other ISO and International Standards and European Regulations 1
A How to make sure whether the predicated device is marked with EC certificate? EU Medical Device Regulations 7
R Identify Medical Device characterstics as Annex C of ISO 14971 Risk Management ISO 14971 - Medical Device Risk Management 5
shimonv FDA guidance on Multiple Function Device Products Other US Medical Device Regulations 0
O ANATEL certification of Medical Device Other Medical Device Regulations World-Wide 0
B MDR equvalent device - can you use a medicine as an equivalent 'device' EU Medical Device Regulations 8
Ajit Basrur FDA News Harmonizing and Modernizing Regulation of Medical Device Quality Systems (7-2020) US Food and Drug Administration (FDA) 0
K Is UDI (Unique Device Identifier) symbol mandatory under EU MDR? EU Medical Device Regulations 1
A Legal Manufacturer Medical device US Food and Drug Administration (FDA) 2
D Device functionality over service life - Objective evidence required? Design and Development of Products and Processes 10
F How to register class I device to all European Competent Autorities(CA) when we are European manufacturer EU Medical Device Regulations 4
K Sterile barrier system symbols - Label redlines for a Class III device (Kit Box) EU Medical Device Regulations 0
D Class 3 Device - Change of supplier of material Canada Medical Device Regulations 6
N Usability testing required for FDA IDE (investigational device exemption)? Human Factors and Ergonomics in Engineering 3
K EU MDR Annex 1 Chapter III: Information in the Instructions for Use-23.4 (e) the performance characteristics of the device; EU Medical Device Regulations 1
R Shall a new UDI-DI be required when stand-alone software device's version is updated? EU Medical Device Regulations 1
S Looking for Quality Content to Build Medical Device Curriculum - Concept to Commercialization Training - Internal, External, Online and Distance Learning 4
A Interpretation of GMP Requirements for class 1 medical device manufacturer (device GMP exempt, only General controls applicable) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
A Legal Manufacturer of a medical device Vs Legal Manufacturer of MDSW EU Medical Device Regulations 7
T Loaded spring device - Active medical device? EU Medical Device Regulations 1
MDD_QNA Medical Device Software - Is a Help Button required? IEC 62304 - Medical Device Software Life Cycle Processes 1
H MDD Article 12 Labeling for Class IIa Medical Device - Please Advise EU Medical Device Regulations 2
J Does Pakistan Medical Device Import License allows parallel import? Other Medical Device Regulations World-Wide 0
K China Medical Device Labeling requirements - Language China Medical Device Regulations 3
F Software as a Medical Device (SaMD) Technical File Requirements Manufacturing and Related Processes 1
A Sample Agreement available for Outsourcing Medical Device Design activity? ISO 13485:2016 - Medical Device Quality Management Systems 1
dgrainger Informational EU medical device website change from 'Growth' to 'Health and Food Safety' (6/2020) Medical Device and FDA Regulations and Standards News 1
M Requirements for an MDD approved device after 26/05/2021 EU Medical Device Regulations 3
F When to market and distribute device in EU? EU Medical Device Regulations 3
B How to classify a medical device based on MDR? EU Medical Device Regulations 3
J Need for a cleanroom in the manufacture of a medical device for a clinical trial EU Medical Device Regulations 4
Similar threads


















































Top Bottom