Analytical Balance Calibration and Verification

H

htquality

#1
Our lab is using an analytical balance and it is calibrated annually by an external provider. The calibration covers a range starting with a 10 mg weight, and the provider will not go any lower (I believe thay don't have the weights to do so).
Before each use, the lab verifies accuracy of the balance with calibrated standard weights. They go down to 1 mg and always bracket the sample weights to be weighed.
Do you see any potential issue by weighing below the minimum weight used by the external calibration lab or if the fact that we verify our range before use covers us? Any risk with regulatory agencies (FDA, EMA, ANVISA, others)?
Thank you for your help!
 
Elsmar Forum Sponsor
#2
Our lab is using an analytical balance and it is calibrated annually by an external provider. The calibration covers a range starting with a 10 mg weight, and the provider will not go any lower (I believe thay don't have the weights to do so).
Before each use, the lab verifies accuracy of the balance with calibrated standard weights. They go down to 1 mg and always bracket the sample weights to be weighed.
Do you see any potential issue by weighing below the minimum weight used by the external calibration lab or if the fact that we verify our range before use covers us? Any risk with regulatory agencies (FDA, EMA, ANVISA, others)?
Thank you for your help!
Calibration is defined as the set of operations, under specified conditions, establishing the relationship between values of quantities by a measuring instrument or measuring system, or values represented by a material measure or reference material, and the corresponding values realized by standards.

Thus the process of calibration establsihes that all the measurements within the range are "true" and hence using the lower and higjher weights in calibration validates your routine measurements.
 

BradM

Leader
Admin
#3
Do you use this balance for assay weighing? Could this be a GLP balance? If so, then USP41 will come into play here.

I would be very interested to know why your provider can't verify that range. How low of weight is the balance used for?


What is the full range of the balance?
 
H

htquality

#5
Hello -
I believe the calibration company does not have the 1 mg capability and are not interested in expanding their range.
As for the balance, it is a Mettler-Toledo MX5 microbalance:
  • Maximum Capacity5.1 g
  • Readability1 µg
  • Taring range5.1 g
  • Repeatability0.8...0.9 µg
  • Settling time (typical)< 10s
  • Adjustment with internal weights proFACT, fully automatic time- and temperature-controlled adjustment
  • Adjustment with external weights custom weight
  • Linearizationautomatic linearization with each internal adjustment
  • Sensitivity temperature drift± 1.0 ppm/°CSensitivity stability± 0.0001%
Thank you!
 
Last edited by a moderator:

BradM

Leader
Admin
#6
Hello -
I believe the calibration company does not have the 1 mg capability and are not interested in expanding their range.

As for the balance, it is a Mettler-Toledo MX5 microbalance:
  • Maximum Capacity5.1 g
  • Readability1 µg
  • Taring range5.1 g
  • Repeatability0.8...0.9 µg
  • Settling time (typical)< 10s
  • Adjustment with internal weights proFACT, fully automatic time- and temperature-controlled adjustment
  • Adjustment with external weights custom weight
  • Linearizationautomatic linearization with each internal adjustment
  • Sensitivity temperature drift± 1.0 ppm/°CSensitivity stability± 0.0001%
Thank you!
:agree1: Thanks for the details.

Is this balance used for GLP studies?
What is the lowest weight that you measure on the balance?
 
H

htquality

#7
Yes, the balance is used for GLP activities. The lab would like to be able to use it down to 1 mg.
 

Ninja

Looking for Reality
Trusted Information Resource
#8
The calibration covers a range starting with a 10 mg weight, and the provider will not go any lower (I believe thay don't have the weights to do so).

Before each use, the lab verifies accuracy of the balance with calibrated standard weights. They go down to 1 mg and always bracket the sample weights to be weighed.
Okay, what am I missing?

They have calibrated standard weights down to 1mg, but dont have weights below 10mg.

I've gotta be missing something obvious again...
 
#9
Yes, the balance is used for GLP activities. The lab would like to be able to use it down to 1 mg.
The AB has Max capacity of 5.1g. What is the Min capacity?
The load cell will be capable to work within its Min and Max capacity. Below Min it does not respond within its accuracy and above Max its an over load and the display indicates such.
Between the Min value and 5.1g its resolution can be like .01g , .001g or higher.
If this Min value is 10mg, then in no way you can use this for any weights less than 10mg.
 

BradM

Leader
Admin
#10
If you are performing GLP studies, then you most likely will need to comply with USP 40.

You will need to determine the minimum weight capability of your balance by establishing precision using the ten readings, standard deviation, etc.

The balance needs to be verified across the use range, and corner load verifications should be performed also.

Now... if you are performing verifications before/after use using weights covering your entire range of use, then the specific test points the outside contractor is performing is less significant. I would venture they are primarily calibrating the internal weights on the balance.

However, I still think it shouldn't be a big problem for them to check at lower points.
 
Thread starter Similar threads Forum Replies Date
J Log Book for Analytical Balance Calibration Verification General Measurement Device and Calibration Topics 4
V Operating Range & Calibration for Analytical Balance General Measurement Device and Calibration Topics 3
D Tolerance when verifying an analytical balance General Measurement Device and Calibration Topics 3
V Drift and Variation over Time of an Analytical Balance. Qualification and Validation (including 21 CFR Part 11) 2
F How to determine Minimum Weight of analytical Balance Measurement Uncertainty (MU) 11
R Document for daily check of analytical balance and criteria General Measurement Device and Calibration Topics 2
K Analytical Method Qualification Vs Validation expectations ISO 13485:2016 - Medical Device Quality Management Systems 1
V Handling deviations during analytical method validations Qualification and Validation (including 21 CFR Part 11) 4
R MSA Studies for chemical analytical instruments in QC Lab Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
R Tolerances for Analytical Balances General Measurement Device and Calibration Topics 3
C FDA on changing acceptance criteria re: analytical method validation US Food and Drug Administration (FDA) 1
M Analytical Test Instruments Used in V&V Manufacturing and Related Processes 2
K QMS Software Recommendations for an Analytical Laboratory General Measurement Device and Calibration Topics 2
V For Analytical Laboratory Equipment, is PQ = Calibration Criteria? Qualification and Validation (including 21 CFR Part 11) 2
K Validation of an Analytical Method that relies on Multi-Variate Calibration General Measurement Device and Calibration Topics 7
V Pharmaceutical Analytical Method Validation and Gauge R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
K Do we need to perform Analytical Method Validation upon Relocation? General Measurement Device and Calibration Topics 5
AnaMariaVR2 Innovation In Assays And Analytical Methods Urgently Needed [article] US Food and Drug Administration (FDA) 0
I Statistical Stability for the PQ of Analytical Equipment Qualification and Validation (including 21 CFR Part 11) 1
A Writing a Procedure for Validation of an Analytical Method General Measurement Device and Calibration Topics 5
Q FDA Bacteriological Analytical Manual (BAM) - Health Care Items Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
J Flexural Strength Testing - Analytical Laboratory recommendations for Testing Alumina Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
Q Analytical Tools to increase or improve Analytical Skills Problem Solving, Root Cause Fault and Failure Analysis 1
E Process Capability Analytical Tools Statistical Analysis Tools, Techniques and SPC 1
B Analytical Lab Environmental Condition (Temperature & Humidity Control Limit) General Measurement Device and Calibration Topics 5
smryan Analytical Scale - Seeking brand and buying advice General Measurement Device and Calibration Topics 7
G Analytical Balances - Looking for Variance Source Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 14
H Analytical Environmental Laboratory ISO 17025 Registration Scope ISO 17025 related Discussions 3
J Association of Analytical Communities validation study data - How deep should I go? General Measurement Device and Calibration Topics 1
Geoff Cotton NWA (Northwest Analytical, Inc.) SPC software - Your Opinion, Review and Experiences Statistical Analysis Tools, Techniques and SPC 2
C Time Crunch on ASQ's More Analytical Exams Professional Certifications and Degrees 12
O Analytical Method Transfer Approval and Use Document Control Systems, Procedures, Forms and Templates 4
B Validation vs. Confirmation in the context of an Analytical Laboratory General Measurement Device and Calibration Topics 10
C AIAG Attribute MSA - Analytical Method - Page 135 of MSA manual 3rd edition Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
R MSA Analytical Method Acceptance Criteria and Requirements Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P Pareto Charts - Analytical tools - Form Attached Quality Tools, Improvement and Analysis 5
H Food and Drink in Production Areas - Manufacturer of FDA analytical and diagnostic ISO 13485:2016 - Medical Device Quality Management Systems 5
S Tolerance limits for micro balance General Measurement Device and Calibration Topics 1
T MSE for dynamic balance measurement? Manufacturing and Related Processes 1
B FDA Philippines has no check or balance outside the health department Supplier Quality Assurance and other Supplier Issues 3
R Fixing of tolerance limit - Electronic weighing balance with the capacity of 150 Kgs General Measurement Device and Calibration Topics 5
M Establishment of Acceptance Criteria for a Weigh Balance General Measurement Device and Calibration Topics 3
P Applying Combined Tolerances of Two Scales - Balance Tolerance Question General Measurement Device and Calibration Topics 9
J Weighing Balance Accuracy for Plastic Moulded Parts variation study Manufacturing and Related Processes 3
C Basic Internal Calibration Procedure to check the Accuracy of a Balance Calibration Frequency (Interval) 6
N Internal Calibration vs. Accredited Calibration for Balance - Differences Calibration Frequency (Interval) 3
J Line Balance Study and maybe Takt Time and Cycle Time Lean in Manufacturing and Service Industries 3
C Balance Calibration SOP for a Pharmaceutical Environment General Measurement Device and Calibration Topics 4
R Line Balance Simulation when Designing a New Production Line Lean in Manufacturing and Service Industries 1
J Balance Calibration Verification question General Measurement Device and Calibration Topics 9