Hello all,

We are interested in running a test to demonstrate repeated use of our product does not increase temperature beyond some threshold. We purchased a FLIR camera to take the thermal measurements. What if any responsibility do we have to demonstrate the thermal camera is working correctly? Do we need to validate the camera and associated data acquisition software? The manufacturer of the camera had no comments.

Any guidance would be greatly appreciated,

kind regards,

MDD
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#3
Hello all,

We are interested in running a test to demonstrate repeated use of our product does not increase temperature beyond some threshold. We purchased a FLIR camera to take the thermal measurements. What if any responsibility do we have to demonstrate the thermal camera is working correctly? Do we need to validate the camera and associated data acquisition software? The manufacturer of the camera had no comments.

Any guidance would be greatly appreciated,

kind regards,

MDD
Hi,

My :2cents: :

1. The set-up is a measuring device and needs R&R analysis including calibration.

2. The software part needs software validation since it's not an integral part of the camera. Perhaps the software supplier can help with this part.

Cheers,
Ronen.
 
Thread starter Similar threads Forum Replies Date
R Tolerances for Analytical Balances General Measurement Device and Calibration Topics 3
D Tolerance when verifying an analytical balance General Measurement Device and Calibration Topics 3
C FDA on changing acceptance criteria re: analytical method validation US Food and Drug Administration (FDA) 1
K QMS Software Recommendations for an Analytical Laboratory General Measurement Device and Calibration Topics 2
V For Analytical Laboratory Equipment, is PQ = Calibration Criteria? Qualification and Validation (including 21 CFR Part 11) 2
K Validation of an Analytical Method that relies on Multi-Variate Calibration General Measurement Device and Calibration Topics 7
V Pharmaceutical Analytical Method Validation and Gauge R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
V Drift and Variation over Time of an Analytical Balance. Qualification and Validation (including 21 CFR Part 11) 2
K Do we need to perform Analytical Method Validation upon Relocation? General Measurement Device and Calibration Topics 5
AnaMariaVR2 Innovation In Assays And Analytical Methods Urgently Needed [article] US Food and Drug Administration (FDA) 0
I Statistical Stability for the PQ of Analytical Equipment Qualification and Validation (including 21 CFR Part 11) 1
A Writing a Procedure for Validation of an Analytical Method General Measurement Device and Calibration Topics 5
J Log Book for Analytical Balance Calibration Verification General Measurement Device and Calibration Topics 4
H Analytical Balance Calibration and Verification General Measurement Device and Calibration Topics 9
Q FDA Bacteriological Analytical Manual (BAM) - Health Care Items Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
J Flexural Strength Testing - Analytical Laboratory recommendations for Testing Alumina Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
Q Analytical Tools to increase or improve Analytical Skills Problem Solving, Root Cause Fault and Failure Analysis 1
V Operating Range & Calibration for Analytical Balance General Measurement Device and Calibration Topics 3
E Process Capability Analytical Tools Statistical Analysis Tools, Techniques and SPC 1
F How to determine Minimum Weight of analytical Balance Measurement Uncertainty (MU) 11
B Analytical Lab Environmental Condition (Temperature & Humidity Control Limit) General Measurement Device and Calibration Topics 5
smryan Analytical Scale - Seeking brand and buying advice General Measurement Device and Calibration Topics 7
G Analytical Balances - Looking for Variance Source Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 14
H Analytical Environmental Laboratory ISO 17025 Registration Scope ISO 17025 related Discussions 3
J Association of Analytical Communities validation study data - How deep should I go? General Measurement Device and Calibration Topics 1
Geoff Cotton NWA (Northwest Analytical, Inc.) SPC software - Your Opinion, Review and Experiences Statistical Analysis Tools, Techniques and SPC 2
C Time Crunch on ASQ's More Analytical Exams Professional Certifications and Degrees 12
O Analytical Method Transfer Approval and Use Document Control Systems, Procedures, Forms and Templates 4
B Validation vs. Confirmation in the context of an Analytical Laboratory General Measurement Device and Calibration Topics 10
C AIAG Attribute MSA - Analytical Method - Page 135 of MSA manual 3rd edition Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
R Document for daily check of analytical balance and criteria General Measurement Device and Calibration Topics 2
R MSA Analytical Method Acceptance Criteria and Requirements Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P Pareto Charts - Analytical tools - Form Attached Quality Tools, Improvement and Analysis 5
H Food and Drink in Production Areas - Manufacturer of FDA analytical and diagnostic ISO 13485:2016 - Medical Device Quality Management Systems 5
M Test failure Root cause not found Customer Complaints 10
Moumen H Variations between ASTM A29 Standard for steel bars and Mill test certificates specs Manufacturing and Related Processes 1
D Tip test 10° -> 5° IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M Sponsor name presented in the test report EU Medical Device Regulations 4
M Packing Validation & Accelerated Aging Test ISO 13485:2016 - Medical Device Quality Management Systems 2
S OEM full range calibrations vs Validation special test points Medical Device and FDA Regulations and Standards News 0
C IEC 60601 - 8.8.3 Dielectric Strength test. 4kv being applied to the ground conductor?! IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
S Mechanical Test Under FDA Freedom of Information Act Medical Device and FDA Regulations and Standards News 5
S Critical characteristic on manufacturing operations test procedure Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
R IEC 60601-1 Clause 15.3.2, Push test IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S Knee Implant (Femoral -Cobalt chrome)-Sub chronic toxicity test (ISO 10993-11)choice of root Medical Device and FDA Regulations and Standards News 2
A Defining a lower ESD test level in IEC 60601 safety test IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Is there a need for clinical test of Class IIa products (for MDR)? EU Medical Device Regulations 2
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Non parametric test for semi-quantitative data. Statistical Analysis Tools, Techniques and SPC 5

Similar threads

Top Bottom