Applicability of 60601 vs. 60335

#1
I have a client that wants to release a hand-held massager with a heating function. Currently they make "muscle scrapers" and they want to expand their product line. They market their products to the cross-fit community for "muscle recovery" and muscle soreness and these are explicitly intended solely for home use by consumers (not medical professionals or medical environments).

The US and Canada explicitly regulate hand-held massagers as medical devices. This makes sense because there's really no such thing, beyond an erotic vibrator, that classifies as a "hand-held massager" which doesn't intend to sooth muscle aches and soreness. My confusion isn't about the regulation it's around the test requirements.

Because they're medical devices in many jurisdictions everyone keeps saying that they should test to IEC 60601. But there's a standard for massagers, IEC 60335-2-32, which explicitly covers:
– foot massagers;
– water filled foot massagers;
– hand-held massagers;
– massage beds;
– massage belts;
– massage chairs;
– massage pads
Moreover, the Scope of this document does NOT explicitly exclude "appliances for medical purposes (IEC 60601)" like the normative 60335-1 standard. To me that's because, as mentioned, at some level basically all massagers are intended for soreness which makes them implicitly medical.

There is no 60601 series standard for massagers which complicates this tremendously. But there is for a similarly-intentioned device: heating pads (60601-2-35). Interestingly, in the scope of that document it explicitly excludes "Heating devices intended for physiotherapy"--so basically if you're using a heating pad at home for muscle aches and soreness then 60601 doesn't apply and presumably the 60335-2-17 does instead. The home-based 2-17 excludes "appliances specifically intended for use under medical supervision" which makes sense. I feel like this langauge would also be in the massager 2-35 standard if there was a similar version in the 60601 series.

So, why wouldn't a regulatory authority accept testing to 60335-2-35? The FDA will accept the very similar UL 1647, but I'm not sure about the EU or Canada which are both intended markets.

Any thoughts or comments are greatly appreciated,
~Josh
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Super Moderator
#2
The applicability of a regulation is decided not on the name of the device (e.g. "massager") but the intended purpose as stated in the instructions for use and marketing material. So you can have an identical massager but in one the manual says "for relaxation and rejuvenation" which is not medical and another which is "for pain relief" or to "speed up recovery from muscle tear" which are medical.

So could have a three scenarios:
Manufacturer A: does not want to apply medical device regulations, and takes great care to avoid mentioning any specific medical treatments for the device.

Manufacturer B: does not want to apply medical device regulations, but is lazy and wants to sell more so includes all sorts of weird and wonderful claims (e.g. cures arthritis) many of which could reasonably be considered medical but are unsupported. That's illegal, by the way.

Manufacturer C: does want the device to be medical, makes specific claims and backs them up with clinical literature or clinical evidence.

So ... that's the regulatory side. And the FDA is sure to follow this as well.

For the standards, if it is non-medical then it the IEC 60335 series is fine.

If it is medical, technically IEC 60601-1 applies but there could be arguments for using IEC 60335-2-35 if this fits better. As an experienced user of IEC 60601-1, there's a lot of stuff in there that doesn't really make sense for a home use massager and I would suspect that the 335 would actually be more effective in addressing the safety issues specific to massagers, in particular the mechanical aspects.
 
Thread starter Similar threads Forum Replies Date
S Applicability of the standard IEC 60601-2-18 to an arthroscopy pump IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A EN 60601-2-51 applicability to an ECG Event Recorder IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
M DIN EN 60601-1-2 applicability in the world and in the US IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Medical device applicability to WEEE Other ISO and International Standards and European Regulations 2
S Defibrillator Protection applicability When using Multiple applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D ISO 14971 applicability in ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 7
F Applicability of CSR's IATF 16949 - Automotive Quality Systems Standard 1
A Applicability of Photobiological Safety Evaluation for LED used in medical devices Reliability Analysis - Predictions, Testing and Standards 2
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
K Applicability of Cybersecurity EU MDR 2017/745 Annex 1 23.4(ab), 14.2(d) CE Marking (Conformité Européene) / CB Scheme 3
K Applicability of eIFU as per EU MDR 2017/745 Annex 1 23.1 CE Marking (Conformité Européene) / CB Scheme 0
R MDD x PPE Directive - Statement of Non-Applicability EU Medical Device Regulations 3
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
DuncanGibbons Understanding the applicability of Design of Experiments to the IQ OQ PQ qualification approach Qualification and Validation (including 21 CFR Part 11) 5
A Software as Medical Device (SaMD) definition and its applicability Other Medical Device and Orthopedic Related Topics 4
N EU MDR - Applicability Article 22 Systems and Procedure Packs CE Marking (Conformité Européene) / CB Scheme 4
R Applicability of new non-harmonized standards (MDD/MDR) EU Medical Device Regulations 14
A MDR Article 22 applicability - Legal manufacturer EU Medical Device Regulations 8
K Diagnostic X-ray devices - Applicability of Biocompatibility Testing per ISO 10993-1 Manufacturing and Related Processes 7
V Applicability of IEC 62366-1 usability to resorbable bone substitutes IEC 62366 - Medical Device Usability Engineering 9
E Commission Regulation (EU) No. 207/2012 E-IFU Applicability EU Medical Device Regulations 5
M ROHS : Applicability for AIMDs RoHS, REACH, ELV, IMDS and Restricted Substances 2
L Dying a slow SCOPE death - NEW ISO APG Paper on Scope and Applicability May 2020. AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
M Applicability of Means of Protection, working voltage in an Automated External Defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Applicability of IEC 60950-1 Table K2 - Minimum clearances between circuits Various Other Specifications, Standards, and related Requirements 0
R Applicability of ISO 14708-3 RF Transmitter and Implantable RF Receiver Other Medical Device Related Standards 1
R Applicability of ISO 14708-3 - RF transmitter and implantable RF receiver Other Medical Device and Orthopedic Related Topics 1
M Oxygen enriched environment applicability - Operating table used in general surgeries in hospital IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
E CB scheme applicability for battery powered medical devices CE Marking (Conformité Européene) / CB Scheme 0
I IATF 16949 - 8.4.2.3 - Applicability to Shipping Suppliers IATF 16949 - Automotive Quality Systems Standard 11
S CQI-12 - Applicability to Raw Materials Customer and Company Specific Requirements 1
C Applicability of ISO 9001:2015 Section 8.5 - Engineering company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
H Applicability of EN 13795 for Incise Drape EU Medical Device Regulations 0
Q Partial non applicability of ISO 9001 Cl. 7.1.5 (not having services)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M User Interface of Unknown Provenance (UOUP) applicability IEC 62366 - Medical Device Usability Engineering 8
E Applicability of Low Voltage Directive (LVD) for 12V equipment CE Marking (Conformité Européene) / CB Scheme 4
Y Counterfeit Parts - AS9100 D Cl. 8.1.4 - Can I claim no applicability? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
M IEC 62304 Applicability - GUI Control Software IEC 62304 - Medical Device Software Life Cycle Processes 3
D Clarification of Applicability of TS 16949 Requirements to a Non-Automotive Business IATF 16949 - Automotive Quality Systems Standard 13
Richard Regalado Sharing a Statement of Applicability (SOA) for ISO/IEC 27001:2013 IEC 27001 - Information Security Management Systems (ISMS) 2
G Not quite old news: Statement of Applicability IEC 27001 - Information Security Management Systems (ISMS) 3
G Question on Statement of Applicability (SOA) IEC 27001 - Information Security Management Systems (ISMS) 2
L TS 16949 applicability for Supplier's Supplier IATF 16949 - Automotive Quality Systems Standard 8
R ISO 22000:2005 Clause 7.10.4 Withdrawal - Applicability and documentation required? Food Safety - ISO 22000, HACCP (21 CFR 120) 3
A Applicability of electrical safety standards to a dedicated PC IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
A Help me understand the applicability of 21 CFR Part 820.198 (Customer Complaints) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
Y New Draft FDA Guidance for Applicability of GLP in Medical Device Submission Other US Medical Device Regulations 2
G TS 16949 applicability for Bulk Material Product IATF 16949 - Automotive Quality Systems Standard 25
Q Two AS9102 Applicability and Interpretation Questions AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
R HIPAA (Health Insurance Portability and Accountability Act) applicability Other US Medical Device Regulations 3

Similar threads

Top Bottom