Dear Experts,
I am currently designing an AED device.
My device is internally powered and has a plastic enclosure. the patient connections are defib electrodes, used for ECG acquisition and defib shock.
In my design, the high-voltage (2000V) and low-voltage (18V) parts have a common Ground plane (GND). I sent my design for validation to a test lab.
But reviewing my architecture, the test lab for certification considers that ALL the electronic board has a WORKING VOLTAGE of (2000V) due to this common ground.
So the test lab considers that I need MOP 2000V isolation between the high-voltage circuit and electrodes, and also MOP 2000V isolation between the ECG path and electrodes, which means that the ECG path resistors should withstand 4kV D.C. because bridging a 2000V MOP (according to 60601-1 ed.3 sect. 4.8).
On my side I have a totally different interpretation of the regulation:
- I consider the WORKING VOLTAGE of my device to be 18V ( according to 60601-2-4 chapter 201.8.8.3 : For the dielectric tests of the general standard, the WORKING VOLTAGE is determined without regard to the presence of defibrillation or pacing voltages.)
- I consider that High-voltage is isolated from Low-voltage of circuit (even if common ground). This is proven by validation test of 60601-2-4 201.8.8.3
- So I don't need MOP 2000V isolation between ECG path and electrodes
- I don't need MOP 2000V isolation between High-Voltage Circuit and electrodes, because current is intended to flow through patient (60601-1 ed.3 sect. 8.4.1)
- BUT I need to pass all the tests of 60601-2-4.
Could someone tell me if my approach is ok or am I totally wrong in my comprehension of 60601-1 and 60601-2-4?
Thanks in advance!
Mars
I am currently designing an AED device.
My device is internally powered and has a plastic enclosure. the patient connections are defib electrodes, used for ECG acquisition and defib shock.
In my design, the high-voltage (2000V) and low-voltage (18V) parts have a common Ground plane (GND). I sent my design for validation to a test lab.
But reviewing my architecture, the test lab for certification considers that ALL the electronic board has a WORKING VOLTAGE of (2000V) due to this common ground.
So the test lab considers that I need MOP 2000V isolation between the high-voltage circuit and electrodes, and also MOP 2000V isolation between the ECG path and electrodes, which means that the ECG path resistors should withstand 4kV D.C. because bridging a 2000V MOP (according to 60601-1 ed.3 sect. 4.8).
On my side I have a totally different interpretation of the regulation:
- I consider the WORKING VOLTAGE of my device to be 18V ( according to 60601-2-4 chapter 201.8.8.3 : For the dielectric tests of the general standard, the WORKING VOLTAGE is determined without regard to the presence of defibrillation or pacing voltages.)
- I consider that High-voltage is isolated from Low-voltage of circuit (even if common ground). This is proven by validation test of 60601-2-4 201.8.8.3
- So I don't need MOP 2000V isolation between ECG path and electrodes
- I don't need MOP 2000V isolation between High-Voltage Circuit and electrodes, because current is intended to flow through patient (60601-1 ed.3 sect. 8.4.1)
- BUT I need to pass all the tests of 60601-2-4.
Could someone tell me if my approach is ok or am I totally wrong in my comprehension of 60601-1 and 60601-2-4?
Thanks in advance!
Mars