Hi guys,
as the name of the thread already states, I am trying to figure out how to apply the IEC 60601-1 and IEC 60601-2-2 to the device I'm developing.
The device is a laparoscopic instrument which has some low voltage motorized parts inside of the handle unit. The handle unit has a cable which splits up into two connectors: One for a external power source for the motors and one which attaches to an ESU.
kolorado22 already addressed this a bit in this post: HF surgery device accessory insulation requirements
Hi William:
The answer you received about no creepage and clearance requirements for HF Surgical Accessories is correct ONLY for that part of your accessory that is transmitting HF energy to the patient. If your Accessory is POWERED by anything other that the HF surgical current from the ESU then the IEC 60601-1 clauses 8.8.1 and 8.5.1.2 apply. Further if there is any protective or functional earth present in the accessory, that earth connection must be protected from the HF circuits according to 201.8.8.3.101. An example would be an accessory within a robot arm or with a camera built in.
My theory is that the part of the device that is motorized and isolated from the HF circuit can be seen as a F-TYPE APPLIED PART of my power source (60601-1, 8.5.2.1):
F - TYPE ISOLATED ( FLOATING ) APPLIED PART (herein F - TYPE APPLIED PART )
APPLIED PART in which the PATIENT CONNECTIONS are isolated from other parts of the ME EQUIPMENT to such a degree that no current higher than the allowable PATIENT LEAKAGE CURRENT flows if an unintended voltage originating from an external source is connected to the PATIENT , and thereby applied between the PATIENT CONNECTION and earth NOTE F- TYPE APPLIED PARTS are either TYPE BF APPLIED PARTS or TYPE CF APPLIED PARTS .
It doesn't technically make contact with the patients body during normal use however because the instrument shaft wouldn't be considered part of it. That makes me unsure...
APPLIED PART
part of ME EQUIPMENT that in NORMAL USE necessarily comes into physical contact with the PATIENT for ME EQUIPMENT or an ME SYSTEM to perform its function
If however - as kolorado22 said - the part of the device that is transmitting the HF energy from the ESU to the patient can be considered an ACTIVE ACCESSORY (60601-2-2) of the ESU I wonder which quote from the standards gives the possibility to split up the device.
I also wonder what quote from the standards this ("Further if there is any protective or functional earth present in the accessory, that earth connection must be protected from the HF circuits according to 201.8.8.3.101.") is based on. More precisely I would like to know how the two parts of the device need to be insulated from each other and what tests need to be performed regarding this.
I would really appreciate it if anyone could help!
as the name of the thread already states, I am trying to figure out how to apply the IEC 60601-1 and IEC 60601-2-2 to the device I'm developing.
The device is a laparoscopic instrument which has some low voltage motorized parts inside of the handle unit. The handle unit has a cable which splits up into two connectors: One for a external power source for the motors and one which attaches to an ESU.
kolorado22 already addressed this a bit in this post: HF surgery device accessory insulation requirements
Hi William:
The answer you received about no creepage and clearance requirements for HF Surgical Accessories is correct ONLY for that part of your accessory that is transmitting HF energy to the patient. If your Accessory is POWERED by anything other that the HF surgical current from the ESU then the IEC 60601-1 clauses 8.8.1 and 8.5.1.2 apply. Further if there is any protective or functional earth present in the accessory, that earth connection must be protected from the HF circuits according to 201.8.8.3.101. An example would be an accessory within a robot arm or with a camera built in.
My theory is that the part of the device that is motorized and isolated from the HF circuit can be seen as a F-TYPE APPLIED PART of my power source (60601-1, 8.5.2.1):
F - TYPE ISOLATED ( FLOATING ) APPLIED PART (herein F - TYPE APPLIED PART )
APPLIED PART in which the PATIENT CONNECTIONS are isolated from other parts of the ME EQUIPMENT to such a degree that no current higher than the allowable PATIENT LEAKAGE CURRENT flows if an unintended voltage originating from an external source is connected to the PATIENT , and thereby applied between the PATIENT CONNECTION and earth NOTE F- TYPE APPLIED PARTS are either TYPE BF APPLIED PARTS or TYPE CF APPLIED PARTS .
It doesn't technically make contact with the patients body during normal use however because the instrument shaft wouldn't be considered part of it. That makes me unsure...
APPLIED PART
part of ME EQUIPMENT that in NORMAL USE necessarily comes into physical contact with the PATIENT for ME EQUIPMENT or an ME SYSTEM to perform its function
If however - as kolorado22 said - the part of the device that is transmitting the HF energy from the ESU to the patient can be considered an ACTIVE ACCESSORY (60601-2-2) of the ESU I wonder which quote from the standards gives the possibility to split up the device.
I also wonder what quote from the standards this ("Further if there is any protective or functional earth present in the accessory, that earth connection must be protected from the HF circuits according to 201.8.8.3.101.") is based on. More precisely I would like to know how the two parts of the device need to be insulated from each other and what tests need to be performed regarding this.
I would really appreciate it if anyone could help!