Application of Risk Management - ISO 14971:2007(E) Medical Devices

G

Gert Sorensen

#11
Re: ISO 14971:2007(E) Medical devices — Application of risk

Does anyone have a copy of ISO 14971-:cfingers: .
Yes, of course there are. Among others I can mention: Danish Standards Association and IHS. It does not cost more than app. 50$ so it is hardly worth not buying.

Please refrain from asking for copies of copy right protected material. We do not endorse this. :bigwave:
 
Last edited by a moderator:
Elsmar Forum Sponsor
J

JohnM

#12
Re: ISO 14971:2007(E) Medical devices — Application of risk

Wow, $50 is a good deal. I paid a lot more.

I'm stuck on the requirement for the risk management to be an input to product specifications. How do you evaluate the risk on something before you know what it is you are making?

My thought was to have the draft version of the hazard analysis be a part of design phase 1 and the final hazard analysis along with the FMEA being part of design phase 2.

Any thoughts?
 
G

Gert Sorensen

#13
Re: ISO 14971:2007(E) Medical devices — Application of risk

Wow, $50 is a good deal. I paid a lot more.
You're right, I hit the $ sign, should have been £. Still is cheap though. :)

I'm stuck on the requirement for the risk management to be an input to product specifications. How do you evaluate the risk on something before you know what it is you are making?
When you put it like that I have to agree. If you have no grasp of what it is you are going to produce, then of course you cannot evaluate the risk involved. :notme:

My thought was to have the draft version of the hazard analysis be a part of design phase 1 and the final hazard analysis along with the FMEA being part of design phase 2.
And, presto, by now you do know what you are going to produce.... FMEA is your friend. A good FMEA in the design phase will give you the opportunity to improve your design, by reducing or removing potential problems. The problems that are not being "R or R'ed" will be highlighted and that is input for your process and your process control.
:bigwave:
 
R

ron_whitehead - 2009

#14
Re: ISO 14971:2007(E) Medical devices — Application of risk

Design is already done when I start the mfg. process.
Should this risk assessment be the responsibility of the design activity?


Anyone with a good idea or developed risk management tool? I am interested in incorporating this for the initial assessment of the process for manufacturing and the continued residual risk evaluation. My plan is to incorporate the tool into the quality plan.

Thanks
 

Marcelo

Inactive Registered Visitor
#15
Re: ISO 14971:2007(E) Medical devices — Application of risk

John

FMEA should really be used in a late design phase because, as you said, it´s difficult to analyse failure modes of something not yet designed....earlier in the design phase you should use some top-down approach such as Hazop or FTA.

Ron

I think the best use of risk management using it some kind of product life-cycle management (PLM) conformity approach. Risk management isn´t the responsability of the design activity because it has to be used in the product lifecycle.
 
V

Valery

#16
Re: ISO 14971:2007(E) Medical devices — Application of risk

Hello WBubel

Good document!

But i think you really tried to make a guide to a risk management file, not to a risk management report. I´m saying this because you seem to mix the concepts in the document. For example, the POTENTIAL FAILURE, CAUSE OF
FAILURE, EFFECTS OF FAILURE, RESPONSE, RISK INDEX and ACTION TAKEN applies to hazardous situations, not to hazards... hazards are the general "problem", and the hazardous situations are the ways in which the problem can come to pass.

For example:

Patient leakage current is an energy hazard. Thare are X situations in which a potential failure (for example in a component) caused by, for example a wrong line voltage, can lead to a hazardous situation of a patient being exposed to a source of leakage current, which could lead to harm (effects, for example, shock or burn). If the risk of the hazardous situation (potential failure + potential situation + effects) is unacceptable, you have to control the risk (response - for example, limiting the impedance) taking an action, and after that verify that the risk is acceptable (risk index?).

This is the general proccess which have to be performed and recorded in the risk management file. The risk management report, on the other hand, is a summary of the risk management file (the stadard has been revised to remove the traceability requirement on the report because the sizae of reposrt, when you put all the hazardous situations - which are many many many - was too big, which was not the intention of the original requirement).

So, in my opinion, you shoul remove the POTENTIAL FAILURE, CAUSE OF
FAILURE and EFFECTS OF FAILURE (and page 3 too!) as they are not applied to the hazards described. Besides that, you doc lacks the information on implementation of the risk management plan, and the information about methods of obtaining production and post-production information (i´m assuming the RISK INDEX is the acceptable risk defined by the policy of 3.2 a), but i think you´ll have to explicit it better anyway - what you could to is to put a column on "overrall residual risk", and then compare it with the risk index.

Cheers!
Hello Mmantunes,

Could you send me a Risk Management File template? I am in the very beginning of the risk management procedure and would like to be on the right way from the start.
Thank you in advance.
Best regards,
Valery.
 

Marcelo

Inactive Registered Visitor
#17
Re: ISO 14971:2007(E) Medical devices — Application of risk

Hello Valery

I do not have a risk management file template, but you can use the one from WBubel above as a start.

Cheers.
 
J

jkonholm

#18
Re: ISO 14971:2007(E) Medical devices — Application of risk

i'm trying again. but a little bit difference: ISO 14971_2007 voting results (RESULTS OF VOTING AND COMPILATION OF COMMENTS ON
ISO/DIS and IEC/CDV 14971, Medical devices – Application of risk management to medical devices).
and Mr. Gert Sorensen, this is not copyright material. thank you...
Ps: i don't understand if we haven't seen the ISO_DIS docs and how we are think about it!! (and sorry for my worst english) :topic:
 

Attachments

Marcelo

Inactive Registered Visitor
#19
Re: ISO 14971:2007(E) Medical devices — Application of risk

Although ISO and IEC documents that are not drafts are not copyrighted, they are to be used by National Commitees only in the process of standard´s development. You can only have access to these documents through National Commitees and usually if you are participating in the development process. Sharing these documents means that participation is not really necessary and thus weakens the development process (an in countries like Brasil where i live, people want the documents but don´t like to participate).
 
Thread starter Similar threads Forum Replies Date
B Application of Risk Management - ISO 14971 for a Tooling Manufacturer ISO 14971 - Medical Device Risk Management 18
V ISO 14971:2007 Application of Risk Management to Medical Devices ISO 14971 - Medical Device Risk Management 9
J FMEA "Types" - ISO 14971 (Medical Devices - Application of Risk Management) ISO 14971 - Medical Device Risk Management 7
I Fault Tree Analysis (FTA) - Application - Preventive, Corrective or Risk Analysis? Quality Tools, Improvement and Analysis 10
T Google Fit Health datas and Medical devices Application Medical Information Technology, Medical Software and Health Informatics 0
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
lanley liao Are there guidelines for application of all API monogrammal product? Oil and Gas Industry Standards and Regulations 2
G Application to Chartered Quality Institute - MCQI ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 2
supadrai Possible to Submit a 510(k) if you have never, do not and cannot produce the device for which you're submitting the application? US Food and Drug Administration (FDA) 4
M Surgical angle measurement guide device with an application software Medical Device and FDA Regulations and Standards News 1
Sidney Vianna IATF 16949 News Update on the IATF CARA Project (“Common Audit Report Application”) - 12/2020 IATF 16949 - Automotive Quality Systems Standard 1
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
U GMDN CODES application ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
P Monolithic application and medical devices ISO 13485:2016 - Medical Device Quality Management Systems 2
R CAMBODIA - Application of AMDD Other Medical Device Regulations World-Wide 3
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 3
C New Class III medical device application for Health Canada Canada Medical Device Regulations 6
T Commission proposal being drawn up for postponement for date of application of MDR (2017/745) EU Medical Device Regulations 7
N Program or application for standard time measurement in industry from video surveillance? Human Factors and Ergonomics in Engineering 1
J Are DAM tools, Mobile application, clinical dashboard, etc. considered accessories to medical device or supportive functions? Manufacturing and Related Processes 4
J Application of mains voltage on SIP/SOP connectors of medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P SOUP anomaly evaluation for MMA (Mobile Medical Application) IEC 62304 clause 7.1.3 IEC 62304 - Medical Device Software Life Cycle Processes 6
M Informational TGA presentation: How to submit an effective good manufacturing practice clearance application Medical Device and FDA Regulations and Standards News 0
C ISO 9001:2015 Monitoring and measuring resources. Application a service industry ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Z Role of economic operators (EO) post the EU MDR Date of application in May 2020 EU Medical Device Regulations 1
P Scope of application for IEC 60601-1-11 Medical electrical equipment — Part 1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Informational EU – Application of transitional provisions concerning validity of certificates issued in accordance to Directives 90/385/EEC and 93/42/EEC Medical Device and FDA Regulations and Standards News 2
D FDA Biomarker Qualification Program - Has anyone prepared an application? Medical Device and FDA Regulations and Standards News 0
A Validating my own application developed with Agile Qualification and Validation (including 21 CFR Part 11) 5
M Informational EU – Guidance document Implant Card relating to the application of Article 18 Regulation (EU) 2017/745 of the European Parliament and of the Council o Medical Device and FDA Regulations and Standards News 2
Sidney Vianna Informational ISO 18091 - Guidelines for the application of ISO 9001 in local government Other ISO and International Standards and European Regulations 1
S Internal Audit Checklist for Application/Software development IEC 27001 - Information Security Management Systems (ISMS) 1
M Informational MDCG 2019-2 Guidance on application of UDI rules to device-part of products referred to in Article 1(8), 1(9) and 1(10) of Regulation 745/2017 Medical Device and FDA Regulations and Standards News 0
D I'm new in IMDS special using the IMDS application Manufacturing and Related Processes 8
M Medical Device News NB-Med and Team-NB have jointly edited a white paper entitled “White Paper – One Year of Application” Medical Device and FDA Regulations and Standards News 0
S IATF 16949 9.1.1.3 Application of statistical concepts - Our technicians are quizzed for statistical knowledge IATF 16949 - Automotive Quality Systems Standard 3
supadrai Informational Japan Minor Change Notification Application Form - Found it? Japan Medical Device Regulations 0
M Some possible misunderstandings on the application of ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 10
K Need feedback: FMEA Web Application FMEA and Control Plans 5
I FAR Part 43 and Application to 17025 Accredited Cal Lab Federal Aviation Administration (FAA) Standards and Requirements 2
B Request for Information; Application for Accreditation Japan Medical Device Regulations 0
Y Application of IEC/EN 62304 at an advanced stage of software development IEC 62304 - Medical Device Software Life Cycle Processes 4
M What is considered an "Audit Day" for OASIS AEA application? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
Ajit Basrur CDRH Premarket Approval Application Critical to Quality Pilot Program US Food and Drug Administration (FDA) 0
Sidney Vianna Blockchain Technology - Any examples of practical application? The Reading Room 21
V Delete Restrictions to be enabled AT BOTH Operating System & Application Levels Qualification and Validation (including 21 CFR Part 11) 1
R Is it legal to attend a trade show while you have a 510(k) application pending? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
D Zinc Oxide as a Topical Application EU Medical Device Regulations 4
C Application of ISO 9001:2015 Clause Interpretations for Service Industries Service Industry Specific Topics 2