Application of Risk Management - ISO 14971 for a Tooling Manufacturer

B

Bud S.

This might not be the correct location for this but please understand, I am reasonably new to Elsmar.
I have been asked by a friend to give his small company a little guidance toward ISO 13485 implementation and eventual certification.
I have led several facilities to 9001:2000 certification and figured I could help my friend before his company spent a bundle on consultants.
Now for my question: While I find it easy to help them with most of their documentation and preparation, I'm stuck on Risk Management. They DO NOT manufacture any medical products. They make tooling (for injection molders). They were asked by a couple of their customers to be certified 13485 because the customers make medical products with the tooling (molds/dies) my friends company designs and builds. How extensive must they be in Risk Management if most of it "does not apply"??

Thank you,
Bud S.
 
Last edited by a moderator:

somashekar

Leader
Admin
This might not be the correct location for this but please understand, I am reasonably new to Elsmar.
I have been asked by a friend to give his small company a little guidance toward ISO 13485 implementation and eventual certification.
I have led several facilities to 9001:2000 certification and figured I could help my friend before his company spent a bundle on consultants.
Now for my question: While I find it easy to help them with most of their documentation and preparation, I'm stuck on Risk Management. They DO NOT manufacture any medical products. They make tooling (for injection molders). They were asked by a couple of their customers to be certified 13485 because the customers make medical products with the tooling (molds/dies) my friends company designs and builds. How extensive must they be in Risk Management if most of it "does not apply"??

Thank you,
Bud S.
OK., while I have reservations about 13485 for such an organization, perhaps the customers drive this from the CMDCAS view point as I guess. Now coming to the point of risk management, a good process FMEA should answer the requirement. Can this make you more comfortable .... The FMEA ?
 

jkuil

Quite Involved in Discussions
First of all, ISO 13485 only applies to manufacturers of medical devices and not to their suppliers. And especially when they make tooling, they are way of the scope of ISO 13485. It is adequate to have ISO 9001, in which you seem experienced. You have to comply with the latest revision which is ISO 9001:2008.

For risk management, that would only apply to their manufacturing process. Again ISO 14971 is the risk management standard applicable to medical devices and does not apply to a tooling manufacturer. It is good reading though, because it helps you understand that risk management is more than a FMEA. Risks should be managed thoughout the life-time of a product.
 
B

Bud S.

Many thanks to you. While it is true that 9001:2008 would be sufficicient for them at this time, I am led to believe their customers will be wanting them to do some molding of subcomponents which will later be used in medical device assemblies. While this is not yet the case, it is a probable activity in the future.

I think I will simply try to figure out a way to set up an in-process inspection format for them in Excel and recommend they read up on Risk MAnagement somewhere.

As for FMEA, I don't know how to steer them into that. Fmea is not used in most of the companies I've helped through ISO 9001:2008 certification. I have always thought that to be more for automotive parts manufacturing.

Thanks again,
Bud S.
 
R

Roland Cooke

First of all, ISO 13485 only applies to manufacturers of medical devices and not to their suppliers.

This is a dangerous statement.

Sterilization facilities, plastic component molders, electronic equipment assemblers, cleanroom packaging companies, translation agencies, regulatory consultancies all (can) hold ISO13485. None of these organizations makes or markets a medical device. Many would not be in business without holding ISO13485.

That said, I would agree that a tooling manufacturer is indeed too far down the food chain to warrant specifically holding ISO13485.
 
B

Bud S.

Thank you Mr. Cooke.

While 13485 is almost a mirror of 9001:2008 it is a bit over the top for a tooling manufacturer. I am guessing that the customers do drive this from a CMDCAS viewpoint as one respondant suggested. Also, there is the probability that they will be asked to mold, or assemble some medical subcompomants in the not-to-distant future.
I will confer again with their chosen registrar to be certain.

As a quality professional, I am always pleased to see a company work toward a viable QMS.

Many thanks to you and all who have responded in this thread.

Respectfully,
Bud S.
 

Ajit Basrur

Leader
Admin
This is a dangerous statement.

Sterilization facilities, plastic component molders, electronic equipment assemblers, cleanroom packaging companies, translation agencies, regulatory consultancies all (can) hold ISO13485. None of these organizations makes or markets a medical device. Many would not be in business without holding ISO13485.

That said, I would agree that a tooling manufacturer is indeed too far down the food chain to warrant specifically holding ISO13485.

I do not think this is true - ISO 13485 is only applicable for medical device manufacturers and not to plastic component manufacturers (that go into medical device) or molders. Having said this, yes, if the system is aligned towards the ISO 13485 requirements, it would be easy for them to do business with device manufacturers.
 
Thread starter Similar threads Forum Replies Date
V ISO 14971:2007 Application of Risk Management to Medical Devices ISO 14971 - Medical Device Risk Management 9
W Application of Risk Management - ISO 14971:2007(E) Medical Devices ISO 14971 - Medical Device Risk Management 69
J FMEA "Types" - ISO 14971 (Medical Devices - Application of Risk Management) ISO 14971 - Medical Device Risk Management 7
I Fault Tree Analysis (FTA) - Application - Preventive, Corrective or Risk Analysis? Quality Tools, Improvement and Analysis 10
P Application of 60601-1 and 60601-2-2 to a hand-held laparoscopic unstrument with motorisation in the handle IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
D CFG application issues Other US Medical Device Regulations 0
T Monolithic medical application ISO 13485:2016 - Medical Device Quality Management Systems 1
D GR&R for Torque application Manufacturing and Related Processes 6
J Has the facility been inspected? question in Certificate to Foreign Government application Other US Medical Device Regulations 0
Cats Clause ISO 9001:2015 7.1.5 Monitoring and measuring resources: Application to training providers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
M Can you import medical device not FDA approved into the USA under an IND application? US Food and Drug Administration (FDA) 2
E Do I need test reports for all standards mentioned in my 510k application? Other Medical Device Related Standards 8
S Taiwan QSD application Other Medical Device Regulations World-Wide 1
shimonv Where do you place Human Factors report in a 510(k) application? Other US Medical Device Regulations 3
Jessesun The application of ozone sterilization in medical treatment US Food and Drug Administration (FDA) 0
L API Q1 Application Oil and Gas Industry Standards and Regulations 2
I Software (SaMD) mobile application verification testing: objective evidence Medical Information Technology, Medical Software and Health Informatics 6
D Dental application - medical product or not? Other US Medical Device Regulations 8
T Including a class 1 camera in the 510k application of class 2 therapeutic technology? US Food and Drug Administration (FDA) 18
shimonv Can you submit a traditional application for a class I device? Other US Medical Device Regulations 7
A Should we assign the PRRC before the date of application of MDR (26 May 2021)? EU Medical Device Regulations 0
T Google Fit Health datas and Medical devices Application Medical Information Technology, Medical Software and Health Informatics 2
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
lanley liao Are there guidelines for application of all API monogrammal product? Oil and Gas Industry Standards and Regulations 2
G Application to Chartered Quality Institute - MCQI ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 2
supadrai Possible to Submit a 510(k) if you have never, do not and cannot produce the device for which you're submitting the application? US Food and Drug Administration (FDA) 4
M Surgical angle measurement guide device with an application software Medical Device and FDA Regulations and Standards News 4
Sidney Vianna IATF 16949 News Update on the IATF CARA Project (“Common Audit Report Application”) - 12/2020 IATF 16949 - Automotive Quality Systems Standard 1
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
U GMDN CODES application ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
P Monolithic application and medical devices ISO 13485:2016 - Medical Device Quality Management Systems 2
R CAMBODIA - Application of AMDD Other Medical Device Regulations World-Wide 3
D Does Manufacture can submit CE mark application under MDD with NB for his New product after May 2020? EU Medical Device Regulations 4
C New Class III medical device application for Health Canada Canada Medical Device Regulations 6
T Commission proposal being drawn up for postponement for date of application of MDR (2017/745) EU Medical Device Regulations 7
N Program or application for standard time measurement in industry from video surveillance? Human Factors and Ergonomics in Engineering 1
J Are DAM tools, Mobile application, clinical dashboard, etc. considered accessories to medical device or supportive functions? Manufacturing and Related Processes 4
J Application of mains voltage on SIP/SOP connectors of medical device IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
P SOUP anomaly evaluation for MMA (Mobile Medical Application) IEC 62304 clause 7.1.3 IEC 62304 - Medical Device Software Life Cycle Processes 6
M Informational TGA presentation: How to submit an effective good manufacturing practice clearance application Medical Device and FDA Regulations and Standards News 0
Cats Clause ISO 9001:2015 Monitoring and measuring resources. Application a service industry ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Z Role of economic operators (EO) post the EU MDR Date of application in May 2020 EU Medical Device Regulations 1
P Scope of application for IEC 60601-1-11 Medical electrical equipment — Part 1-11 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Informational EU – Application of transitional provisions concerning validity of certificates issued in accordance to Directives 90/385/EEC and 93/42/EEC Medical Device and FDA Regulations and Standards News 2
Dobby1979 FDA Biomarker Qualification Program - Has anyone prepared an application? Medical Device and FDA Regulations and Standards News 0
A Validating my own application developed with Agile Qualification and Validation (including 21 CFR Part 11) 5
M Informational EU – Guidance document Implant Card relating to the application of Article 18 Regulation (EU) 2017/745 of the European Parliament and of the Council o Medical Device and FDA Regulations and Standards News 2
Sidney Vianna Informational ISO 18091 - Guidelines for the application of ISO 9001 in local government Other ISO and International Standards and European Regulations 1
S Internal Audit Checklist for Application/Software development IEC 27001 - Information Security Management Systems (ISMS) 1

Similar threads

Top Bottom