Are disposable electrodes in the scope of WEEE Directive?

Mark Meer

Trusted Information Resource
#1
The Directive 2012/19/EU on Waste Electrical and Electronic Equipment (WEEE) defines "Electrical and Electronic Equipment" (EEE) as:

"...equipment which is dependent on electric currents or electromagnetic fields in order to work properly and equipment for the generation, transfer and measurement of such currents and fields and designed for use with a voltage rating not exceeding 1000 volts for alternating current and 1500 volts for direct current;" (Article 3, 1(a))

We've got some disposable electrodes made of only plastic, cotton and stainless steel.

Sure, when connected to a parent stimulator, these electrodes transfer current, and hence appear to fit the EEE definition. ...but given their material composition, it seems strange to have to subject them to special disposal instructions...
 
Elsmar Forum Sponsor

AgnieszkaSz

Involved In Discussions
#3
In 2012 update of WEEE directive, in Article 1 (Scope) clause 4 says that "This shall not apply to [...] (g) medical devices and in vitro diagnostic medical devices, where such devices are expected to be infective prior to end of life, and active implantable medical devices". So to me it seems that used disposable electrodes are out of scope of WEEE.
 
Last edited by a moderator:
Thread starter Similar threads Forum Replies Date
T Do Disposable Electrodes need CT-Compatible Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
T No Defined Shelf Life/Expiration Date - Disposable single-use, non-invasive, non-sterile Other Medical Device Regulations World-Wide 2
T 5ºC Storage Temperature Validation - Urology disposable devices EU Medical Device Regulations 3
T Must all Disposable Medical Devices be DEHP Free for MDR? EU Medical Device Regulations 1
S Scope change in Japan - Class I disposable medical devices Japan Medical Device Regulations 0
Y Shelf Life of Non-Sterile Disposable Drug Inhaler 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
P Single Use Disposable Dental Syringe - CE Marking Advice EU Medical Device Regulations 4
S IEC 60601-1 - Label Physical Requirements for Disposable (Single Use) Medical Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
N Reliability Testing of a Single-Use (Disposable) Medical Device Reliability Analysis - Predictions, Testing and Standards 4
B PPAP for Disposable Tooling APQP and PPAP 2
S Food Safe Ink for reusable and disposable gloves US Food and Drug Administration (FDA) 2
A Calibration of Disposable Weight Scales General Measurement Device and Calibration Topics 2
S Resterilization - Disposable devices Imported Legacy Blogs 0
S Resterilization of Disposable Medical Devices by a Distributor Other Medical Device and Orthopedic Related Topics 5
D Class II Disposable Medical Device Label Content Requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N Accelerated Aging Test for Disposable Protective Garment for Incontinence Other US Medical Device Regulations 1
S Disposable Loop Electrode - GMDN/or UMDNS code? Imported Legacy Blogs 3
M RoHS for Disposable Medical Device Assemblies RoHS, REACH, ELV, IMDS and Restricted Substances 1
L Predicate Device for 510(k) - Disposable vs Reusable 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
D Disposable Medical Device Sampling Plans per ISO 2859-1 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
F HACCP Certification for Single Use Disposable Food Contact Grade Gloves Food Safety - ISO 22000, HACCP (21 CFR 120) 1
B Blood Bag Classification - with Disposable Centrifugal Unit for processing blood EU Medical Device Regulations 4
C ISO17020 Required to ship disposable "coveralls/overgarments" into Egypt General Measurement Device and Calibration Topics 6
S Surgical Instrument Luminaries Standards - Single Use Disposable Surgical Instrument IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Ajit Basrur Manufacturing of sterile single use disposable devices in a single plant? ISO 13485:2016 - Medical Device Quality Management Systems 4
D Battery Powered Disposable Pump - What IEC 60601 testing is required, if any? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Batch Processing Record for Medical Devices - Disposable Plastic Syringes Other Medical Device Regulations World-Wide 1
somashekar Waste Categorization - Medical Device (Disposable Sensor) Miscellaneous Environmental Standards and EMS Related Discussions 4
N RFID on a Medical Disposable Kit - Active Medical Device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
M Classification of a disposable Screen Pointer for use by surgeons on monitors? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
W Distributor Requirements for Sterile Disposable Medical Devices EU Medical Device Regulations 8
M Cleanroom Gowning Disposable Lab Coats - Class 8 Cleanroom ISO 13485:2016 - Medical Device Quality Management Systems 4
A Removing the liner from disposable nicotine patch without using hand ISO 13485:2016 - Medical Device Quality Management Systems 11
Jim Wynne We have entered the age of disposable laser printers After Work and Weekend Discussion Topics 23
D Defibrillation protection for patient connection except for electrodes IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S IEC 60601-2-25 ECG 4 electrodes and derived leads IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
N ECoG/EEG electrodes - Required IEC 60601-1 tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Are EEG electrodes considered as one applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Standards for testing of adhesive electrodes for use with TENS machines IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Electrosurgical Electrodes - Stability Test Other Medical Device Related Standards 1
B Applications required for main accessories (electrodes) of CE Medical Products EU Medical Device Regulations 4
N Must electrodes for Rx/prescription stimulators be Rx only? Other US Medical Device Regulations 2
J IEC 60601-1 for TENS-electrodes (CE-marking) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
J Sourcing Rubber or Silicone Electrodes for Medical Device Other ISO and International Standards and European Regulations 2
A Classification of ECG/EKG Electrodes according to MDD 93/42/EEC? EU Medical Device Regulations 2
D ECG Electrodes - Duration of Use (How long they are expected to stay on the patient) Other US Medical Device Regulations 2
D Independent Testing of ECG Electrodes in accordance with AAMI standards Reliability Analysis - Predictions, Testing and Standards 5
N CE Mark (CE directives) for Welding Electrodes CE Marking (Conformité Européene) / CB Scheme 5
bio_subbu FDA Issues Guidance Document on Electrocardiograph (ECG) Electrodes Other US Medical Device Regulations 0
J Pull test standard? Wires coming off the electrodes at the weld ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5