Are EEG electrodes considered as one applied part?

EMengineer

Involved In Discussions
Hi,

I am a little confused and would like to ask about EEG electrodes, are they considered as one applied part or every single electrode is different applied part? Applied Leakage current test methods depends on how these electrodes are defined.

What I found in 601-1 3rd ed. is that "the APPLIED PART includes the electrodes and those parts of the patient cable that need to physically contact the patient in normal use" (explanation above A.7 pic.). It explains about ECG electrodes as APPLIED PART (not parts), but I assume in my situation it is the same.

Could someone confirm or deny it?
 
M

MedMartin

Hi EMengineer,

yes, from my experience designing ECGs:
Your EEG electrodes are considered one applied part.

A device with e.g. two applied parts could be a fictive monitor with EEG and ECG inputs.
See also in IEC 60601-1 the explanation to subclause 3.8:
"c) If there is an applied part, there can be one or more patient connection(s). ..."

Best regards
Martin
 

Souheil

Registered
You may consider each electrode as distinct applied part, but in this case, that electrode shall be separated from the other AP (other electrodes) in terms of leakage current, dielectric test and creepage distances (see clause 8.5.2.1)
 
Thread starter Similar threads Forum Replies Date
R Qualify EEG surface electrodes as AP acc 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N ECoG/EEG electrodes - Required IEC 60601-1 tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N USB Powered EEG - USB provides 1 MOPP want 2 MOPP IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Is an EEG (Electroencephalogram) an IVD or a Medical Device? Other US Medical Device Regulations 14
A 60601-1 : Integrated Dry ECG Electrodes = 2 Patient connections inside 1 applied part? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A 60601-1-2 : Integrated Dry Electrodes = PATIENT COUPLED PORT? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D Defibrillation protection for patient connection except for electrodes IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
S IEC 60601-2-25 ECG 4 electrodes and derived leads IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
F Standards for testing of adhesive electrodes for use with TENS machines IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M Electrosurgical Electrodes - Stability Test Other Medical Device Related Standards 1
M Are disposable electrodes in the scope of WEEE Directive? Other ISO and International Standards and European Regulations 2
B Applications required for main accessories (electrodes) of CE Medical Products EU Medical Device Regulations 4
N Must electrodes for Rx/prescription stimulators be Rx only? Other US Medical Device Regulations 2
T Do Disposable Electrodes need CT-Compatible Testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J IEC 60601-1 for TENS-electrodes (CE-marking) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
J Sourcing Rubber or Silicone Electrodes for Medical Device Other ISO and International Standards and European Regulations 2
A Classification of ECG/EKG Electrodes according to MDD 93/42/EEC? EU Medical Device Regulations 2
D ECG Electrodes - Duration of Use (How long they are expected to stay on the patient) Other US Medical Device Regulations 2
D Independent Testing of ECG Electrodes in accordance with AAMI standards Reliability Analysis - Predictions, Testing and Standards 5
N CE Mark (CE directives) for Welding Electrodes CE Marking (Conformité Européene) / CB Scheme 5
bio_subbu FDA Issues Guidance Document on Electrocardiograph (ECG) Electrodes Other US Medical Device Regulations 0
J Pull test standard? Wires coming off the electrodes at the weld ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
V What can be considered as a Hazard ISO 14971 - Medical Device Risk Management 4
T What should be considered or asked to certified body auditors before selecting them for AS9100 Audit? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
B Under MDR, are VATS and MICS still considered just class Ir? ISO 13485:2016 - Medical Device Quality Management Systems 1
M Are we still considered a distributor in this scenario? EU Medical Device Regulations 7
Q What is considered "Multimedia equipment" (60601-1-2 section 7.1.3) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
G What is considered good (and practical) evidence of supervision ISO 17025 related Discussions 4
L Automatically considered Critical Component when within a fire ENCLOSURE? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
S Is Amalgam Separator considered a Medical Device in US and EU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
A Are defective returned spare parts to be considered as a complaint? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
J Are items with a NSN considered COTS Items? Manufacturing and Related Processes 2
GStough Objective Evidence - Are Interviews Still Considered as Objective Evidence? General Auditing Discussions 15
B A.I. diagnostic software is considered as medical device in FDA? US Food and Drug Administration (FDA) 6
Y Software updates considered servicing (7.5.4) ISO 13485:2016 - Medical Device Quality Management Systems 4
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
D CB and customer audits considered as internal audits? General Auditing Discussions 9
U NOC - What is considered a "design change" EU Medical Device Regulations 5
S What is considered a "core algorithm"? (From an FDA guidance document) Medical Information Technology, Medical Software and Health Informatics 4
MrTetris Should potential bugs be considered in software risk analysis? ISO 14971 - Medical Device Risk Management 5
D CE Marked product considered a Drug in India Other Medical Device Regulations World-Wide 0
J Are DAM tools, Mobile application, clinical dashboard, etc. considered accessories to medical device or supportive functions? Manufacturing and Related Processes 4
S Is any dissatisfaction over a Medical Device considered as a complaint? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 8
I Is highlighting on a printed document considered a change? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
M Are mortuary/autopsy tables considered to be medical devices EU Medical Device Regulations 7
K AS9100D Clause 7.5.2.a) - What is considered to be "documented information"? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
V Which batches should or could be considered for design validation and design verification? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
M Professional Use Medical Software French Labeling for Canada -- Not Considered Medical Device Canada Medical Device Regulations 2
S Tools and equipment provided by customer - Considered as external provider? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
S What is considered the complete software medical device? Medical Information Technology, Medical Software and Health Informatics 6

Similar threads

Top Bottom