Are power distribution units Multiple Socket Outlets (MSOs)?
Good day,
I am new here on the forum and in the world of IEC 60601!
We are currently designing a medical system which is a medical imaging scanner. The system is composed of the actual scanner, a power supply and a 19in rack with some non medical equipment such as network switches and a power distribution unit (PDU). The PDU has a 3 phase 120/208 VAC input and multiple standard 120VAC outlets to connect the network switches.
My question is whether a PDU should be regarded as a MSO (Multiple socket outlet). I believe the answer is yes. But according to 16.9.2.1 c), the MSO must comply with IEC 60884-1. My problem is I cannot find any PDU manufacturer that explicitly says it complies with this standard.
Has anyone had any experience with PDUs in a medical equipment/system?
Thank you!
Olivier
Peter Selvey
Are the outlets integrated in your device? (e.g. buying an outlet with bare terminals as a electrical part which is then screwed in the wall of your device, terminals connected to internally to your wiring by solder/screw etc). If it is, then the it's best to follow the requirements in the standard.
If it's not, the standard has a mistake in attempting to apply new requirements in 16.9.2.1 c) to an already complete device that is placed on the market by another manufacturer. Legally and logistically this is not possible.
The MSO should comply with the normal legal requirements in the country of sale (or use), and the manufacturer of the MSO should be the one responsible for that. The medical device manufacturer should not modify the MSO in any significant way, just make sure that it is used within ratings/specifications.
The medical device manufacturer can still add labels, use isolation transformer or prevent access etc as mentioned in the standard in order to manage the special risks in a medical environment, such as cumulative earth leakage.
R Medical Equipment (ME) used in different Power Distribution Systems IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
A Can a power Supply be an accessory to a medical device, if it is an 'off-the-shelf' product. IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Judy Abbott The impact of laser power on biocompatibility of the products Manufacturing and Related Processes 6
C Power Supply EN60601-1 vs EN62368-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S DC-AC Power inverter with two MOPPs IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
J Surge Protectors and other power protection Manufacturing and Related Processes 7
Sidney Vianna Informational APQP4Wind - Advanced Product Quality Planning for the Wind Power Supply Chain APQP and PPAP 3
U Power supply - Change From 160 to 300 Watts - Significant change? EU Medical Device Regulations 2
M MS Power Automate/Sharepoint Files Document Control Systems, Procedures, Forms and Templates 1
L Medical-grade power adapter labeling Other US Medical Device Regulations 2
A Calibration of Power Supplies AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
B IEC 60601 Cl. 8.4.3 - ME EQUIPMENT intended to be connected to a power source by a plug / Plug discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Proud Liberal Power Query to combine multiple tables of unequal number of columns Statistical Analysis Tools, Techniques and SPC 1
J Universal Power Cord with Adapters - IVD IEC 61010-2-101 CE Marking (Conformité Européene) / CB Scheme 6
L External power supplies: How close does the safety report have to match the end-use application? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
L IEC 62368-1 / Electrical Energy Source vs. Power Source? CE Marking (Conformité Européene) / CB Scheme 2
B Need For BIS Standard Mark? Imported OEM's Power supply,Li-Ion Battery Other Medical Device Regulations World-Wide 0
I Containment procedure for parts after a power outage. IATF 16949 - Automotive Quality Systems Standard 1
P UDI labeling for power module IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
E IEC 61010-1 Dielectric strength test of inner power supply Other ISO and International Standards and European Regulations 0
S Power rating for medical devices in the US - Radiotherapy devices US Food and Drug Administration (FDA) 0
T Is anyone working with N299.1 (Supply/service to nuclear power plants)? Various Other Specifications, Standards, and related Requirements 0
E IEC 60335-1 power input deviation - Table 1 Values Other ISO and International Standards and European Regulations 0
H External power supply - requirements in safety/EMC test? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K Power point training slides on DOE (Design of Experiments) Wanted Training - Internal, External, Online and Distance Learning 2
Marc Renewables, led by wind, provided more power than coal in Germany in 2018 Sustainability, Green Initiatives and Ecology 2
S Responsibilities between government owner and private subcontractor of a power plant seeking ISO 9001 certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
C Electrical Power Plant Supply failure in PFMEAs FMEA and Control Plans 10
H IEC 60601-1 Main Power Cord Requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Infusion Pump Power Failure Alarm IEC 60601-2-24 IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
F Power Cord Length Issue - Medical Device Electrical Safety IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
C ISO 9001:2015 - Non Applicable Design and Development (Power Plant Operator) Design and Development of Products and Processes 2
H Enclosed vs. Open-Frame Power Supplies - Overdoing it? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
J Pneumatic & Electrical Torque Tool - Calibration/Verification of 'Power Tools' General Measurement Device and Calibration Topics 2
G Medical Device Power Inlet/Outlet Voltage Labeling IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
x-files [Thermal Power plant KOSTOLAC, SERBIA] Sucessfully certified to ISO 50001 Sustainability, Green Initiatives and Ecology 1
M Is IEC 60601 compliance possible when only generic power-supply is specified? IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
U IEC 60601-1, Class 1 or Class 2 on emergency power system IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
A Power Traceability certificate for values for voltage, current and phase General Measurement Device and Calibration Topics 1
D NQA-1 Standard Applied to Services rather than Items for Nuclear Power Various Other Specifications, Standards, and related Requirements 3
R USB power between Medical Device Data Collection Device Component IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
F Power and sample size for factorial design Using Minitab Software 4
P Medical Device Marking - Power Input Ports IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
I Reaction plan in case of power outage Misc. Quality Assurance and Business Systems Related Topics 7
B Detachable Power Supply Cable Connection ESD Clarification IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
K How to determine rated input power for the markings? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
R Selling of Medical Power Supply in Australia/New Zealand and Japan Other Medical Device Regulations World-Wide 5
M Multiple versions of device due to power connectors? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2