Hi,
I'd like to better understand the Clinical Investigation mandate under Art. 61. In my reading - I understand that equivalence to another manufacturers device is likely impossible. However, where a manufacturer claims equivalence to it's own devices (Class III) this is an acceptable path to go the clinical eval route.
We are developing a newer version of a device currently approved under 93/42/EEC. We intend to go under the MDR with this newer version - and believe we'd be able to prove equivalence from a biological, technical and clinical route. Conducting an animal study and biocomp testing.
For a new Class III device (not novel) - how does Art. 61 (a) apply? Is this saying that we can't go the clinical eval route since the new version of the device is not an MDD device?
Art. 61(a):
6. The requirement to perform clinical investigations pursuant to paragraph 4 shall not apply to implantable devices
and class III devices:
(a) which have been lawfully placed on the market or put into service in accordance with Directive 90/385/EEC or
Directive 93/42/EEC and for which the clinical evaluation:
— is based on sufficient clinical data, and
— is in compliance with the relevant product-specific CS for the clinical evaluation of that kind of device, where
such a CS is available
Any guidance would be helpful!!
I'd like to better understand the Clinical Investigation mandate under Art. 61. In my reading - I understand that equivalence to another manufacturers device is likely impossible. However, where a manufacturer claims equivalence to it's own devices (Class III) this is an acceptable path to go the clinical eval route.
We are developing a newer version of a device currently approved under 93/42/EEC. We intend to go under the MDR with this newer version - and believe we'd be able to prove equivalence from a biological, technical and clinical route. Conducting an animal study and biocomp testing.
For a new Class III device (not novel) - how does Art. 61 (a) apply? Is this saying that we can't go the clinical eval route since the new version of the device is not an MDD device?
Art. 61(a):
6. The requirement to perform clinical investigations pursuant to paragraph 4 shall not apply to implantable devices
and class III devices:
(a) which have been lawfully placed on the market or put into service in accordance with Directive 90/385/EEC or
Directive 93/42/EEC and for which the clinical evaluation:
— is based on sufficient clinical data, and
— is in compliance with the relevant product-specific CS for the clinical evaluation of that kind of device, where
such a CS is available
Any guidance would be helpful!!