"As Found" (aka "As Received") Calibration Data - ISO 13485 Requirements

M

MEDQA

Hello -

When sending instruments / equipment out for calibration, do you request "as found" data?

If not, how do you satisfy the following 13485 requirement?

In addition, the organization shall assess and record the validity of the previous measuring results when the equipment is found not to conform to requirements.

Just wondering if I am missing something simple here..

thanks!
 
P

Phil Fields

Re: "As Found" Calibration Data - ISO 13485 Requirements

We get information as "found within specification" or "found not within specification". When "found not within spedcification" data is provided by the calibration service. The data is then evalauted to determine what effect the maeasuremnts may have on the final product.
 

Marc

Fully vaccinated are you?
Leader
Re: "As Found" Calibration Data - ISO 13485 Requirements

"As Found" data should always be required. If you calibrate a measurement device and "As Found" data shows the instrument was out of calibration, all measurements made since the last calibration are suspect. I can't speak to ISO 13485 requirements specifically, but this is standard in calibration.

What to do when a measurement device is found to be out of calibration has been discussed here a number of times, and how far you go depends upon the product and measurement being made, but the bottom line is you essentially have to go back and determine whether it is critical enough for you to recall all product which was measured with the measurement device. This is where traceability comes into play.

I'm sure some others here will add to my brief reply.
 
M

MEDQA

Re: "As Found" Calibration Data - ISO 13485 Requirements

I agree with you entirely. -thanks
 
A

arios

Re: "As Found" Calibration Data - ISO 13485 Requirements

" I can't speak to ISO 13485 requirements specifically, but this is standard in calibration.

ISO 13485 has he same requirement as in ISO 9001, except that 13485 requires a documented procedure.

The "as found" condition is called by some "As received condition". The standard does not say that you have to require it literally, but this information is absolutely necessary, otherwise you will not be able to assess the status of your instrument prior to recalibration. Bottom line, ask for it.

Don't be satisfatied only with an statement on the calibration certifcate saying that the instrument was found "in tolerance". It is important to check at the actual data on the report. I have seen for example calibration certs for pin gage sets where the calibration lab says that the "set" was in tolerance, but there are some pin gages out of range in the box and noted as such on the detail report.
 

BradM

Leader
Admin
Re: "As Found" Calibration Data - ISO 13485 Requirements

"As Found" data should always be required. If you calibrate a measurement device and "As Found" data shows the instrument was out of calibration, all measurements made since the last calibration are suspect. I can't speak to ISO 13485 requirements specifically, but this is standard in calibration.

What to do when a measurement device is found to be out of calibration has been discussed here a number of times, and how far you go depends upon the product and measurement being made, but the bottom line is you essentially have to go back and determine whether it is critical enough for you to recall all product which was measured with the measurement device. This is where traceability comes into play.

I'm sure some others here will add to my brief reply.

:agree1: Yep. Pretty much covered it, Marc. Nice reply.

Additionally from Section 7.6:

The organization shall take appropriate action on the equipment and any product affected.


I would argue that taking appropriate action would depend on the magnitude of the error. If I have 4 to 1 ratio, and I have a thermometer with a .25 degree tolerance. Being out of specification with .36 degree error is totally different than being out of specification by .92 degree error; as far as investigation/remedial action.


Too, if I get the report with the .92 degree error, something is wrong with the measurement system.

Even if the instrument does not fail, having the values let's you know if you are at the optimal calibration interval, by how much error is found at calibration.

Now... NOTE: I believe (don't quote me) that 17025 requires the calibration supplier to provide the data should the instrument be found out of specification. But, that is certainly something to arrange with your calibration vendor. So if the vendor provides data if it is out of tolerance, then there is no need to provide as-found data with every calibration.:)
 

Marc

Fully vaccinated are you?
Leader
Re: "As Found" Calibration Data - ISO 13485 Requirements

By the way - This is an aspect many people do not take into account when setting calibration interval (aka 'frequency') for a measurement device. You may have a stable measurement device which is always found to be in calibration, but in some cases calibration frequency (aka 'interval') is more important with respect to the potential costs of a recall, not to mention liability aspects. I'm sure in the medical device industry this is a more critical factor than in many other industries.
 

Marc

Fully vaccinated are you?
Leader
Re: "As Found" Calibration Data - ISO 13485 Requirements

<snip> Now... NOTE: I believe (don't quote me) that 17025 requires the calibration supplier to provide the data should the instrument be found out of specification. But, that is certainly something to arrange with your calibration vendor. So if the vendor provides data if it is out of tolerance, then there is no need to provide as-found data with every calibration.:)
My training (high reliability/mission critical military aerospace and marine electronic devices) was you always required "As Found" ("As Received") data, even if the device is found within calibration specs, to ensure you could track any drift or trend of the instrument. If the device needs any adjustment, it has to be watched. This is in part useful to watch for wear and related factors, as much as for use in determining calibration interval. It also goes to knowing the stability of the instrument over time.
 

BradM

Leader
Admin
Re: "As Found" Calibration Data - ISO 13485 Requirements

My training (high reliability/mission critical military aerospace and marine electronic devices) was you always required "As Found" ("As Received") data, even if the device is found within calibration specs, to ensure you could track any drift or trend of the instrument. If the device needs any adjustment, it has to be watched. This is in part useful to watch for wear and related factors, as much as for use in determining calibration interval. It also goes to knowing the stability of the instrument over time.

We're in agreement on this one.:agree1:

Which illustrates the difference of having instruments calibrated, and managing your calibration program. As with any process, you need data to make effective management decisions; no different with calibrations.
 
Thread starter Similar threads Forum Replies Date
P No trouble found (NTF) IATF 16949 - Automotive Quality Systems Standard 6
Vader22 As-Found FAILED calibration ISO 14001:2015 Specific Discussions 16
M Test failure Root cause not found Customer Complaints 13
M Vigilance - incident found in an article EU Medical Device Regulations 4
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
A Non-Conformances Found After 3rd Party Sorting Supplier Quality Assurance and other Supplier Issues 12
S ISO 9001 7.1.5.2 - Actions Taken When Measuring Equipment is Found to be Unfit for Use General Measurement Device and Calibration Topics 13
G Correction fluid found on documents that are reviewed Qualification and Validation (including 21 CFR Part 11) 11
Ed Panek ISOTECH TTI-10 found out of calibration - Calibration house was wrong? General Measurement Device and Calibration Topics 7
supadrai Just found out the hard way that only one registration per manufacturer is allowed in Yemen, regardless of OEM or private label arrangement. Other Medical Device Regulations World-Wide 0
supadrai Informational Japan Minor Change Notification Application Form - Found it? Japan Medical Device Regulations 0
M Can we change the ProCode if a better one is found? Other US Medical Device Regulations 3
GStough Pushback on a Nonconformance Found in a Supplier Audit General Auditing Discussions 22
S Recapture Program for Measurement Devices found Out of Calibration General Measurement Device and Calibration Topics 8
GStough Nonconformance Found Outside the Audit Scope - Supplier Audit General Auditing Discussions 16
T Are internally found Medical Device Software "Bugs" Complaints? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
A Defect Rate for asking for an RMA - Company wants an RMA for every defect found Supplier Quality Assurance and other Supplier Issues 10
Stijloor Inmate-Produced Combat Helmets Found Defective Supplier Quality Assurance and other Supplier Issues 4
C Calibration Certificate "As Found" condition - TS16949 Section 7.6.2 General Measurement Device and Calibration Topics 7
T What to do if parameters are found to be Outside of Control Limits Document Control Systems, Procedures, Forms and Templates 5
A Does anybody know in which Standard can this attached Alarm Symbol be found? Other Medical Device Related Standards 3
B Nonconforming Parts found by our Service Department Nonconformance and Corrective Action 3
D As Found Condition - In or out of tolerance? General Measurement Device and Calibration Topics 4
J Customer found Nonconforming Product Metrics ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
R Internal Audit Cancellation when Nonconformances are Found Quality Manager and Management Related Issues 13
D "As Found" Condition and Calibration Interval Changes Calibration Frequency (Interval) 5
N What Would You Do If You Found Treasure - But it's not yours... Coffee Break and Water Cooler Discussions 17
D What is "As Found" in Calibration of a Measurement Device? General Measurement Device and Calibration Topics 8
R "As Found" Calibration Results for HPLCs General Measurement Device and Calibration Topics 22
Q Auditor found Form with Identical Contents but different Format - is this an NC? Document Control Systems, Procedures, Forms and Templates 28
D What should the QMS Auditor do if a Major Nonconformity is found during an Audit? Internal Auditing 11
C National/International Standard For Calibration "As Found" Readings General Measurement Device and Calibration Topics 4
Marc So, I woke up this morning and found this tree growing... Coffee Break and Water Cooler Discussions 4
Y What if the Root Cause can't be found? Nonconformance and Corrective Action 6
R Information found missing in a Medical Device Record Other US Medical Device Regulations 4
Hami812 Predicting the outcome using high No Trouble Found Rate Statistical Analysis Tools, Techniques and SPC 4
A Help needed for Fighting a specifically identified Organism found during Dose Audit Miscellaneous Environmental Standards and EMS Related Discussions 2
R Suppliers Defects found during Production Process - Supplier Scorecards Quality Tools, Improvement and Analysis 6
B CMM Stylus Calibration - Required to record the As Found condition AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
S Want to buy Pinhole Diaphragm(S) per ASTM E1165-04 - Found it! Various Other Specifications, Standards, and related Requirements 3
Q Torque Wrench Issue - Incorrect method to collect the "as found data" Problem Solving, Root Cause Fault and Failure Analysis 7
S Top 10 areas where most NC's are found during CB audits? General Auditing Discussions 4
S External Auditor Findings when an Internal Audit found a Nonconformance Yesterday Document Control Systems, Procedures, Forms and Templates 11
D Found Watch - Watch is "Element" and says "New York" Coffee Break and Water Cooler Discussions 16
S What is the proper flow to feedback the defective found during QA sampling check? Quality Tools, Improvement and Analysis 8
Marc Tar balls found along shore in Key West - 18 May 2010 World News 6
H Elsmar Cove Statistics Found Online Coffee Break and Water Cooler Discussions 2
J Positive Corrective Action - 3 parts found out of 20,000 pcs 9 months after delivery Inspection, Prints (Drawings), Testing, Sampling and Related Topics 33
L No Trouble Found (NTF) concept applied in Automotive Industry Nonconformance and Corrective Action 6
A Injection Mold 100% Visual Inspection - Short Shots and Debris not being found Inspection, Prints (Drawings), Testing, Sampling and Related Topics 22

Similar threads

Top Bottom