As Found Condition - In or out of tolerance?
dwperron
We have been having a philosophical discussion here. When an instrument arrives at the lab and it is inoperative (dead power supply, bad display, hangs on self test, etc) so that you are unable to record "as found" results, and you repair the instrument, and find that after the repair it is within tolerance without having to make any adjustments. Would you report the as-found condition to be in or out of tolerance?
I have heard that since you could not have taken an out of tolerance measurement because the instrument was inoperative that you should not report it as found out of tolerance (mainly so that the user does not have to deal with an "out of tolerance" result).
The other side feels that since you could not make a measurement how can you call the instrument in tolerance?
How do your organizations handle this issue? Is there any guidance in the standards?
BradM
Hmmm.... that's a good question.
I would document as an as-found as "fail"; if you cannot assert the instrument is in tolerance at the time you receive it. Mark on the documentation what occurred, the nature of the repair, etc.; make the repair, and then record as-left. I would note: inoperable; detectable to the user. Then, the user can write up in the failure notice follow-up that it was not used, no impact or whatever.
After all, the instrument could have been working fine prior to shipping; or they could have had intermittent issues affecting their process.
Overall, I think it's... better.... for a repair/ metrology function to report on what they found and what they did; and not have to try to assess if there was impact or not.
Hershal
Metrologist-Auditor
I would agree with John. Especially as repair is typically billed at a different reate than calibration.
DietCokeofEvil
At my facility, if something is not operable upon receipt, the as found readings are "inoperable". If it can be repaired, the repaired and calibrated readings go in the as left column.
C Determination of M&TE out of tolerance condition and what to do when found General Measurement Device and Calibration Topics 6
C Calibration Certificate "As Found" condition - TS16949 Section 7.6.2 General Measurement Device and Calibration Topics 7
D "As Found" Condition and Calibration Interval Changes Calibration Frequency (Interval) 5
B CMM Stylus Calibration - Required to record the As Found condition AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
M Test failure Root cause not found Customer Complaints 10
M Vigilance - incident found in an article EU Medical Device Regulations 3
Y We found out we have been using a equipment without validation for past 4 years Quality Manager and Management Related Issues 6
A Non-Conformances Found After 3rd Party Sorting Supplier Quality Assurance and other Supplier Issues 12
S ISO 9001 7.1.5.2 - Actions Taken When Measuring Equipment is Found to be Unfit for Use General Measurement Device and Calibration Topics 13
G Correction fluid found on documents that are reviewed Qualification and Validation (including 21 CFR Part 11) 11
Ed Panek ISOTECH TTI-10 found out of calibration - Calibration house was wrong? General Measurement Device and Calibration Topics 7
supadrai Just found out the hard way that only one registration per manufacturer is allowed in Yemen, regardless of OEM or private label arrangement. Other Medical Device Regulations World-Wide 0
supadrai Informational Japan Minor Change Notification Application Form - Found it? Japan Medical Device Regulations 0
M Can we change the ProCode if a better one is found? Other US Medical Device Regulations 3
GStough Pushback on a Nonconformance Found in a Supplier Audit General Auditing Discussions 22
S Recapture Program for Measurement Devices found Out of Calibration General Measurement Device and Calibration Topics 8
GStough Nonconformance Found Outside the Audit Scope - Supplier Audit General Auditing Discussions 16
T Are internally found Medical Device Software "Bugs" Complaints? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
A Defect Rate for asking for an RMA - Company wants an RMA for every defect found Supplier Quality Assurance and other Supplier Issues 10
Stijloor Inmate-Produced Combat Helmets Found Defective Supplier Quality Assurance and other Supplier Issues 4
T What to do if parameters are found to be Outside of Control Limits Document Control Systems, Procedures, Forms and Templates 5
A Does anybody know in which Standard can this attached Alarm Symbol be found? Other Medical Device Related Standards 3
B Nonconforming Parts found by our Service Department Nonconformance and Corrective Action 3
J Customer found Nonconforming Product Metrics ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
R Internal Audit Cancellation when Nonconformances are Found Quality Manager and Management Related Issues 13
N What Would You Do If You Found Treasure - But it's not yours... Coffee Break and Water Cooler Discussions 17
D What is "As Found" in Calibration of a Measurement Device? General Measurement Device and Calibration Topics 8
R "As Found" Calibration Results for HPLCs General Measurement Device and Calibration Topics 22
Q Auditor found Form with Identical Contents but different Format - is this an NC? Document Control Systems, Procedures, Forms and Templates 28
D What should the QMS Auditor do if a Major Nonconformity is found during an Audit? Internal Auditing 11
C National/International Standard For Calibration "As Found" Readings General Measurement Device and Calibration Topics 4
Marc So, I woke up this morning and found this tree growing... Coffee Break and Water Cooler Discussions 4
Y What if the Root Cause can't be found? Nonconformance and Corrective Action 6
R Information found missing in a Medical Device Record Other US Medical Device Regulations 4
Hami812 Predicting the outcome using high No Trouble Found Rate Statistical Analysis Tools, Techniques and SPC 4
A Help needed for Fighting a specifically identified Organism found during Dose Audit Miscellaneous Environmental Standards and EMS Related Discussions 2
R Suppliers Defects found during Production Process - Supplier Scorecards Quality Tools, Improvement and Analysis 6
S Want to buy Pinhole Diaphragm(S) per ASTM E1165-04 - Found it! Various Other Specifications, Standards, and related Requirements 3
Q Torque Wrench Issue - Incorrect method to collect the "as found data" Problem Solving, Root Cause Fault and Failure Analysis 7
S Top 10 areas where most NC's are found during CB audits? General Auditing Discussions 4
S External Auditor Findings when an Internal Audit found a Nonconformance Yesterday Document Control Systems, Procedures, Forms and Templates 11
D Found Watch - Watch is "Element" and says "New York" Coffee Break and Water Cooler Discussions 16
M "As Found" (aka "As Received") Calibration Data - ISO 13485 Requirements ISO 13485:2016 - Medical Device Quality Management Systems 8
S What is the proper flow to feedback the defective found during QA sampling check? Quality Tools, Improvement and Analysis 8
Marc Tar balls found along shore in Key West - 18 May 2010 World News 6
H Elsmar Cove Statistics Found Online Coffee Break and Water Cooler Discussions 2
J Positive Corrective Action - 3 parts found out of 20,000 pcs 9 months after delivery Inspection, Prints (Drawings), Testing, Sampling and Related Topics 33
L No Trouble Found (NTF) concept applied in Automotive Industry Nonconformance and Corrective Action 6
A Injection Mold 100% Visual Inspection - Short Shots and Debris not being found Inspection, Prints (Drawings), Testing, Sampling and Related Topics 22
T In-House Calibration - Digital Micrometer found Out of Calibration General Measurement Device and Calibration Topics 10