For years we have been testing our pouches using Seal Strength testing according to ASTM F88 and we always used multiple samples per pouch during packaging validation testing:
- the left seal, right seal, top seal which we all made by the manufacturer of the pouch (so 3 samples in total)
- the bottom seal which we made after placing our product in the seal (another 1 sample)
However, our testing partner now suggest that during packaging validation testing only
- one samples of the manufacturer seal (arbitrarily taken from the left seal, right seal or top seal)
- one sample of the bottom seal
My reaction was to disagree but looking into ASTM F88, ISO 11607-1 & ISO 11607-2 I could not find a requirement on this.
What is your experience with this? Did I miss the requirement somewhere? Or is it just a 'best practice'?