Comments, folks?
----snippo----------
Date: Mon, 12 Jun 2000 10:49:57 -0400
From: Bruno Brazauskas
To: Greg Gogates
Subject: Re: GAGE STABILITY AND LINEARITY RE4
Per the previously recommended AIAG manual:
The data is the result of measurements taken with the gage being evaluated, utilizing the appropriate gaging instructions. The samples, which are measured, may be either production parts or masters. Collect the data in a manner which is similar to, but in terms of repetition not quite the same as a gage R&R study.
For Linearity, the samples should be taken from the expected operating range of the gage. The bias on each of these parts is determined to be the difference between the reference value and the observed average measurement. The reference value for each part is determined through a different means. These means may be plate or other inspection machine layout or through an independent or calibration lab.
For Stability, it is not necessary to have parts from the expected operating range of the gage but it is still a good idea. As the AIAG MSA states, sample size and the frequency of the measurements should be based on your knowledge of the measurement system. So you may use the same parts as you used for the Linearity Evaluation. For this study, the measurements are taken over time (i.e.. the same parts are measured again and again over a predetermined time).
Bruno Brazauskas
Inspection Technologies, Inc.
Greg Gogates wrote:
> Date: Fri, 09 Jun 2000 09:05:37 -0400
> Subject: Re:GAGE STABILITY AND LINEARITY RE3
>
> Forgive me but I'm still confused on the data collection. Is this data from
> actual production part measurement, calibration data, Gage R&R data or how do I
> obtain the data? I guess I'm a little dumb but I'm not sure what data to use for
> these studies.
>
> ____________________Reply Separator____________________
> Subject: GAGE STABILITY AND LINEARITY RE3
> Author: Greg Gogates
> Date: 6/5/00 11:38 AM
>
> Date: Mon, 05 Jun 2000 08:44:18 -0400
> From: "Bruno J. Brazauskas"
> To: Greg Gogates
> Subject: Re: GAGE STABILITY AND LINEARITY
>
> These types of studies may be simple or more involved. The simple
> versions are as follows and reference GM's Statistical Process Control
> Manual:
>
> Basically, "Gage Stability" is the difference (range) in the average of
> a minimum of two sets of measurements taken with the same gage on the
> same samples at different times. For example, using gage "A" and sample
> "1" take 10 readings today and average them. Repeat this test tomorrow
> under the same conditions. The difference between today's and tomorrow's
> averages is the stability of gage "A" over one day.
>
> "Gage Linearity" is the range of values derived from at least groups of
> data. It is a little more complicated and requires that the actual
> values of samples to be tested are known. A known sample at the low end
> of the specification is measured with the gage and the difference
> between actual and measured readings is recorded. This is repeated with
> a sample from the high end of the specification. The range obtained
> from these deviations is the gage linearity.
>
> A more involved approach may be found within the AIAG's "Measurement
> Systems Analysis (MSA)." It includes more information in the form of
> "Guidelines for Determining Linearity" and "Guidelines for Determining
> Stability."
>
> Bruno Brazauskas
> Inspection Technologies, Inc.
>
> Greg Gogates wrote:
> >
> > Date: Wed, 31 May 2000 13:34:59 -0400
> > Subject: GAGE STABILITY AND LINEARITY
> >
> > I have been requested by our QS9000 registrar to begin performing Gage
> > Stablility and Linearity Studies. I have performed Gage R&R studies but never
> > linearity or stability studies. Can anyone give me some detailed information
> on
> > what type of data collection is needed to perform these studies and how to
> > perform these studies? I have software (Gagetrak 4.0) for analyzing the data
> but
> > it doesn't really specify exactly how and what type of data is needed.
> > Thank you.