Automated Measurement Machine MSA - Small Parts

A

arthurs

#1
We're thinking of purchasing a new tool that automatically measures small parts. When I say automatically I really mean that the Operator is taken out of the equation. The only input the operator has is to dump the parts in a bin. The machine does the rest. I can perform R&R study but the operator portion should contribute 0% as it make no odds to the machine who loads it. That said, I expect the R&R to show any run to run variation.

What else can I do wrt consistency trial or is an R&R good enough.
 
Elsmar Forum Sponsor

Proud Liberal

Quite Involved in Discussions
#2
I just bought a $50k vision system that claimed parts could be randomly arranged on the stage and found R&R values far in excess of 30% when they were. The only way I could get the unit to repeat was to fixture the parts in a fixed orientation.

I guess what I'm saying is don't believe the salesman. Make the do the R&R for you before you commit.
 

Miner

Forum Moderator
Staff member
Admin
#3
We're thinking of purchasing a new tool that automatically measures small parts. When I say automatically I really mean that the Operator is taken out of the equation. The only input the operator has is to dump the parts in a bin. The machine does the rest. I can perform R&R study but the operator portion should contribute 0% as it make no odds to the machine who loads it. That said, I expect the R&R to show any run to run variation.

What else can I do wrt consistency trial or is an R&R good enough.
Replace the Operators field with run to run. When you interpret the Reproducibility remember that it represents the run to run instead of operator to operator.
 
A

arthurs

#4
Well, I'm thinking of doing some kind of Bias study and also correlation to a known good measurement. I'm not too clear on the criteria for success on a consistency study or method. I think that we should demonstrate that the tool is accurate to 10% of the process tolerance too.
 
Thread starter Similar threads Forum Replies Date
C MSA (Measurement System Analysis) on automated machine Capability, Accuracy and Stability - Processes, Machines, etc. 1
Ron Rompen MSA on automated measurement system - Multiple Step Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
T Automated and Manual Measurement System Comparison (Correlation) Quality Tools, Improvement and Analysis 3
C How to perform Gauge R & R on Automated Test & Measurement Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
E MSA (Measurement System Analysis) Study on an Automated Process Quality Tools, Improvement and Analysis 4
I Gage R&R for Automated Measurement Equipment Systems Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
B Automated Vision Inspection System MSA (Measurement Systems Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
J Automated In-Line Inspection MSA (Measurement System Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
B MSA (measurement system analysis) on CCA Automated Test Equipment (ATE) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
B Verifying Accuracy Automated Measurement Device - Width of a strip of steel Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
P Measurement Resolution of In-Process Automated Inspection System - Stamping House Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
gramps What do you think about automated QA testing For software app industry? Misc. Quality Assurance and Business Systems Related Topics 5
L Gage RandR on automated equipment. IATF 16949 - Automotive Quality Systems Standard 7
J Sub-supplier change from manual to automated process - same specs - Report to FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
P Images of Product for Automated Process - How long to keep? US Food and Drug Administration (FDA) 2
T Correct MSA study for an automated camera system which makes attribute inspection Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
M Applicability of Means of Protection, working voltage in an Automated External Defibrillator IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
T Tests IEC 60601-1 - Grounding Resistance, Leakage Current and Dielectric Rigidity in an automated way IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Ron Rompen GR&R on Automated Inspection Cell Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 15
S What are you looking for in an automated quality management system? Quality Assurance and Compliance Software Tools and Solutions 8
F Automated Platform (IVD) Classification/Registration (New to RA) EU Medical Device Regulations 3
M Is Automated Internal Auditing Possible? Internal Auditing 13
Crimpshrine13 MSA for a part Automated and part Manual Inspection Machine IATF 16949 - Automotive Quality Systems Standard 7
B How to do an MSA on an Automated Torque / Angle System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
A Automated calibration of Digital Multimeter without communication port Calibration and Metrology Software and Hardware 2
S Bench Marking Automated Data Collection Benchmarking 4
P Looking for recommendations for an automated vision measuring system Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
M Automated Revision Updates - Notifications of new revisions for ISO's to subscribers Document Control Systems, Procedures, Forms and Templates 2
T Tools Six Sigma for my Automated Machine Six Sigma 8
C Gage R&R on In-Process Automated Inspection Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 13
Q 21 CFR 820.30 - Automated with Computer Software - Applicable? US Food and Drug Administration (FDA) 5
S How to calibrate AOI (Automated Optical Inspection) for IOPQ Qualification and Validation (including 21 CFR Part 11) 3
P GR&R on automated test system which produces hundreds of results Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
C Multiple Variation Gage R&R for Automated Testers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
J Classes for Hydraulic Calibration - Dead Weight Tester or Automated Controller General Measurement Device and Calibration Topics 1
Q Automated Test Software and System Validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
L Automated Calculators like EXCEL sheets for different processes Quality Assurance and Compliance Software Tools and Solutions 2
L Dry Lube Application (Ever lube 620) - Manual Application vs. Automated Spray Process Manufacturing and Related Processes 1
Hami812 Gage R&R on Automated Tool that supports Pass/Fail and some Parametric Testing Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
G Automated Production Line Quandary... FMEA and Control Plans 1
R Use of Automated Liquid Handling Robot for a PCR based Assay CE Marking (Conformité Européene) / CB Scheme 2
D Automated Vision Systems - Cost, Value and Payback of Implementation Statistical Analysis Tools, Techniques and SPC 2
A Classification of Automated Gait Training Device for Rehabilitation EU Medical Device Regulations 5
R GAMP (Good Automated Manufacturing Practice) simplified please. Quality Manager and Management Related Issues 3
G Is Agilent?s automated network analyzer verification ISO/IEC-17025 compliant? ISO 17025 related Discussions 4
T Gage Repeatability for an Automated Leak Tester Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
sagai Automated System's Signature on Electronic Record - PART11 Qualification and Validation (including 21 CFR Part 11) 4
C Multiple Fixture Automated Vision System GR&R (Gage R&R) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M GR&R (Gage R&R) for Automated Electrical Testing Equipment Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
B Automated paint process control on conveyerized lines - Million dollar question Capability, Accuracy and Stability - Processes, Machines, etc. 14

Similar threads

Top Bottom