Avoiding an Ultrasound System HIPOT Test during Manufacture

P

Phloogy

#1
I am developing an ultrasound system and am looking to see if getting the power supply or the entire system UL certified will allow the device to skip HIPOT testing during manufacture. Is this a true statement? Is this a good idea? Any feedback is appreciated!
 
Elsmar Forum Sponsor
I

isoalchemist

#2
Just because the system is approved doesn't mean you manufactured it correctly and no danger exists. Every computer we made had HIPOT testing and the tester was one of the first things checked every time UL (our electrical assessor) was in.
 
P

PaulJSmith

#3
Correct. The UL certification only approves the design based on the performance of the submitted sample unit. HIPOT testing will confirm that there are no failures during the manufacturing process on each individual unit.

I'm not sure why you would want to avoid it. That's a very important test, and takes only a few seconds to perform (takes longer to connect it than to test it).
 
P

Phloogy

#4
Thanks for the feedback! I agree, there may be issues/mistakes during manufacturing where HIPOT would be beneficial. We are looking to be as lean as a possible and cut out the non-value added steps.

This is something we are only considering if our PFMEA supports it. Ex. the occurrence of relevant failure modes are low... So if we need HIPOT as mitigation, then we will have it.
 

somashekar

Staff member
Super Moderator
#5
Thanks for the feedback! I agree, there may be issues/mistakes during manufacturing where HIPOT would be beneficial. We are looking to be as lean as a possible and cut out the non-value added steps.

This is something we are only considering if our PFMEA supports it. Ex. the occurrence of relevant failure modes are low... So if we need HIPOT as mitigation, then we will have it.
Think of the risks along with the lean and non - value as you say of testing for safety.
If your product is going to be with a safety mark on it (Ex. UL mark) then the UL product file will contain what you are required to test on 100% of manufactured devices. Certainly it would contain the HIPOT, along with conditions of testing. The followup services inspection will be keen to see these records of devices made and dispatched.
 
#6
Remember that hipot (dielectric strength tests) only apply to solid insulations. Where insulation is provided by creepage and clearance distances it is evaluated by measurements.

Also, where the dielectric test of 60601-1 says that the high voltage should be maintained for 1 minute (with no breakdown) this is not suitable for a production test. Instead the high voltage should only be held for 5 seconds.

(The parts that I test for dielectric strength during a "Type Test" I consider to have invisibly damaged insulation and quarantine them from further use).

It is also worth realising that, particularly with modern power supplies and isolation systems, the value of a "front-to-back" dielectric test is very limited. When the patient isolation is comprised of several parts (so MOPPs are in more than one place) any high voltage applied from the power supply input to the applied parts will be (unevenly) divided across those insulations.

I now only do dielectric tests on component parts, although we do retain the test as a brief production test.
 
H

hcrook

#7
Additionally, IEC 62354 3.0 ed Annex K provides some guidance on production line hipot testing, such as having a 1 second hipot test duration at a higher voltage. 60601-1 also mentions in Subclause 5.1 that tests from this standard shouldn't necessarily be repeated, especially in regards to hipot. "Production tests need not be identical with TYPE TESTS, but can be adapted to manufacturing conditions and possibly invoking less RISK for the quality of the insulation or other characteristics important for BASIC SAFETY and ESSENTIAL PERFORMANCE."
 

Candi1024

Quite Involved in Discussions
#8
Also keep in mind that 100% testing your product, MAY keep certain manufacturing steps from being "special". Special meaning that the output is not 100% verified. If they are special processes, you need to do process qualification. This may or may not apply to you.
 
Thread starter Similar threads Forum Replies Date
O Informational Scaling back internal audits due to corona virus while avoiding a NC Internal Auditing 7
M Avoiding Pitfalls in ISO 9001:2015 Transition - Free webinar Training - Internal, External, Online and Distance Learning 0
L Avoiding Recurrence of the Supplier?s Quality Problems Supplier Quality Assurance and other Supplier Issues 5
B Avoiding Rust on 410 Stainless Steel Design and Development of Products and Processes 15
C Guidance for avoiding human error Quality Tools, Improvement and Analysis 24
Claes Gefvenberg ISO 9001: Avoiding Over Documentation Document Control Systems, Procedures, Forms and Templates 26
A Avoiding mixup - Similar parts for two different customers QS-9000 - American Automotive Manufacturers Standard 1
A Avoiding ISO 9000 Registration - Hiding behind 'Customer Wishes' ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
M Informational US FDA final guidance – Marketing Clearance of Diagnostic Ultrasound Systems and Transducers Medical Device and FDA Regulations and Standards News 0
S South Korea Ultrasound Software Requirements Other Medical Device Regulations World-Wide 1
J Level of Concern - Ultrasound Medical Device Systems ISO 14971 - Medical Device Risk Management 2
J Shipping Liquids (Small Bottles of Ultrasound Gel) to Europe EU Medical Device Regulations 4
T Canada Device Licence of Accessory - Ultrasound Therapy Devices Canada Medical Device Regulations 4
S Internal Files for Medical Device Ultrasound System Licence Other US Medical Device Regulations 6
S The relationship between IEC 60601-2-18 and IEC 60601-2-37 - Ultrasound Project IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L FDA issued new Guidance for Ultrasound Diagnostic Devices. 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T More on Biocompatible Ultrasound Gel ISO 13485:2016 - Medical Device Quality Management Systems 3
T Standard(s) used to Test Sterile Ultrasound Gel for use on Heart ISO 13485:2016 - Medical Device Quality Management Systems 4
S Searching for TOYOTA Floor Management Development System (FMDS) board Customer and Company Specific Requirements 3
M Procedure for vigilance system according to new MDR EU Medical Device Regulations 0
bobdoering Most typical universal Quality System Funny Stuff - Jokes and Humour 8
G Organizing internal audit program for an Integrated QHSE Management System Internal Auditing 13
K Bioreactor/ incubator system in tissue engineering EU Medical Device Regulations 9
D Integrated Management System Software Quality Manager and Management Related Issues 2
E System&Procedure pack or accessory EU Medical Device Regulations 0
J Informal vs formal scope creep... managing non-medical devices through system processes ISO 13485:2016 - Medical Device Quality Management Systems 2
C Quality Management System Question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 32
V Csv, excel format - audit trail file of HPLC system ( Empower, openlab, EZchrom or any other ) Qualification and Validation (including 21 CFR Part 11) 0
I Class VI certification on seals, o-ring and plastic for a WFI system Qualification and Validation (including 21 CFR Part 11) 0
A System for Supplier Documents Quality Assurance and Compliance Software Tools and Solutions 7
S MDR - System and procedure pack article 22 and all sub processes that apply ISO 13485:2016 - Medical Device Quality Management Systems 0
T Device & Accessory, Components in device, or System EU Medical Device Regulations 0
M Go Live With New ERP System before Recertification Audit General Auditing Discussions 6
John C. Abnet Terms- Different Items in a system ISO 26262 - Road vehicles – Functional safety 0
T Controlling Expandable Forms in Paper-Based Document Control System Document Control Systems, Procedures, Forms and Templates 10
G Management Review (integrated system) Management Review Meetings and related Processes 17
M Unique Quality Management System for 2 sites ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S Orthopedic Implants (Knee system) Sterilization Indicators Manufacturing and Related Processes 6
C Projects in the CAPA system Preventive Action and Continuous Improvement 6
E Insulation diagram for ECG monitoring and diagnosis system. IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
R Quality System Functional Safety Checklist / Guidance IATF 16949 - Automotive Quality Systems Standard 0
J Mastercontrol or other system ISO 13485:2016 - Medical Device Quality Management Systems 0
O Any info on release date of FDA “Computer Software Assurance for Manufacturing and Quality System Software” document? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
P MSA - what exactly mean "system" and master sample Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
P MSA study for visual system with artifical inteligence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
D Big companies suffer from quality management system? ISO 13485:2016 - Medical Device Quality Management Systems 1
qualprod Assign a name to a home-made ERP system? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
H IEC 60601-1 ME equipment or ME system IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K Biological indicator system classification under MDR EU Medical Device Regulations 2
N Chemical Testing on Medical Devices - Solutions in a container closure system (bag) EU Medical Device Regulations 1

Similar threads

Top Bottom