Barometer Calibration and Traceablilty

Marc

Fully vaccinated are you?
Staff member
Admin
#1
From: "Gust, Jeff" - gte.com
To: iso25
Subject: traceability of barometer
Date: Mon, 2 Aug 1999 12:49:00 -0500

Any barometer has the ability of becoming a traceable barometer, provided that you can develop an appropriate method of calibrating the device to the manufacturers (or other acceptable metrological) specifications.

Some barometers are very difficult to calibrate because they do not have a pressure port that is easily interfaceable to a pressure calibration system. If you beat the interface problem, the rest of the calibration process is relatively easy.

As an example, many barometers are built into a "clock" configuration. They are sealed, and there is no access to the pressure port. Our laboratory uses a bell Jar to calibrate these barometers. We place the whole barometer in the bell jar, and vary the pressure within the bell jar, which is monitored by our pressure calibration system (capacitance manometers or deadweight piston gage). We use this method also to check the barometer above atmospheric pressure, but we have to strap the bell jar down to its base because the increased pressure will lift the bell jar up until it leaks the excess pressure.

Please do not send your barometer out to get a one point calibration, which is simply a single measurement at the ambient pressure of the calibration lab.

Several manufacturers produce great barometers that can easily calibrated. The Davis Instruments Catalog lists several. I personally use the Druck DPI 740 in our electrical standards lab to monitor the barometric pressure as it applies to my standard resistance measurement process. It is easy to calibrate, and is very accurate.

Jeff Gust
Staff Engineer-Metrology
GTE Electronic Repair Services
NVLAP Lab Code 200352-0
 
Elsmar Forum Sponsor
Thread starter Similar threads Forum Replies Date
Mikey324 External calibration - Finding in our 3rd party audit General Measurement Device and Calibration Topics 9
L A certificate of conformance used to support a calibration General Measurement Device and Calibration Topics 6
A AS9100 - 7.1.5.2 In-House Calibration AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 25
B UV-A and Illuminance meter Calibration Setup General Measurement Device and Calibration Topics 0
S Calibration Related Posters General Measurement Device and Calibration Topics 2
F Quality Control in Electrical Calibration ISO 17025 related Discussions 0
C Narrow definition of IMTE or calibration possible? ISO 13485:2016 - Medical Device Quality Management Systems 8
dhuot1990 In-House Calibration Program - Biotech Industry General Measurement Device and Calibration Topics 3
M Reference part calibration or verification General Measurement Device and Calibration Topics 12
M Calibration of equipment IATF 16949 - Automotive Quality Systems Standard 4
Q Gage calibration / tracking software General Measurement Device and Calibration Topics 5
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
S Calibration Due date-Critical Connie General Measurement Device and Calibration Topics 16
Q Request Calibration Info Calibration and Metrology Software and Hardware 16
R Calibration Log - Archive Manufacturing and Related Processes 2
U Calculate acceptance of equipment reading after calibration General Measurement Device and Calibration Topics 2
D Question on Documented Calibration versus ISO 17025 Accredited Calibration ISO 13485:2016 - Medical Device Quality Management Systems 2
U Calculate acceptance of calibration reading Measurement Uncertainty (MU) 1
D Air hood calibration ISO 13485:2016 - Medical Device Quality Management Systems 1
K Ground Bond Test for Class I Medical Electrical Equipment - calibration problems IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
E Calibration/Verification Report Durometer ISO 17025 related Discussions 2
S Medical device equipment calibration Qualification and Validation (including 21 CFR Part 11) 1
J Scrap Material Scale Calibration Required? IATF 16949 - Automotive Quality Systems Standard 21
S 7.1.5.2.1 Calibration/Verification Records IATF 16949 - Automotive Quality Systems Standard 6
K Is Calibration Required for Non-Adjustable Commercial Inspection Devices? General Measurement Device and Calibration Topics 11
F Calibration in GPC/SEC General Measurement Device and Calibration Topics 1
Q Changing calibration frequency Calibration and Metrology Software and Hardware 18
M IATF 16949 SI # 10: Integrated self-calibration of measurement equipment- Needs for explanation IATF 16949 - Automotive Quality Systems Standard 2
M Storing Calibration Certificate Document Control Systems, Procedures, Forms and Templates 7
B Risk analysis for defective measuring or measuring equipment out of calibration General Measurement Device and Calibration Topics 2
S Placing a calibration sticker with someone other than your initials Records and Data - Quality, Legal and Other Evidence 8
J IATF 16949 Calibration/Verification records question ISO 26262 - Road vehicles – Functional safety 6
M Approval of Calibration Certificates General Measurement Device and Calibration Topics 1
lanley liao Question regarding the calibration of monitoring and measure equipment. Oil and Gas Industry Standards and Regulations 3
N IATF Calibration Lean in Manufacturing and Service Industries 5
Q Do these certificates of calibration meet ISO 9001 requirements for traceability to NIST? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
T Plug Gage Calibration Calibration and Metrology Software and Hardware 1
M Load Cell Calibration using a totalizer on a flow meter General Measurement Device and Calibration Topics 0
E Calibration Records needed ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
D Limited Range Calibration - 5000 lb Industrial floor scale General Measurement Device and Calibration Topics 3
D Calibration of Small Scales General Measurement Device and Calibration Topics 32
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
I IQOQ or just initial calibration required? General Measurement Device and Calibration Topics 3
B Calibration in real life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
J Calibration cycle for monitoring & measuring tools used in medical device manufacturing General Measurement Device and Calibration Topics 5
A Is calibration of test weight required General Measurement Device and Calibration Topics 4
S Calibration Frequency for Slip Gauge Kit used for CMM Calibration? General Measurement Device and Calibration Topics 0
S Calibration/Verification of customer fixtures IATF 16949 - Automotive Quality Systems Standard 6
Ron Rompen Calibration by manufacturer ISO 17025 related Discussions 4
Q Calibration verification records 7.1.5.2.1 IATF 16949 - Automotive Quality Systems Standard 2

Similar threads

Top Bottom