Barrier cream classification

MedicalDevice

Involved In Discussions
#1
Good morning everyone!

I came across a specific medical device product that is described as a "Barrier Cream"
Some indications that come with it are the following:
• Suitable for wet areas
• It helps maintain the pH of the skin and skin moisture.
This product is applied to intact and at-risk skin to provide protection from body fluids and moisture. It can be applied around wounds, wound dressings and incontinence pads where a protective barrier will help prevent skin breakdown.

Could someone help me to classify this medical device? From my point of view, it is a (Rule 4) Class I MD. But I have had a discussion with a friend that thinks it's Class II. Any feedback would be helpful.
 
Elsmar Forum Sponsor

Sullen-gent

Starting to get Involved
#2
Hi,

I worked with products previously that were barrier creams and they were class I. However would be careful as it very much depends on the claims and the nature of contact. Above you said 'intact' skin so could argue Rule 1 is appropriate (contacting intact skin). However you also mention wounds and this is more rule 4 territory where the skin is damaged or the dermis is breached or the product manages the wound environment.

This is all MDD and there are some changes under MDR to these rules but they don't look massively different on face value.

I think this cream is likely to be class I, as barrier creams act as a shield to stop healthy skin (rule 1) becoming damaged or protect damaged skin (rule 4) from further damage but it would be the claims of the device that could push it further into risk classification.
 

dgrainger

Trusted Information Resource
#3
However you also mention wounds and this is more rule 4 territory where the skin is damaged or the dermis is breached or the product manages the wound environment.
Yes, I agree, your claims may push the classification up - and it is foreseeable that patients may use it on wounds given your description.

There is an entry in the borderlines manual for zinc oxide creams that may/may not be useful:

4.6. Zinc oxide containing creams
- Background
Products containing zinc oxide are available as creams for local administration.
Some zinc oxide containing products, depending on their claims and intended use, might be covered by Directive 76/768/EEC on cosmetic products.
The discussion below only concerns zinc oxide containing products used to treat or prevent minor skin irritations (e.g. burns, cuts, nappy rash, eczema etc.).
- Outcome
A medical device should not achieve its principal intended action in or on the human body by pharmacological, immunological or metabolic means, but may be assisted in its function by such means.
For zinc oxide containing products, according to the literature, a pharmacological and metabolic action is demonstrated, e.g. may play a role in enzymatic processes, support of wound granulation.
The pharmacological action may, however, be ancillary when the product concerned is primarily a barrier cream.
In such cases the qualification of zinc-oxide containing products is defined taking into account the claims, the intended purpose and the relevant primary mode of action. Some products that act primarily as a barrier may therefore be acceptable as Class III medical devices in accordance with rule 13 of Annex IX of Directive 93/42/EEC.
In accordance with Article 2(2) of Directive 2001/83/EC, in case of doubt where taking into account all product characteristics, and provided that the concerned product meets both definitions of a medicinal product and of a medical device, the provisions of Directive 2001/83/EC shall apply.
 
Thread starter Similar threads Forum Replies Date
K ISO 15223-1 Sterile barrier system - The 3-layer symbol EU Medical Device Regulations 1
M GSPR 23.3 (h) Manufacturing date on sterile barrier EU Medical Device Regulations 0
M Barrier Free vs. Wheelchair Accessible - Differences Occupational Health & Safety Management Standards 1
O F-typed isolated barrier of medical electrical equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
V Language is not a barrier if you know the "thing/subject" and "art of communicating" Coffee Break and Water Cooler Discussions 8
A Minimum Testing Requirements for a Clear Protective Equipment Barrier (Class II 510K) Other US Medical Device Regulations 1
D Packaging Addition - Non-sterile barrier to protect the whole package 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 12
F Sterile Barrier Package (Tyvek Pouch) Validation Question ISO 13485:2016 - Medical Device Quality Management Systems 7
M Split labels? Split label approach on the sterile barrier packaging ISO 13485:2016 - Medical Device Quality Management Systems 4
N Insulation Barrier Testing - According IEC 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 22
S Sterile Barrier System (SBS) Testing ISO 13485:2016 - Medical Device Quality Management Systems 5
D Use of hand cream inside gloves in the cleanroom for medical devices Occupational Health & Safety Management Standards 7
A FDA Class Classification for a cabinet 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
dinaroxentool Question about FDA Classification of a Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N What is our product classification? (Does Unclassified classification still exists) Other US Medical Device Regulations 14
nadhar2 Classification of Action Items Misc. Quality Assurance and Business Systems Related Topics 3
A Class medical device (MDD) - Classification help EU Medical Device Regulations 1
D IEC 62304 Risk Classification - With and without hardware control IEC 62304 - Medical Device Software Life Cycle Processes 2
M Classification of Clean Rooms Other Medical Device Related Standards 9
A ISO 10993-7:2008/ Amd.1:2019 - Classification of special populations Other Medical Device Related Standards 1
DuncanGibbons Classification of aerospace parts depending on their risk and criticality etc. Federal Aviation Administration (FAA) Standards and Requirements 3
N Medical Device Classification under MDR - Rule 21 EU Medical Device Regulations 11
M Risk Classification For Supplier - Clinical Research Organisation (CRO) Supply Chain Security Management Systems 3
D Classification of Syringe (nozzle) of needle free injection system. EU Medical Device Regulations 8
S CE Mark - Classification Confusion EU Medical Device Regulations 12
S Medical Device Product Classification in South Korea Other Medical Device Regulations World-Wide 1
M Informational US FDA – Ear, Nose, and Throat Devices; Classification of the Self-Fitting Air-Conduction Hearing Aid Medical Device and FDA Regulations and Standards News 0
D MDR 2017/745 Classification Guide EU Medical Device Regulations 8
M Informational EU – Guidance on Qualification and Classification of Software in Regulation (EU) 2017/745 – MDR and Regulation (EU) 2017/746 – IVDR Medical Device and FDA Regulations and Standards News 2
dgrainger Informational MDCG 2019-11: Guidance on Qualification and Classification of Software in Regulation (EU) 2017/745 – MDR and Regulation (EU) 2017/746 – IVDR Medical Device and FDA Regulations and Standards News 0
M Informational EU MDR Classification Rule 11 – what??? Medical Device and FDA Regulations and Standards News 9
M Informational US FDA Final Guidance – Acceptance Review for De Novo Classification Requests Medical Device and FDA Regulations and Standards News 1
A Fire hazard classification - Clause in IEC 60601 about gauze and electricity ISO 14971 - Medical Device Risk Management 0
J MDR Annex VIII, Rule 6 Classification - Implication for lower risk CV products? CE Marking (Conformité Européene) / CB Scheme 3
A Device Classification - India CDCSO - How to classify the device? Other Medical Device Related Standards 1
B ASTM F2924-14 /ASTM F3001 - Room temperature classification criteria Other US Medical Device Regulations 0
M Informational EU – MANUAL ON BORDERLINE AND CLASSIFICATION IN THE COMMUNITY REGULATORY FRAMEWORK FOR MEDICAL DEVICES Version 1.22 (05-2019) Medical Device and FDA Regulations and Standards News 2
E Wristwatch Skin-proximal glucose sensor - Classification under MDR EU Medical Device Regulations 4
K MEDDEV 2.4/1 rev. 9 Classification EU Medical Device Regulations 1
M Informational USFDA – Medical Devices; Anesthesiology Devices; Classification of the Ventilatory Electrical Impedance Tomograph Medical Device and FDA Regulations and Standards News 0
N Medical Device Re-Classification EU Medical Device Regulations 2
B Class IIB Device - IEC 62304 Software Classification IEC 62304 - Medical Device Software Life Cycle Processes 13
C Suppliers re-classification from Critical to Significant or from Significant to Non-critical Supply Chain Security Management Systems 0
J Medical Device re-classification and marketing ISO 13485:2016 - Medical Device Quality Management Systems 13
M APQC PCF (Process Classification Framework) and ISO 9001 - Processes Based Approach ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Informational TGA – Several proposed changes to classification to better align with the EU MDR Medical Device and FDA Regulations and Standards News 0
N Medical Device Accessory Classification - Software as a Medical Device Other Medical Device Regulations World-Wide 5
G EU MDR 2017/745 Rule 11 interpretation - Re-classification of a Software as Medical Device Other Medical Device Related Standards 0
M Medical Device News Manual On Borderline And Classification In The Community Regulatory Framework For Medical Devices Medical Device and FDA Regulations and Standards News 0
M FDA News USFDA Final Rule – Medical Device Classification Procedures: Incorporating Food and Drug Administration Safety and Innovation Act Procedures Medical Device and FDA Regulations and Standards News 0
Similar threads


















































Top Bottom