Battery Enclosure Requirements
H
himalayan
Hi all,
I'm new here and already appreciating the wealth of experience and knowledge being shared. I come from a consumer product design background, but have recently made the switch to designing medical devices at quite a young company, and could use a steer in the right direction!
I'm in the early stages of designing a class II medical device, which has an internal Li Ion battery (already approved for use in this context). The product will always remain below the 100VA threshold discussed in clause 13.1 (by means of a fuse), which as I understand it, suggests that the risk of 'ignition' is acceptably low, and I therefore do not have to meet the flammability requirements of Sub Clause 11.3, as the enclosure is not considered a fire enclosure.
I expected there to be a section specifically regarding the risks surrounding batteries and the requirements of a battery enclosure (from a flammability perspective) but I appear to be wrong in that assumption. The battery itself has 3 means of protection in its circuitry. Am I correct in my understanding that the requirements placed on the parts that form an enclosure for the battery, should be determined by a risk analysis? I.e, if through the risk analysis, it is deemed that the 3 means of protection on the battery circuitry, plus the 100VA limit, plus the mechanical strength of the battery enclosure are sufficient to minimize any risk, that's acceptable?
Or, have I missed some section that specifically relates to battery enclosures or misunderstood 60601 entirely?
My initial reaction was that anything enclosing the battery should be UL 94 V0 rated, but I can't seem to see that in the standard. (If it was an option, we'd go for this anyway but for specific reasons, its not that straightforward).
Thanks for any advice!
F MOPP in battery powered device with metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C Lithium battery powered Medical Device - 15.3.3 Impact test, broken enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
D LiPo battery minimum requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Z Is Op Amp Common-Mode Rejection Ratio (CMRR) test applicable to battery operated devices? Other Medical Device Related Standards 0
L EN 62368 (In an Annex - M.4.2.1) Secondary Lithium Battery Charging Safeguards CE Marking (Conformité Européene) / CB Scheme 2
S Battery powered device - electrical protection requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 18
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Q EU Rep for Packaging waste, WEEE and Battery disposal CE Marking (Conformité Européene) / CB Scheme 0
J FDA wants electrical safety testing on battery powered medical device US Food and Drug Administration (FDA) 11
B Need For BIS Standard Mark? Imported OEM's Power supply,Li-Ion Battery Other Medical Device Regulations World-Wide 0
T Battery Powered Device - MOP and Patient Auxiliary Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F IEC 62133 cell vs. Battery Pack - Certification on the cell only IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E IEC 60335-1 li-ion battery requirements Other ISO and International Standards and European Regulations 0
E CB scheme applicability for battery powered medical devices CE Marking (Conformité Européene) / CB Scheme 0
G Assembly of cells into battery - Medical Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Leakage Current test and Dielectric strength for Battery operated ECG device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H Classification of a device that uses an existing battery IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Dielectric Test of Battery Powered Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Electrical safety (battery-powered medical device) EU Medical Device Regulations 1
L Battery requirements for medical device (US/Canada/EU) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Testing ME equipment powered by Battery EU Medical Device Regulations 0
K IEC 62133 Test Report for Lithium Battery Certification IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601 Main requirements for Lithium Ion Battery Powered Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Testing a Low Current (Battery) DC Medical Device Other Medical Device Related Standards 4
M Battery Powered Measurement Equipment and Low Voltage Directive CE Marking (Conformité Européene) / CB Scheme 1
B Primary lithium battery protection and vented gas IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Measuring LiON Battery State of charge - Transporting batteries by Air RoHS, REACH, ELV, IMDS and Restricted Substances 1
W EMC, ESD for battery operated apparatus Other Medical Device and Orthopedic Related Topics 1
A IEC60601-1 - Battery Charger Requirement - Home Healthcare Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F Ambulance Battery Isolation - IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601 Charger requirement for Lithium Ion Battery Powered Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
C Battery Powered Beauty Products minimum Legal Certifications Requirements CE Marking (Conformité Européene) / CB Scheme 7
E Requirements for primary battery powered ME with BF-applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Requirements for Using Lithium battery instead of Alkaline IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
X IEC60601 Battery powered devices intended for measuring bio-impedance IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Demonstrating battery safety in 510(K) submission for the FDA Other US Medical Device Regulations 9
D Battery Powered Medical Device Battery Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
B Flammability Ratings for Internal Battery & Hand-Held ME Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
thisby_ Medical Device Battery Powered Equipment connected to Charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T Polymer Lithium Ion Battery and Ventilation Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
M IEC60601-1 Battery Powered Body Worn Device Leakage Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M IEC 60601-1 testing for device with its own 2V battery power supply necessary? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
T IEC 62133 Certified Battery Manufacturers Other Medical Device Related Standards 3
K Li Ion battery for medical device: transportation safety Other Medical Device Related Standards 2
A Boeing 787 Battery Investigation - NTSB findings to date AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
L Medical Device Battery Labeling Requirements in the US US Food and Drug Administration (FDA) 1
R Does applied part of battery powered TENS unit need to be designed as type BF? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
K NiMH Battery Use in Medical Devices - What Certifications are Required? EU Medical Device Regulations 2
K Small Sterilized Battery Operated Light - Is this a medical Device? ISO 13485:2016 - Medical Device Quality Management Systems 5