Battery overload in IEC 60601-1 vs 62133-2
FrancoisA
Hello
How do you interpret "Overload" in IEC 60601-1 chapter 15.4.3.3 ?
1) As excessive charging current ?
2) As excessive current consumption of the downstream circuit ?
3) As excessive charging time, as defined in IEC 62133-2 chapter 7.3.6 ?
Thanks ;-)
Peter Selvey
15.4.3.3 is "overcharge" not overload. Typically this is due to failure of the charging circuit but can also be a consideration for non-rechargeable cells that are connected into circuits that are powered independently, and faults in the circuit could lead to reverse charging. The need for and type of protection against overcharging varies depending on type of battery.
FrancoisA
FYI, the French version uses the term "surcharge" which translates to either "overcharge" or "overload", so my mistake. The English version is less ambiguous ;-)
F Interpretation of 60601-1 Figure J.5 (Isolation, example 5) for a rechargeable, battery operated device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
L Single-cell lithium ion battery regulations IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
M MOPs of power socket for battery powered device IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
G User Ejecting Lithium Battery from Device Other Medical Device and Orthopedic Related Topics 4
I Lithium Battery removal ISO 14971 - Medical Device Risk Management 1
F Lithium polymer battery, venting and CE Marking CE Marking (Conformité Européene) / CB Scheme 4
D LiPo battery minimum requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
Z Is Op Amp Common-Mode Rejection Ratio (CMRR) test applicable to battery operated devices? Other Medical Device Related Standards 0
L EN 62368 (In an Annex - M.4.2.1) Secondary Lithium Battery Charging Safeguards CE Marking (Conformité Européene) / CB Scheme 2
S Battery powered device - electrical protection requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 18
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Q EU Rep for Packaging waste, WEEE and Battery disposal CE Marking (Conformité Européene) / CB Scheme 0
J FDA wants electrical safety testing on battery powered medical device US Food and Drug Administration (FDA) 11
B Need For BIS Standard Mark? Imported OEM's Power supply,Li-Ion Battery Other Medical Device Regulations World-Wide 0
T Battery Powered Device - MOP and Patient Auxiliary Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F IEC 62133 cell vs. Battery Pack - Certification on the cell only IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E IEC 60335-1 li-ion battery requirements Other ISO and International Standards and European Regulations 0
E CB scheme applicability for battery powered medical devices CE Marking (Conformité Européene) / CB Scheme 0
F MOPP in battery powered device with metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
G Assembly of cells into battery - Medical Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Leakage Current test and Dielectric strength for Battery operated ECG device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H Battery Enclosure Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
C Lithium battery powered Medical Device - 15.3.3 Impact test, broken enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H Classification of a device that uses an existing battery IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
B Dielectric Test of Battery Powered Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Electrical safety (battery-powered medical device) EU Medical Device Regulations 1
L Battery requirements for medical device (US/Canada/EU) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Testing ME equipment powered by Battery EU Medical Device Regulations 0
K IEC 62133 Test Report for Lithium Battery Certification IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S IEC 60601 Main requirements for Lithium Ion Battery Powered Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Testing a Low Current (Battery) DC Medical Device Other Medical Device Related Standards 4
M Battery Powered Measurement Equipment and Low Voltage Directive CE Marking (Conformité Européene) / CB Scheme 1
B Primary lithium battery protection and vented gas IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Measuring LiON Battery State of charge - Transporting batteries by Air RoHS, REACH, ELV, IMDS and Restricted Substances 1
W EMC, ESD for battery operated apparatus Other Medical Device and Orthopedic Related Topics 1
A IEC60601-1 - Battery Charger Requirement - Home Healthcare Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F Ambulance Battery Isolation - IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M IEC 60601 Charger requirement for Lithium Ion Battery Powered Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
C Battery Powered Beauty Products minimum Legal Certifications Requirements CE Marking (Conformité Européene) / CB Scheme 7
E Requirements for primary battery powered ME with BF-applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
E Requirements for Using Lithium battery instead of Alkaline IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
X IEC60601 Battery powered devices intended for measuring bio-impedance IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Demonstrating battery safety in 510(K) submission for the FDA Other US Medical Device Regulations 9
D Battery Powered Medical Device Battery Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
B Flammability Ratings for Internal Battery & Hand-Held ME Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
thisby_ Medical Device Battery Powered Equipment connected to Charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
T Polymer Lithium Ion Battery and Ventilation Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
M IEC60601-1 Battery Powered Body Worn Device Leakage Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
M IEC 60601-1 testing for device with its own 2V battery power supply necessary? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1