Battery Powered Disposable Pump - What IEC 60601 testing is required, if any?

D

deviceguy

#1
My company is developing a medical device that uses a battery powered DC (brush) motor to drive a pump to deliver saline to the catheter. There is no electrical patient connection. The entire unit is disposed of after the procedure. the batteries are alkaline and are removable for proper disposal. What IEC 60601 requirements would I be expected to meet?

thanks.
 
Elsmar Forum Sponsor

Peter Selvey

Staff member
Moderator
#2
There would be a few applicable requirements such as marking, instruction manual and handling of batteries, but the bulk of the standard would not be applicable (or at least, not applicable in a way that significantly influences the design).

Unfortunately, despite the standard being mostly not applicable, the only way to handle it is to check all items one by one to confirm what is or is not applicable. For the 2nd edition there are about 450-500 discrete items, for the 3rd edition around 1,500.

For an experienced test engineer, it would take around 2-3hrs to scan all the requirements (maybe 4-5 for the 3rd edition).

I suggest to shop around for a test lab that will not charge a huge amount, but keep in mind paperwork takes time no matter how simple the task. Also, ask the test lab how they will handle the 3V battery part. Sometimes they may go into overkill on the electrical isolation part. Look for a lab that applies some common sense.
 
A

AmirBaron

#3
[FONT=Arial,Bold][FONT=Arial,Bold]Don't forget to check if 60601-2-24 is relevent for your device.[/FONT][/FONT]
[FONT=Arial,Bold][FONT=Arial,Bold]It's a particular requirements for the safety of infusion[/FONT]

[FONT=Arial,Bold]pumps and controllers.[/FONT]

[/FONT]
 
E

ejacob3

#5
I have a similar application but instead am applying suction to a catheter (non-venous). I have a AA battery powered pump that is not applying any sort of voltages to the patient nor is even considered an applied part.

Do I have to consider other than the -8 disposal:
EMC, leakage to operator etc?
 
D

deviceguy

#6
You may be surprised by what is considered an applied part. It is often easier to do the testing than justify the exemption from the clause. You will need to consider the entire standard and produce a base document that defines your position on inclusion or exclusion based on the type of device. You will need to do EMC, RFI all the standard stuff as well as risk management for 3rd edition which can be a pain. You new risk documents (FMEA etc) should consider all hazards described in the standard, even if you don't test to them you still need to examine them.

Hope this helps.
 
Thread starter Similar threads Forum Replies Date
S Battery powered device - electrical protection requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
J FDA wants electrical safety testing on battery powered medical device US Food and Drug Administration (FDA) 11
T Battery Powered Device - MOP and Patient Auxiliary Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E CB scheme applicability for battery powered medical devices CE Marking (Conformité Européene) / CB Scheme 0
F MOPP in battery powered device with metal enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
C Lithium battery powered Medical Device - 15.3.3 Impact test, broken enclosure IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
B Dielectric Test of Battery Powered Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Electrical safety (battery-powered medical device) EU Medical Device Regulations 1
E Testing ME equipment powered by Battery EU Medical Device Regulations 0
S IEC 60601 Main requirements for Lithium Ion Battery Powered Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
M Battery Powered Measurement Equipment and Low Voltage Directive CE Marking (Conformité Européene) / CB Scheme 1
M IEC 60601 Charger requirement for Lithium Ion Battery Powered Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
C Battery Powered Beauty Products minimum Legal Certifications Requirements CE Marking (Conformité Européene) / CB Scheme 7
E Requirements for primary battery powered ME with BF-applied parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
X IEC60601 Battery powered devices intended for measuring bio-impedance IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
W Is IEC 60601-1-2 compliance/testing required for battery powered Class I Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
D Battery Powered Medical Device Battery Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
thisby_ Medical Device Battery Powered Equipment connected to Charger IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M IEC60601-1 Battery Powered Body Worn Device Leakage Currents IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
R Does applied part of battery powered TENS unit need to be designed as type BF? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
S Medical Device Battery Powered Equipment with Battery Charger Classification IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
Z Is Op Amp Common-Mode Rejection Ratio (CMRR) test applicable to battery operated devices? Other Medical Device Related Standards 0
L EN 62368 (In an Annex - M.4.2.1) Secondary Lithium Battery Charging Safeguards CE Marking (Conformité Européene) / CB Scheme 2
A IEC 60601-1 Dielectric Strength test for battery operated devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
Q EU Rep for Packaging waste, WEEE and Battery disposal CE Marking (Conformité Européene) / CB Scheme 0
B Need For BIS Standard Mark? Imported OEM's Power supply,Li-Ion Battery Other Medical Device Regulations World-Wide 0
F IEC 62133 cell vs. Battery Pack - Certification on the cell only IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E IEC 60335-1 li-ion battery requirements Other ISO and International Standards and European Regulations 0
G Assembly of cells into battery - Medical Devices IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Leakage Current test and Dielectric strength for Battery operated ECG device? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
H Battery Enclosure Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
H Classification of a device that uses an existing battery IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
L Battery requirements for medical device (US/Canada/EU) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
K IEC 62133 Test Report for Lithium Battery Certification IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Testing a Low Current (Battery) DC Medical Device Other Medical Device Related Standards 4
B Primary lithium battery protection and vented gas IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Measuring LiON Battery State of charge - Transporting batteries by Air RoHS, REACH, ELV, IMDS and Restricted Substances 1
W EMC, ESD for battery operated apparatus Other Medical Device and Orthopedic Related Topics 1
A IEC60601-1 - Battery Charger Requirement - Home Healthcare Medical Device IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
F Ambulance Battery Isolation - IEC 60601 IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
E Requirements for Using Lithium battery instead of Alkaline IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
M Demonstrating battery safety in 510(K) submission for the FDA Other US Medical Device Regulations 9
B Flammability Ratings for Internal Battery & Hand-Held ME Equipment IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
T Polymer Lithium Ion Battery and Ventilation Requirements IEC 60601 - Medical Electrical Equipment Safety Standards Series 14
M IEC 60601-1 testing for device with its own 2V battery power supply necessary? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
T IEC 62133 Certified Battery Manufacturers Other Medical Device Related Standards 3
K Li Ion battery for medical device: transportation safety Other Medical Device Related Standards 2
A Boeing 787 Battery Investigation - NTSB findings to date AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
L Medical Device Battery Labeling Requirements in the US US Food and Drug Administration (FDA) 1
K NiMH Battery Use in Medical Devices - What Certifications are Required? EU Medical Device Regulations 2

Similar threads

Top Bottom